Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • memento
    Member
    • Nov 2010
    • 49

    #1

    Stopped MiSeq run

    Hi,
    I experienced first problems with our MiSeq. This morning with new run loaded there was a flow rate error - first thought, valve.....
    Did re-test at 250ul (failed) and 500ul (machine freezed) so had to restart. While restart I re-seated the flow cell and I managed to pass the flow rate check (no other issues). After 1h run stopped - error msg that was thrown is related to "no clusters detected". We managed to do couple runs on it with no problems (300 cycle / 500 cycle / nano and regulars). This is updated v2 version (with recently updated software).
    Is this a prelude to valve problems (already reported in our collaborators lab) ?
    Thanks!
  • kwaraska
    Senior Member
    • Nov 2008
    • 131

    #2
    A no cluster error is thrown when there is nothing to see OR a sample with no diversity. If it sees all the same base at each position for the first 5 bases, it assumes everything is junk and will stop the run.

    I know this because it happened to me. When we spoke to the customer (we are a core) they told us there was no diversity on the first 10 or so bases. When we reran with a PhiX spike in-no problem.

    Comment

    • memento
      Member
      • Nov 2010
      • 49

      #3
      There was big diversity of the amplicons sequenced (more than 300 totally different ones) plus some phix... Thanks for a quick reply!

      Comment

      • kcchan
        Senior Member
        • Jul 2012
        • 186

        #4
        Were you using the correct sequencing primers for your amplicons? That would also give you a no clusters error message.

        MiSeq v2 has a newly designed valve, so it shouldn't have the same problems the earlier instruments experienced.

        Comment

        • memento
          Member
          • Nov 2010
          • 49

          #5
          kcchan,

          I checked the setting under sample sheet. last 3 runs (successful!) I had a different "adapter" sequence in my sample sheet csv - would that cause that??

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            I suppose you have already contacted Illumina tech support/local FAS. If there is something wrong with your MiSeq hardware then none of us are going to be of much help.

            Comment

            • memento
              Member
              • Nov 2010
              • 49

              #7
              Techsup asked me to run few tests. Volume test failed on PR2 sipper....I guess no wonder there was nothing detected if the incorporation buffer didnt find its way to a flow cell...
              looks like hardware.
              are they any fast with repairs ?
              thanks!

              Comment

              • ECO
                --Site Admin--
                • Oct 2007
                • 1360

                #8
                PR2 sipper fails often and is not a cause for concern.

                Did it make any images? I'm sure tech support will ask for your MCS and cycle logs...but you can poke around in there to see what the error might be (searching for "ERR:" is most helpful.

                Comment

                • memento
                  Member
                  • Nov 2010
                  • 49

                  #9
                  MCS was not able to eastablish focus model as the result of low intensity...
                  they also said that PR2 is a common issue and not a real problem...

                  Comment

                  • memento
                    Member
                    • Nov 2010
                    • 49

                    #10
                    I used these primers to enrich my library:
                    Forward:
                    aatgatacggcgaccaccgagatctacactctttccctacacgacgctcttc
                    reverse:
                    caagcagaagacggcatacgagatcggtctcggcattcctgctgaacc

                    Comment

                    • memento
                      Member
                      • Nov 2010
                      • 49

                      #11
                      what is more interesting. If I used i7 barcode (index1):
                      Reverse primer used to enrich:
                      caagcagaagacggcatacgagatCCTGCGAcggtctcggcattcctgctgaacc

                      Do I have to use custom index1 primer to read it or the MiSeq flow cell primer will work with this ?
                      THanks!

                      Comment

                      • MLog
                        Member
                        • Jan 2010
                        • 36

                        #12
                        Hi memento,

                        have you found what was the reason of failed run? Was this an instrument issue or rather reagents/library?
                        I'm asking because I got the same error from our Miseq last week. If this is a reagent issue, I would like to retry with a new kit. The samples need to be sequenced quite urgently.

                        Comment

                        • memento
                          Member
                          • Nov 2010
                          • 49

                          #13
                          @ MLog
                          When this problem arises, usually the flow cell is not seated properly. You have to open the flow cell compartment and re-seat the flow cell. If there was a spill all over the flow cell, You should clean the flow cell first. At that particular problem this helped.
                          Just now, after illumina replaced our v1 valve to v3 valve I experienced the same problem. This time though, re-seating the flow cell didnt help...illumina tech is coming today to help...
                          You can keep the reagents @ 2-8C for ~ 1 week. Flow cell should be intact (unless it is a flowcell's gasket that caused flow-rate error).

                          Comment

                          • MLog
                            Member
                            • Jan 2010
                            • 36

                            #14
                            I checked the flowcell, it looked perfectly OK and was delivering proper volumes of water during washes. So we retried with another kit and it worked without any problems. Looks like this no clusters error might be a reagent issue, too.

                            Comment

                            • d00b
                              Member
                              • Jan 2009
                              • 17

                              #15
                              Your flowcell problem

                              This is flowcell issue. the inlet of the gasket was not well aligned to the injector port. it will make this kind of error. you need to request replacement for the reaction kit and other supply material for index library prep.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              43 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              26 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              74 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...