Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • andibody
    Member
    • Mar 2008
    • 45

    #1

    Troubles with Cluster Density

    Hi all,

    We have had sporadic troubles with cluster density. Our samples usually yield between 150.000 and 210.000 colonies per tile (GA IIx).
    Last weeks flowcell showed only ~90.000 colonies. Samples originated from different preparations making it unlikely them being the cause.
    Our proposed suspects: cluster gen kit / flowcell / sample denaturation.
    Has anybody had similar problems or knows which the critical steps in cluster generation are?
    We check concentration with bioanalyzer after sample prep; didn't include PhiX.


    Thanks
  • mollusc
    Junior Member
    • Oct 2009
    • 9

    #2
    we always aim for 210K clusters which is pretty easily acheived by quantifying the library on QPCR.The numbers are reproducible.variations in the cluster numbers might be because of pipetting errors.We had this problem intially, but overcame this prob by weighing the hyb buffer when making template hybs.May i ask you how much would you load onto flowcell to hit this number?
    Thanks!

    Comment

    • andibody
      Member
      • Mar 2008
      • 45

      #3
      Sorry for the late reply. We use a 6.5 pM but know that we could go a little bit higher. Which protocol do you use for QPCR quantification? At which step do you wheig hyb buffer? Before mixing with NaOH and subsequently with sample?
      The strange thing is we are used to a certain variance but from time to time it drops significantly without obvious reason...

      Comment

      • elaney_k
        Member
        • Mar 2008
        • 55

        #4
        Hi Andibody,

        What type of preps are you doing? I've seen it with Chip-seq and mRNA-seq preps were further analysis of the sequenced reads showed a percentage of adapter-dimers in the samples. This happened when the input was lower than expected for the protocol. QPCR showed up to 3 ct difference between samples even though they were all quantified and diluted to the same concentration (10nM).
        So whilst pipetting errors are definitely something to consider, any large variation in cluster number that I've come across has been due more to the sample prep where low input amounts (or low yield of polyA RNA during RNA-seq prep) leads to adapter-dimers in your final library. I would check the sequences that came off the run for presence of adapters.

        elaine

        Comment

        • andibody
          Member
          • Mar 2008
          • 45

          #5
          Hi Elaine,

          Most of our samples are indeed ChIP-Seq or RNA-Seq and in many cases starting amount is less than 10ng.
          We are at the moment setting up the QPCR hoping to get more reliable concentration estimations.
          We have checked in the past for adaptor contamination but haven't got many hits.
          How do you use Adaptor sequences as template for alignment? As single sequence or concatenated or randomly extended to the read length?

          Best
          Andi

          Comment

          • elaney_k
            Member
            • Mar 2008
            • 55

            #6
            Hi Andi,

            I look for a slightly truncated adapter sequence in the export files using the grep command so for ChIP-seq (single read adapters) the command I use is:
            grep -c "GATCGGAAGAGCTCGTATGCCGTCTTCT" s_N_export.txt
            and for RNA-seq (paired-read adapters) the command is:
            grep -c "GATCGGAAGAGCGGTTCAGCAGGAAT" s_N_export.txt

            The majority of our RNA-seq libraries have a 0 count for the adapter sequence but the odd few, where the final library yield was <5ng/ul resulted in much lower than expected cluster numbers on the flowcell and correlated with detection of adapter sequences in the sequenced reads.

            Elaine

            Comment

            • andibody
              Member
              • Mar 2008
              • 45

              #7
              I used Bowtie for alignment to adaptor sequence and even allowed for 2 mismatches but haven't got hits so I suppose that this is not the cause for my reduced cluster density. I hope that the QPCR will give more insight into this...


              Best
              Andi

              Comment

              • elaney_k
                Member
                • Mar 2008
                • 55

                #8
                Hi Andi,

                Does Bowtie allow you to align reads to a reference that's shorter than your read length?

                Comment

                • andibody
                  Member
                  • Mar 2008
                  • 45

                  #9
                  Hi Elaine,

                  No it doesn't, thanks for that hint.
                  I'll try out your grep command.
                  Why didn't you use the last 4 bases of the adaptor sequence?

                  Comment

                  • elaney_k
                    Member
                    • Mar 2008
                    • 55

                    #10
                    Hi Andi,

                    I was initially looking through data from a ChIP-seq run that had a really low % alignment and by eye I could see the same sequence repeated, when i checked the sequence it turned out to be a slightly shorter version of the adapter sequence. Less than 1% of the reads contained the full adapter sequence but up to 20% contained the truncated form.
                    I tried both full length adapter sequence and a slightly shorter version with the RNA-seq data when the cluster number dropped dramatically and found a much higher percentage of the shorter adapter sequence in the reads than the full-length again. The sequence wasn't just at the end of the reads either. I'm sure there's a logical molecular biology explanation for the phenomenon but I don't know exactly what it is myself.

                    Elaine

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                      by SEQadmin2




                      NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                      10-30-2026, 06:46 AM
                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-28-2026, 09:57 AM
                    0 responses
                    45 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-28-2026, 09:51 AM
                    0 responses
                    26 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    74 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    48 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...