    Tom - the MiSeq moves your sample from the cartridge to teh flowcell through non-disposable tubing, there is a risk some carry-over will happen but this can be mitigated against with several relatively small changes.

    Many applications will be largely unaffected, amplicons in heterogenous samples are likely to suffer the most.
    Send your samples with different barcodes each time if you can.
    Make sure to ask your facility to perform the washes between runs.

    There is a hack to fool the MiSeq into thinking it is clean, don't use it! A maintensance wash is a bit of a pain but we'll be running them until this is fixed.

    Comment

    • thomasblomquist
      Member
      • Jul 2012
      • 68

      #32
      Originally posted by james hadfield View Post
      Tom - the MiSeq moves your sample from the cartridge to teh flowcell through non-disposable tubing, there is a risk some carry-over will happen but this can be mitigated against with several relatively small changes.

      Many applications will be largely unaffected, amplicons in heterogenous samples are likely to suffer the most.
      Send your samples with different barcodes each time if you can.
      Make sure to ask your facility to perform the washes between runs.

      There is a hack to fool the MiSeq into thinking it is clean, don't use it! A maintensance wash is a bit of a pain but we'll be running them until this is fixed.
      Okay, that makes sense as to the source of the inter-run cross-contamination. Does the hi-seq operate in the same fashion?

      Still, that seems like a bit of a design flaw. Why would an engineer design a non-disposable piece of tube for the loading on the chip? For that reason alone I would be hesitent to use this in the clinic.

      -Tom

      Comment

      • mcnelson.phd
        Senior Member
        • Jul 2011
        • 162

        #33
        Man, I need to check these forums more regularly. We've been seeing this problem in all of our amplicon runs. At first we assumed it was contamination during the sample prep, but after running two completely distinct amplicon types (16S and non-ribosomal gene amplicons made using very different fusion primers) we still found low level mixing of the two despite the samples never coming anywhere near each other except that they were both run on the same MiSeq.

        I know my FAS said that the MiSeq tubing can't stand repeated exposures to KOH, so I guess that's out of the question as a cleaning solution, but hopefully they'll come up with something.

        Comment

        • thomasblomquist
          Member
          • Jul 2012
          • 68

          #34
          Just how non-disposable is this tubing? I'd be willing to pay an extra $100 a run just to have contamination free assurance.

          Can this topic be forwarded to someone at Illumina that is responsive to this need?

          -Tom

          Comment

          • mcnelson.phd
            Senior Member
            • Jul 2011
            • 162

            #35
            Originally posted by thomasblomquist View Post
            Just how non-disposable is this tubing? I'd be willing to pay an extra $100 a run just to have contamination free assurance.

            Can this topic be forwarded to someone at Illumina that is responsive to this need?

            -Tom
            I'd say that the tubing is very non-disposable because from what my FAS and one of Illumina's field technicians have told me, in order to replumb the MiSeq you're looking at taking the whole system apart, including the optics, and would then have to put it all back together and re-align the optical table. My FAS said he once nicked one of the lines while doing an optical adjustment, and it took him an entire day to get just the one line replaced (I don't know which it was) and put the optics back in alignment.

            Basically, Illumina will need to put out a new recommendation for a solvent that won't destroy the lines but will also sufficiently clean them to fix this issue.

            Comment

            • thomasblomquist
              Member
              • Jul 2012
              • 68

              #36
              And the hiseq?

              -Tom

              Comment

              • james hadfield
                Moderator
                Cambridge, UK
                Community Forum
                • Feb 2008
                • 224

                #37
                The tubing inside the MiSeq goes from the cartridge via a fixed sipper, through tubing to a vici valve, and through a final tube to the flowcell. It does not appear to be easily replaceable.

                However a better wash is likely to decrease the carry-over.
                Washing more frequently may well help.
                Changing barcodes between runs should help.

                These are all things we are in control of.

                I think HiSeq 2500 comes with the option of cBot clustering if this is the case then there is almost no issue for 2500.

