Cluster Generation Failure
Several things:
1) We also had a valve failure, this is supposedly due to crystals forming in one of the valves and grinding it down - this led to the new wash protocol recommendations in MCS v1.3. It seems this was an endemic problem to the MiSeq.
2) Since getting MCS v1.3, we've noticed a couple of other bugs - wondering if anyone else has the same problem. For one, it constantly alerts that it may need a standby wash because the fluidics haven't been used, even when this is not true. They tell us its a known bug. Second issue is the volume conductance test - it seemed to work fine to diagnose the valve problem when we had v1.2, now they tell us we have to run it twice. The first time it reports everything as failed, then you run it again and it still reports a PR2 failure, but my tech support rep assures me, again, these are "known software bugs" and everything is fine.
3) We've had 2 50 cycle kits fail to generate any clusters (using NexteraXT libraries), and subsequent fluidics tests and reruns have been "fine". The reps ascribe this to a "one time faiure" and happily replace the kits, but we're wondering if there is a bad lot of flowcells or reagent cartridges because it has happened twice now with 50 cycle kits from the same purchase, both expiring around 8/16/2012. The frustrating thing is that due to the nature of the XT libraries being single stranded, there is a significant freeze-thaw lesion when you rerun. They report 10-20% and we have experienced losses equal to 20% or greater in our two rerun experiments.
So I'm wondering if any one else had any 50 cycle kits expiring in August simply fail to generate clusters. The reps seem quick to tell us the instrument is fine and replace the kits, so I wonder if they know more than they're telling us. Both times were the same - no clusters, blank images on the first cycle, run failure. Repeat run generated good clusters.
Frustrating.
Header Leaderboard Ad
Collapse
miSeq failures during cluster generation
Collapse
Announcement
Collapse
SEQanswers June Challenge Has Begun!
The competition has begun! We're giving away a $50 Amazon gift card to the member who answers the most questions on our site during the month. We want to encourage our community members to share their knowledge and help each other out by answering questions related to sequencing technologies, genomics, and bioinformatics. The competition is open to all members of the site, and the winner will be announced at the beginning of July. Best of luck!
For a list of the official rules, visit (https://www.seqanswers.com/forum/sit...wledge-and-win)
For a list of the official rules, visit (https://www.seqanswers.com/forum/sit...wledge-and-win)
See more
See less
X
-
It appears our problem was a valve failure -- this may be a somewhat common problem with the MiSeq?
Illumina replaced the valve and recommended a new wash procedure. The last couple of runs have generated clusters without error. Hopefully that will remain the case. Thanks for all replies.
Leave a comment:
-
Quantify your libraries by QPCR. This is the only way to get consistent clustering results.
Leave a comment:
-
We've also had this issue on multiple runs. Try running the volume test. When this happened to us it was typically a result of a valve failure.
Leave a comment:
-
Thanks. If our clusters were too dense, likely would have been by a factor of 2-4. If 25X excess library didn't trip up the miSeq then I doubt that is our problem.
Leave a comment:
-
Originally posted by greenhilly View PostNo images were generated. Consistent with no clusters -- might the same thing occur if clusters were way too dense?
Here is a shot:
There is a second issue in play that caused the distortions in some of the panes. But I think the center pane is what your MiSeq flowcell tile will look like if you cluster at 250 pM.
It did give a "raw" cluster density number. But none of the clusters passed filter.
--
Phillip
Leave a comment:
-
Ive never gotten it when we've gone too dense, usually that manifests as very low PF.
Without knowing the history of this machine and these samples...spike in 50% phiX and see what happens.
Leave a comment:
-
No images were generated. Consistent with no clusters -- might the same thing occur if clusters were way too dense?
Leave a comment:
-
Post some thumbnails from the first images....if there aren't any, it's likely there were no clusters to find focus. If the images are there and blurry, but density looks good...it's probably an instrument problem.
Leave a comment:
-
miSeq failures during cluster generation
We've experienced a number of miSeq run failures during/immediately after cluster generation. The errors in the log are pasted below. Anybody else get these errors? Anyone know how to interpret them?
Illumina says these are consistent with cluster generation failure, potentially due to low-complexity libraries or other library problems. Curious whether anyone else gets a lot of these, and on their interpretation. FYI running poly-A selected RNA-seq libraries generated with the TruSeq RNA kit, which look great in all the QC we've performed and quantify consistently with KAPA. Have never gotten these errors with phiX, suggesting it is something about the libraries (though we don't have enough data points to state this with high confidence).
