    MiSeq v2 cluster density 1/2?

    Did a run last week on a v1 hardware MiSeq using a v2 chemistry 500 cycle kit on qPCR titrated libraries. Clustered at 10 pM. Instead of the expected 800-1000 clusters/mm^2 we got 425.

    Again these were with KAPA qPCR titrated libraries using the same protocol that usually allows us to nail cluster density. These TruSeq RNAseq libraries were a little odd in that the insert sizes were a mean of 700-800 bp to get the full value of the 250 base reads. But we have gotten dead-on accurate clustering results for libraries with much larger mean inserts sizes than these in the past. (Obviously this does require compensation for the larger amplicon sizes compared to the KAPA standards.)

    This is the first time we have used v2 chemistry, however. But Illumina Tech support says there should be no major difference in cluster densities with v2. I have another run to do with another set of similar libraries, but I don't want to waste the reagents getting 1/2 the amount of data I should. On the other hand clustering at 20 pM seems too aggressive.

    Anyone seeing much lower than expected cluster densities running v2 MiSeq chemistry? Especially with fairly long amplicons?

    --
    Phillip
  • Bucky
    Member
    • Feb 2010
    • 18

    #2
    We had the same problem here. We used to cluster at 8pM and that resulted in about 700k/mm2. Switching to v2, we ended up with 200k/mm2. Loaded the latest run at 15pm and that seemed to do the trick (1000k/mm2). Will probably set up the next run at 13 or 14pM.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      Thanks Bucky,
      I have been stripping my gears trying to figure out what was going on. That makes it seem like v2 performs more like a HiSeq with respect to raw cluster density[1].

      Are you running v1 hardware, or has your instrument been upgraded to v2?

      Anyone else see this or not see this?




      [1]Of course as a matter of efficiency the cBot blows the MiSeq away -- 120 ul/lane for the HiSeq gets you the same density that 600 ul gets you for the MiSeq.

      --
      Phillip

      Comment

      • Bucky
        Member
        • Feb 2010
        • 18

        #4
        Hey Philip,

        We are using v1 hardware. I should also mention that I pooled my sample at 4pm and then diluted it out more than usual in order to get a lower NaOH final concentration. Maybe that helped as well..

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Philip,

          Disclaimer: I do not work in the lab. I am merely passing the following info along.

          We are not seeing this problem with v.2. kits but the inserts in question are not long.

          Comment

          • pmiguel
            Senior Member
            • Aug 2008
            • 2328

            #6
            Originally posted by Bucky View Post
            Hey Philip,

            We are using v1 hardware. I should also mention that I pooled my sample at 4pm and then diluted it out more than usual in order to get a lower NaOH final concentration. Maybe that helped as well..
            Not sure I follow you. Normal protocol is to mix 10 ul of 0.2 M NaOH with 10 ul of 2 nM (total concentration of all libraries), incubate, then neutralize with 980 ul (50xdilution) of HT1. That gives you 1 ml of 20 pM denatured libraries, that you can dilute further, as needed.

            Could you elaborate, please?

            --
            Phillip

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by GenoMax View Post
              Philip,

              Disclaimer: I do not work in the lab. I am merely passing the following info along.

              We are not seeing this problem with v.2. kits but the inserts in question are not long.
              Is this v2 kits on v2 hardware, or v1 hardware?

              --
              Phillip

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                Originally posted by pmiguel View Post
                Is this v2 kits on v2 hardware, or v1 hardware?

                --
                Phillip
                This is v.2. kits on v.2 hardware.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by GenoMax View Post
                  This is v.2. kits on v.2 hardware.
                  Okay, this may be a v1 hardware/v2 reagent thing then. I heard from another core that they saw lower cluster densities ("like 1/2") using v2 chemistry on a v1 instrument. But that apparently went away when they were upgraded to v2 hardware.

                  --
                  Phillip

                  Comment

                  • Bucky
                    Member
                    • Feb 2010
                    • 18

                    #10
                    Originally posted by pmiguel View Post
                    Not sure I follow you. Normal protocol is to mix 10 ul of 0.2 M NaOH with 10 ul of 2 nM (total concentration of all libraries), incubate, then neutralize with 980 ul (50xdilution) of HT1. That gives you 1 ml of 20 pM denatured libraries, that you can dilute further, as needed.

                    Could you elaborate, please?

                    --
                    Phillip
                    Right. Instead of starting with 10ul of 2nM, I start with 10ul of 4nM. So in order to get to my desired final concentration, I need to dilute out more, thus lower NaOH concentration. Does that make sense?

                    Comment

                    • pmiguel
                      Senior Member
                      • Aug 2008
                      • 2328

                      #11
                      Originally posted by Bucky View Post
                      Right. Instead of starting with 10ul of 2nM, I start with 10ul of 4nM. So in order to get to my desired final concentration, I need to dilute out more, thus lower NaOH concentration. Does that make sense?
                      Ah, okay, got it.

                      --
                      Phillip

                      Comment

                      • GW_OK
                        Senior Member
                        • Sep 2009
                        • 411

                        #12
                        Illumina put out a new methodology for denaturation with the v2 chemistry. It's in the new Rev E version of the Miseq users guide.

                        Comment

                        • pmiguel
                          Senior Member
                          • Aug 2008
                          • 2328

                          #13
                          Originally posted by GW_OK View Post
                          Illumina put out a new methodology for denaturation with the v2 chemistry. It's in the new Rev E version of the Miseq users guide.
                          I downloaded it to take a look. Not seeing the difference in the denaturation method though. Looks pretty much the same to me.

                          It is different from the HiSeq -- there the concentration of NaOH used is 1/2 that than for the MiSeq denaturation. Is that new with Rev E of the MiSeq UG?

                          Anyway, we were using the 0.2 N NaOH, just as called for in Rev E.

                          --
                          Phillip

                          Comment

                          • pmiguel
                            Senior Member
                            • Aug 2008
                            • 2328

                            #14
                            After complaining the Illumina about this issue they told me they did not equipment in-house to test it out. (Meaning, I presume, all their in-house MiSeq have v2 hardware.) So they offered to send a couple of MiSeq kits out to test this hypothesis. Just started the v1 kit this morning.

                            Clustered at 10 pM and got 758 Kclusters/mm^2. That seems a little lower than normal, I think. Actually, I ran the same library on v2 chemistry 2 days ago at 15 pM and got a cluster density less than 500 Kclusters/mm^2. But Illumina wants the experiment to be done by the same person, same day, etc.

                            I'll let it run until cycle 26 and then abort the run, so I can get the second run started today.

                            --
                            Phillip

                            Comment

                            • Bucky
                              Member
                              • Feb 2010
                              • 18

                              #15
                              Cool, let us know what you find out!

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              38 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              23 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              72 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...