  • Oliviervg
    Member
    • Sep 2011
    • 13

    #1

    Scriptseq V2 : fragmentation problem ?

    Hi,

    We want to study the transcriptome of breast cancer cell lines and tissue, so we want to do RNA-seq experiment.
    We prepare our sample with Tripure extraction, clean up (Qiagen), DNAse treatment. We check the quality with the Bioanalyser, our samples have a RIN of about 8 - 8,5.
    Then we use the Ribozero kit with 2µg of total RNA as input. We use the clean and concentrator (Zymo) to clean our sample. We obtain about 12ng of ribodepleted RNA (0,6ng/µl in 20µl). We check our samples with the Bioanalyser (see http://www.fichier-pdf.fr/2012/01/11...bozero-090112/), everything seems ok.
    Then we prepare the library (scriptseq V2) with 9,5 µl of our ribodepleted RNA, we exactly follow the protocol (We purify the cDNA with the MinElute PCR purification kit in step 4.D., and we purify the RNA-seq library using AMPure XP purification (step 4.F.) , and we perform 15 cycles for the PCR)
    After that, we quantify with the Qubit, and we obtain +/- 3 to 5 ng/µl of DNA (total amount 60 to 100 ng, it's less than expected).

    When we put 1µl of our sample on the Bioanalyser (DNA 1000), we see nothing, no peak, no concentration.

    We tried one qPCR with this DNA (+/- 5ng DNA per well) , and we obtain good Cq values (we tried Actin (Cq = 17,2), 18S (Cq = 25,5), 28S (Cq = 22,1).

    So we suppose that the fragmentation step didn't work well.
    We did all the experiment twice, and we obtain similar result.

    What do you think about that ?
    Has someone else use this kit ? If yes, do you have any problem ?

    Thank you

    Best regards,
    Last edited by Oliviervg; 01-11-2012, 08:50 AM.
  • huguesparri
    Member
    • May 2008
    • 97

    #2
    I'm not familiar with the scriptseq V2 so what I'm going to say might be silly but did you try to run your library on an Agilent HiSensitivity Chip?
    Maybe your concentration is too low to be detected on DNA1000 Chips...

    Comment

    • Oliviervg
      Member
      • Sep 2011
      • 13

      #3
      Thank you for your answer,

      After the library preparation we get more than 3 ng cDNA / µl, that is sufficient for an analysis with the DNA 1000 on the BioAnalyser.

      We 'll try an other control on a DNA 7000 chip, and hope to see something.

      Does someone already use this kit ? Did you have problem ?

      Comment

      • susma
        Member
        • Mar 2013
        • 11

        #4
        Hi,
        I am using this kit now. It was my first time, but my RNA source is bacterial. I put 40 ng rRNA depleted mRNA as input and I got 30 ng/ul (Total volume 20 ul) DNA at the end, then I run an Agilent HiSensitivity Chip but I got strange results with the fragmentation. I got many fragments!

        I am trying now to ask them what might be wrong.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
          by SEQadmin2



          CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

          Despite this, “CRISPR helped turn genome editing from a specialized technique into
          ...
          07-31-2026, 11:01 AM
        • SEQadmin2
          Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
          by SEQadmin2


          Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

          The systematic characterization of the human proteome has
          ...
          07-20-2026, 11:48 AM
        • SEQadmin2
          Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
          by SEQadmin2



          Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
          ...
          07-09-2026, 11:10 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, Today, 10:13 AM
        0 responses
        13 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-31-2026, 02:55 AM
        0 responses
        23 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-24-2026, 12:17 PM
        0 responses
        19 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 07-23-2026, 11:41 AM
        0 responses
        18 views
        0 reactions
        Last Post SEQadmin2  
        Working...