Thanks for any help/suggestions.
12-05-02 13:39:55.950 -04 MiSeq Control Software 000000029 WRN: !!!!! Warning Occurred: Z Motor: Z Motor: Position error: Target = -0.196095, Position = -0.036164, Error = 0.159931, min/max error = -0.000031/0.159931, tolerance = 0.004001 mm
12-05-02 13:39:55.950 -04 MiSeq Control Software 000000029 ERR: Z Motor: Z Motor: Position error: Target = -0.196095, Position = -0.036164, Error = 0.159931, min/max error = -0.000031/0.159931, tolerance = 0.004001 mm
12-05-02 13:39:55.950 -04 MiSeq Control Software 000000029 WRN: !!!!! Warning Occurred: Z Motor: MoveAbsStepped failed: Z Motor: Position error: Target = -0.196095, Position = -0.036164, Error = 0.159931, min/max error = -0.000031/0.159931, tolerance = 0.004001 mm
12-05-02 13:39:55.950 -04 MiSeq Control Software 000000029 ERR: Z Motor: MoveAbsStepped failed: Z Motor: Position error: Target = -0.196095, Position = -0.036164, Error = 0.159931, min/max error = -0.000031/0.159931, tolerance = 0.004001 mm
12-05-02 13:39:55.950 -04 MiSeq Control Software 000000029 WRN: !!!!! Warning Occurred: Z Motor: MoveAbsScaled failed: Z Motor: MoveAbsStepped failed
12-05-02 13:39:55.950 -04 MiSeq Control Software 000000029 ERR: Z Motor: MoveAbsScaled failed: Z Motor: MoveAbsStepped failed
1Tags: None
Latest Articles
Collapse
-
by seqadmin
Developments in sequencing technologies and methodologies have transformed the field of epigenetics, giving researchers a better way to understand the complex world of gene regulation and heritable modifications. This article explores some of the diverse sequencing methods employed in the study of epigenetics, ranging from classic techniques to cutting-edge innovations while providing a brief overview of their processes, applications, and advances.
Methylation Detect...-
Channel: Articles
05-31-2023, 10:46 AM -
-
Differential Expression and Data Visualization: Recommended Tools for Next-Level Sequencing Analysisby seqadmin
After covering QC and alignment tools in the first segment and variant analysis and genome assembly in the second segment, we’re wrapping up with a discussion about tools for differential gene expression analysis and data visualization. In this article, we include recommendations from the following experts: Dr. Mark Ziemann, Senior Lecturer in Biotechnology and Bioinformatics, Deakin University; Dr. Medhat Mahmoud Postdoctoral Research Fellow at Baylor College of Medicine;...-
Channel: Articles
05-23-2023, 12:26 PM -
-
by seqadmin
Continuing from our previous article, we share variant analysis and genome assembly tools recommended by our experts Dr. Medhat Mahmoud, Postdoctoral Research Fellow at Baylor College of Medicine, and Dr. Ming "Tommy" Tang, Director of Computational Biology at Immunitas and author of From Cell Line to Command Line.
Variant detection and analysis tools
Mahmoud classifies variant detection work into two main groups: short variants (<50...-
Channel: Articles
05-19-2023, 10:03 AM -
ad_right_rmr
Collapse
News
Collapse
Topics | Statistics | Last Post | ||
---|---|---|---|---|
Started by seqadmin, Today, 07:14 AM
|
0 responses
4 views
0 likes
|
Last Post
by seqadmin
Today, 07:14 AM
|
||
Started by seqadmin, Yesterday, 01:08 PM
|
0 responses
6 views
0 likes
|
Last Post
by seqadmin
Yesterday, 01:08 PM
|
||
Started by seqadmin, 06-01-2023, 08:56 PM
|
0 responses
12 views
0 likes
|
Last Post
by seqadmin
06-01-2023, 08:56 PM
|
||
Deep Sequencing Unearths Novel Genetic Variants: Enhancing Precision Medicine for Vascular Anomalies
by seqadmin
Started by seqadmin, 06-01-2023, 07:33 AM
|
0 responses
149 views
0 likes
|
Last Post
by seqadmin
06-01-2023, 07:33 AM
|
Leave a comment: