    Maximum fragment size

    Last thread on a similar topic was back in 2008 so figured I'd check if anything has changed - to save time whilst performing a genomic library prep (TruSeq) I decided to try a 0.6x AMPure cleanup instead of gel-cutting for size-selection. Resulting libraries are a bit bigger than I was expecting - around 700bp enriched product so insert size of ~600bp! Are fragments this size going to cluster correctly?

    I suppose I could burn a MiSeq run to find out before risking them on the HiScan... but any advice appreciated!
  • TonyBrooks
    Senior Member
    • Jun 2009
    • 303

    #2
    They should still cluster OK. If you want to err on the side of caution, you can load at a slightly lower density to compensate for the marginally larger clusters. Although honestly, I don't think that they are sufficiently larger than normal to cause a problem. This only really seems to happen when you get over 1kb inserts.
    Remember than Illumina will (probably) be rolling out 400bp PE on MiSeq soon which will need >700bp inserts to get the most out of the sequencing.

    Comment

    • matth431
      Member
      • Jan 2012
      • 36

      #3
      Originally posted by TonyBrooks View Post
      They should still cluster OK. If you want to err on the side of caution, you can load at a slightly lower density to compensate for the marginally larger clusters. Although honestly, I don't think that they are sufficiently larger than normal to cause a problem. This only really seems to happen when you get over 1kb inserts.
      Remember than Illumina will (probably) be rolling out 400bp PE on MiSeq soon which will need >700bp inserts to get the most out of the sequencing.
      Many thanks - I did wonder regarding the 2x400bp reads but assumed users might also want a bit of overlap between the two reads for error-checking. I have a freebie MiSeq kit I can use to check, will perhaps try a 7pM load to space the clusters a bit.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        For a more extreme example see this thread.

        --
        Phillip

        Comment

        • matth431
          Member
          • Jan 2012
          • 36

          #5
          Thanks again - looks like it should work then but possibly exclude some of the larger inserts... that's not really an issue for these samples (just doing some genomic shotgun to look for SNPs) but worth bearing in mind for future.

          Interesting that the 0.5x AMPure in that case gave such huge fragment sizes - wouldn't have expected such a difference (nearly 3-fold bigger!) to the range I got here, although my DNA was sheared for 500bp mean size...

          Comment

          • hawaii454-0
            Member
            • Dec 2010
            • 23

            #6
            Hello,
            What do you mean by a 0.6x ampure clean up? 1:0.6 ratio?

            Thanks,
            Anna.

            Comment

            • TonyBrooks
              Senior Member
              • Jun 2009
              • 303

              #7
              If your reaction size is 100uL, then add 60uL of Ampure beads during clean up as opposed to the standard 180uL (1.8x). Keep the ratio the same (30uL of beads for 50uL reaction clean ups).
              By reducing the volume of beads, you reduce the buffer concentration which moves the binding titration point up (lower salt conc in the reaction means only more negatively charged DNA (i.e. longer) bind). The beads themselves shouldn't be an issue as 1µL of beads could potentially bind >1ug of DNA.

              Comment

              • hawaii454-0
                Member
                • Dec 2010
                • 23

                #8
                Thanks Tony.

                Does anyone have final libraries which show shoulder peaks on their larger insert libraries?

                Thanks,
                Anna.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by TonyBrooks View Post
                  If your reaction size is 100uL, then add 60uL of Ampure beads during clean up as opposed to the standard 180uL (1.8x). Keep the ratio the same (30uL of beads for 50uL reaction clean ups).
                  By reducing the volume of beads, you reduce the buffer concentration which moves the binding titration point up (lower salt conc in the reaction means only more negatively charged DNA (i.e. longer) bind).
                  The salient component of the "buffer" here is PEG (with NaCl also at high concentration--polynucleotides don't seem to precipitate as free acids, but do as salts). Like alcohols, PEG is said to "occupy" some percentage of the solvation capability of water. The effect is that longer molecules come out of solution preferentially. As more and more of the solvation capacity of water is devoted to PEG (or alcohol), shorter and shorter molecules of polynucleotides can be forced from solution.

                  The binding of this precipitant to the carboxyl groups on the surface of these magnetic beads is another issue, and one that I have not seen an explanation for. But it does seem that this surface has some affinity for polynucleotides. But really it is just a faster method of collecting your precipitate than a hard spin in a centrifuge.

                  I have seen at least one report on seqanswers of using an alcohol precipitation with an Ampure beads collection of the precipitated DNA.

                  --
                  Phillip

                  Comment

                  • captainentropy
                    Member
                    • Mar 2009
                    • 89

                    #10
                    Originally posted by pmiguel View Post
                    The binding of this precipitant to the carboxyl groups on the surface of these magnetic beads is another issue, and one that I have not seen an explanation for.
                    Once I read this post on core-genomics it began to make sense to me how it's working

                    Someone recently asked me, “how do SPRI beads work” and I realized I was not completely sure so I went to find out. My lab uses kits. Lots...

                    Comment

                    • pmiguel
                      Senior Member
                      • Aug 2008
                      • 2328

                      #11
                      Originally posted by captainentropy View Post
                      Once I read this post on core-genomics it began to make sense to me how it's working

                      http://core-genomics.blogspot.com/20...70126511809493
                      Yes, later I quoted the same comment in another thread, here. At the time, I was less than convinced by the commenter's argument. However, re-reading the comment now, I am liking it more.

                      The original SPRI paper, Nucl. Acids Res. (1994) 22 (21): 4543-4544 doi:10.1093/nar/22.21.4543, made it sound like the carboxyl surface of the beads being used was what bound the DNA. But if the authors merely tried a few surface derivatives of these beads and went with the one that gave the highest yields then there is another explanation. That is that any beads derivatized with positively charged molecules would more-or-less irreversibly bind the DNA and therefore appear to give a poor yield. The carboxyl groups would weakly repel the DNA and hence allow for easy elution.

                      This is basically what the anonymous commenter you link to is writing.

                      --
                      Phillip

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM
                      • SEQadmin2
                        New Genomics Technologies Take Aim at Long-Standing Limits
                        by SEQadmin2


                        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                        ...
                        09-28-2026, 10:25 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, Yesterday, 11:13 AM
                      0 responses
                      16 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, Yesterday, 10:48 AM
                      0 responses
                      15 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-15-2026, 10:52 AM
                      0 responses
                      35 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-14-2026, 05:39 AM
                      0 responses
                      42 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...