  • bioinfosm
    Senior Member
    • Jan 2008
    • 483

    #1

    75bp: high error rates on read2

    Is this common? Or there are ways around getting high error rates after cycle 50 for long reads? Especially read2 looks pretty bad!
    Attached Files
    --
    bioinfosm
  • basickler
    Member
    • Mar 2009
    • 26

    #2
    I need a little more info on this. What are you aligning to? What chemistry are you using? Is this PhiX? If you're using the v2 kits and this is PhiX than yes these are very high, it should be below 5% at cycle 75 for both reads on a decent run.

    The reason for this could range from degraded reagents, to out of focus images to a bad turn around on the PE module, etc. etc.

    Brad

    Comment

    • bioinfosm
      Senior Member
      • Jan 2008
      • 483

      #3
      Thanks Brad.
      Its PE mRNA seq genomic DNA on human genome as reference. These are not PhiX, but human mRNA samples.
      Last edited by bioinfosm; 09-09-2009, 08:33 AM. Reason: These are genomic DNA
      --
      bioinfosm

      Comment

      • basickler
        Member
        • Mar 2009
        • 26

        #4
        Well first off you might want to use the human cDNA library as a reference genome. The human genomic reference isn't going to work with ELAND very well. Eland doesn't support gaps (read introns). This still doesn't explain the higher error rates on the paired end read, that's probably a chemistry or technician issue.

        Brad

        Comment

        • cnicolet
          Member
          • Dec 2008
          • 35

          #5
          This is typically what I see regardless of the run length we do, ie, similar curves to what you posted for our 60 x 60 and 85 x 85 runs. I don't think it's a technician issue. Because the curve does look proportionally similar (ie, starts to rise at about the same percentage into the read, not at a particular base) I think it may be a phasing or base calling issue. I'm also suspecting reagent longevity--our next long run will use freshly made topped off reagents like the old days just to see if that helps.

          Comment

          • bioinfosm
            Senior Member
            • Jan 2008
            • 483

            #6
            Originally posted by basickler View Post
            Well first off you might want to use the human cDNA library as a reference genome. The human genomic reference isn't going to work with ELAND very well. Eland doesn't support gaps (read introns). This still doesn't explain the higher error rates on the paired end read, that's probably a chemistry or technician issue.

            Brad
            Sorry for the confusion, these are actually genomic DNA samples, not mRNA. I corrected that in my post above
            --
            bioinfosm

            Comment

            • bioinfosm
              Senior Member
              • Jan 2008
              • 483

              #7
              paired end genomic human dna - 2nd read!

              Hi,

              Do people have comments on this sort of error graph from GA pipeline for a particular tile, running paired 75bp genomic human dna?

              Thanks,
              Attached Files
              --
              bioinfosm

              Comment

              • der_eiskern
                Member
                • Jul 2009
                • 46

                #8
                Originally posted by bioinfosm View Post
                Is this common? Or there are ways around getting high error rates after cycle 50 for long reads? Especially read2 looks pretty bad!
                if i recall correctly we've noticed higher error rates in the second read and attributed it to the lower signal-to-background ratio. i hear that PE constraints on mapping help make up for this problem but i've never seen the data to support that directly.

                Comment

                • GenoMax
                  Senior Member
                  • Feb 2008
                  • 7142

                  #9
                  Originally posted by bioinfosm View Post
                  Is this common? Or there are ways around getting high error rates after cycle 50 for long reads? Especially read2 looks pretty bad!
                  Did this turn out to be a rare observation? These error rates seem rather high. We generally see error rates below 1.5% for 2x76 runs with human samples.

                  Comment

                  • bioinfosm
                    Senior Member
                    • Jan 2008
                    • 483

                    #10
                    yes it was just one of those tiles. We are also getting around 1-2% error rates for 2x76 runs
                    --
                    bioinfosm

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM
                    • SEQadmin2
                      New Genomics Technologies Take Aim at Long-Standing Limits
                      by SEQadmin2


                      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                      ...
                      09-28-2026, 10:25 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, Today, 09:57 AM
                    0 responses
                    1 view
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, Today, 09:51 AM
                    0 responses
                    5 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    62 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    44 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...