Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • koadman
    Member
    • May 2010
    • 65

    #1

    Nextera XT multiplex sample normalization

    Hi all,
    Our lab just did our first Nextera XT run on a MiSeq. I'm just the informatics guy and didn't actually prep the libraries myself, but the person who did says the protocol is nicely straightforward compared to some earlier library protocols.

    The resulting run, unfortunately, had a fairly wide range in sample abundances and I'm wondering if any of you have insight into why or what we might do differently to achieve uniform abundances? I've pasted a histogram of per-sample read counts below. Input material was normalized to 1ng/sample as measured with flourimetry (picogreen or qubit).


    There is a 10-fold difference between the least abundant sample and the most abundant. My (very limited) understanding of the protocol is that normalization happens at the last stage of library prep where tagmented samples are loaded with magnetic beads and shaken at 1800rpm. A limited quantity of sample material is supposed to saturate the binding capacity of the beads in each tube. When the material is released from the beads one obtains about the same amount of material from each sample.

    Does that sound right? Any ideas on what we could do differently to achieve better normalization? Are the results I have here considered "good" normalization?
  • MrGuy
    Member
    • Mar 2009
    • 68

    #2
    Surface capture normalization is based on the concept of overloading to normalize. For this to work, the beads capture X molecules. If you add >X, then you get normalized samples as the excess molecules are discarded in washing. If you add <X, then you get not so normalized samples as the beads capture "all" available molecules.

    My guess is you are below X as you normalized to 1 ng prior to input. You are also gambling on the lot-to-lot reproducibility of the kit at that low of a range. What does the manufacturer recommend as input range?

    Comment

    • koadman
      Member
      • May 2010
      • 65

      #3
      Thanks for the reply MrGuy, it hadn't occurred to me that we might be underloading some samples. The Nextera XT protocol had us tagment 1ng material per sample, then amplify that material with 12 cycles PCR before carrying out the normalization, so in principle there should have been much more than 1ng per sample going into normalization. However, the person making the libs reserved a portion of the PCR product so perhaps it's as simple as just reserving less next time.

      Comment

      • ScottC
        Senior Member
        • Jan 2008
        • 244

        #4
        I'm not saying that this is the solution to your problem, but a couple of things come to mind as possibilities:

        I'd measure the amount of material in the PCR tube before and after cycling to make sure you're actually getting something produced from the PCR reaction and that the yield is good enough. Also measure the size distribution on a bioanalyzer after 'tagmentation' and/or PCR. Sometimes the size range with different samples is very large (i.e. libs with fragments down near 100b, others at 2kb) and this might affect the normalisation too; but, I haven't actually measured the effect of fragment size on normalisation...

        Cheers,

        Scott.

        Comment

        • koadman
          Member
          • May 2010
          • 65

          #5
          Thanks for the reply Scott. Based on your suggestion I decided to look at the relationship between median insert size in mapped reads and the number of clusters per sample. See the attached plot.

          Based on this plot it seems like there might be a relationship, and a Pearson's correlation test seems to confirm it:
          > cor.test(abc$V1,abc$V2)

          Pearson's product-moment correlation

          data: abc$V1 and abc$V2
          t = -2.3818, df = 13, p-value = 0.0332
          alternative hypothesis: true correlation is not equal to 0
          95 percent confidence interval:
          -0.82929338 -0.05423031
          sample estimates:
          cor
          -0.5511812
          We want to process hundreds of samples very quickly, so unfortunately doing a bioanalyzer per-sample is not really a possibility for us. We're thinking instead that we might reduce our estimated input per-sample to 0.5ng in the hope that this will create smaller insert size distributions. We're a bit concerned about the effect this might have on library complexity, it seems like there must be a tradeoff between input quantity and complexity. One further question: when you use lower input amounts, do you add extra PCR cycles to compensate?

          Comment

          • ScottC
            Senior Member
            • Jan 2008
            • 244

            #6
            Are you processing the same sample type, or is each one different? We're a service provision lab so each sample we process can be from a different organism, prepared in a different way by a different set of hands with different reagents... we're running a Bioanalyzer trace for each of them for the time being because we see such huge differences between the libraries.

            If you're processing similar samples, I think you should be able to optimise things a bit and come up with a standard 'apparent' (using your method of quantitation) DNA concentration that will work well each time. When we first started using the XT preps and noticed such a wide variation in sizing and library yield, we did a set of 5 preps ranging from 0.2ng to 1.0ng to find an optimal mass... but, again, that only works if you're processing the same kinds of samples all the time (unfortuantely, we're not).

            We ended up settling on 0.8ng for most preps. We don't increase the PCR cycle number because the yield was good enough and we wanted to avoid too many PCR duplicates.

            Comment

            • koadman
              Member
              • May 2010
              • 65

              #7
              That's a good point, we've noticed similar issues in the past too and the samples in the histogram I posted above are indeed from at least 4 different sample types, at least 4 different DNA extraction methods, and 2 quantitation methods (picogreen on plates read by TECAN and qubit). The samples in the x-y scatterplot (my 3rd post) are a subset of those in the histogram which are all from the same sample type, DNA extraction method, and quantitation method. The hundreds of samples we're processing will all be extracted & quantitated the same way, all the same sample type, so yes, we'll aim for an operational standard optimized for that sample type. Still, it's annoying that extraction method, sample type, and quantitation method seem to have such a strong influence on the quality of tnp-catalyzed preps.

              Do you think adding extra PCR cycles would increase the PCR duplicate rate in resulting sequence data? I was under the impression that reducing starting material would have a stronger effect than adding a few extra cycles. PCR bias on GC content seems like it might be an issue with extra cycles.

              Comment

              • koadman
                Member
                • May 2010
                • 65

                #8
                Hi all, following up to give this thread some closure. We remade libraries for the 16 samples from the xy-plot above (all same sample type, DNA extraction, quantitation) using 0.5ng input and 14 cycles of enrichment PCR. The cluster count was low so it seems our loading concentration was too low. However, the normalization looks much better now:



                And Pearson's shows no correlation between insert size and read count. Thanks for the suggestions! Now just need to work out how to load these on the HiSeq!

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                  by SEQadmin2




                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                  10-30-2026, 06:46 AM
                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-28-2026, 09:57 AM
                0 responses
                45 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-28-2026, 09:51 AM
                0 responses
                26 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                75 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                49 views
                0 reactions
                Last Post SEQadmin2  
                Working...