Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #1

    MiSeq intermittent blurry images.

    Here is a quality plot of our ongoing MiSeq validation run:



    Overall pretty good. But, if you look at one of the more poorly performing tiles you see:



    Seems largely to be a tile-specific issue:



    Any ideas what is causing this? Vibration seems unlikely because I would expect intermittent vibrations to effect the tiles randomly.

    --
    Phillip
  • ECO
    --Site Admin--
    • Oct 2007
    • 1360

    #2
    I would guess that your stage tilt is out of spec or very close. Pretty quick fix for an FSE and tech support will replace the reagents. We had this problem on a couple of ours but after a secondary checkup they've been fine.

    Whenever we see "swirly" images like this, that was the cause.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Originally posted by ECO View Post
      I would guess that your stage tilt is out of spec or very close. Pretty quick fix for an FSE and tech support will replace the reagents. We had this problem on a couple of ours but after a secondary checkup they've been fine.

      Whenever we see "swirly" images like this, that was the cause.
      Thanks. Actually the FAS did the install. I'll mention this to her. (It was just the phiX validation run, so we are not out any reagents.)

      Phillip

      Comment

      • ECO
        --Site Admin--
        • Oct 2007
        • 1360

        #4
        I know...same with ours...the cluster density and other metrics may still pass their specs even with some of the image out of focus. The FAS/FSE had to come back and readjust the stage post install.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by ECO View Post
          I would guess that your stage tilt is out of spec or very close. Pretty quick fix for an FSE and tech support will replace the reagents. We had this problem on a couple of ours but after a secondary checkup they've been fine.

          Whenever we see "swirly" images like this, that was the cause.
          Wow, you are good. FAS ran a check and stage tilt failed! It had passed during the install, but somehow got out of alignment. Should be fixed now.

          We did a 2x25 nt run a couple of days back with the stage tilt presumably out of alignment -- worked fine. Could see no trace of this issue. Probably not that far off spec. Spec is -400 to 400 and the tilt was in the 700s when checked yesterday before it was re-aligned.

          Actually my FAS wants to contact your FAS/FSE to find out if there is someway to lock that alignment down. But maybe it just happens sometimes during instrument "burn in"...

          --
          Phillip

          Comment

          • ECO
            --Site Admin--
            • Oct 2007
            • 1360

            #6
            Originally posted by pmiguel View Post
            Actually my FAS wants to contact your FAS/FSE to find out if there is someway to lock that alignment down. But maybe it just happens sometimes during instrument "burn in"...
            Marc Rosoff is our guy.

            Seems to be starting to happen on the second of our newest instruments as well...I'm hoping something is just loose from initial assembly and not something that will continually go out of spec.

            Saddest part is that even when our runs start to drop to <70% cluster PF because of this problem, the instrument still produces enough in-spec data that ILMN won't replace the reagents.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2



              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection of antimicrobial...
              Today, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            22 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            14 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            67 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            47 views
            0 reactions
            Last Post SEQadmin2  
            Working...