                Comment

                • kameran
                  Junior Member
                  • Mar 2012
                  • 3

                  #38
                  Hi everyone, this is Kameran and I’m part of the Illumina Technical Support team. Illumina has posted a bulletin, “Best Practices for High Sensitivity Applications: Minimizing Sample Carryover”, which can be found by following this link:
                  https://icom.illumina.com/MyIllumina...applications-m


                  In summary, run to run sample carryover is more likely to have an effect on very low detection threshold applications. Our internal testing found that, when the instrument is properly washed and maintained, sample carryover typically remained below 0.1%, representing 1 read in 1000. Carryover may arise from a number of steps in the sequencing workflow, including incomplete removal of template DNA before the next run and various steps of library preparation. Proper maintenance procedures with water and Tween are recommended to minimize carryover. We currently do not advise the use of other wash additives, such as bleach, Triton or other decontamination solutions.

                  Of course, we’re always available by phone or email ([email protected]) to discuss any additional questions or concerns.
                  Kameran with Illumina Technical Support

                  Comment

                  • thomasblomquist
                    Member
                    • Jul 2012
                    • 68

                    #39
                    Originally posted by kameran View Post
                    Hi everyone, this is Kameran and I’m part of the Illumina Technical Support team. Illumina has posted a bulletin, “Best Practices for High Sensitivity Applications: Minimizing Sample Carryover”, which can be found by following this link:
                    https://icom.illumina.com/MyIllumina...applications-m


                    In summary, run to run sample carryover is more likely to have an effect on very low detection threshold applications. Our internal testing found that, when the instrument is properly washed and maintained, sample carryover typically remained below 0.1%, representing 1 read in 1000. Carryover may arise from a number of steps in the sequencing workflow, including incomplete removal of template DNA before the next run and various steps of library preparation. Proper maintenance procedures with water and Tween are recommended to minimize carryover. We currently do not advise the use of other wash additives, such as bleach, Triton or other decontamination solutions.

                    Of course, we’re always available by phone or email ([email protected]) to discuss any additional questions or concerns.
                    Thank you Kameran. What is the rate of decrease over multiple uses and washes? Specifically, using a shared instrument, how many other samples should I interdigitate between returning to previously used barcode combinations to achieve a 1:10,000; 1: 100,000, etc. carry-over?

                    -Tom

                    Comment

                    • kameran
                      Junior Member
                      • Mar 2012
                      • 3

                      #40
                      Hi Tom,



                      Studies regarding barcode integrations and degradation rate for run to run carryover have not been performed. However, customers have reported steady decreases from wash to wash. Anecdotal evidence suggests that, if carryover was 1 in 1000, it would be more like 1 in 1000000 the next time. This assumes there isn’t template that has dried and gotten “stuck” somewhere in the system. The more rinsing we can do on the MiSeq, the lower the carryover rate is expected to be.
                      Kameran with Illumina Technical Support

                      Comment

                      • Harlon
                        Junior Member
                        • Apr 2013
                        • 6

                        #41
                        Originally posted by kameran View Post
                        Hi Tom,

                        Studies regarding barcode integrations and degradation rate for run to run carryover have not been performed. However, customers have reported steady decreases from wash to wash. Anecdotal evidence suggests that, if carryover was 1 in 1000, it would be more like 1 in 1000000 the next time. Thlis assumes there isn’t template that has dried and gotten “stuck” somewhere in the system. The more rinsing we can do on the MiSeq, the lower the carryover rate is expected to be.
                        Hi Kameron

                        Thanks for the link to the technical bulletin. We had already implemented all of the fixes you suggest, including a maintenance wash between every run, prior to reporting our carry-over problem.

                        We have noticed that carry-over disappears within 2-3 runs at the read depths we are doing. Suggesting that additional washes would help. That said, I am really hoping Illumina will come up with a better solution than 'run 2 maintenance washes between runs'. Aside from the inconvenience, it increases the total time for a 150bp paired end run + wash to >24 hours. Since we're running our MiSeq pretty much every day, this would be a big hit to productivity for us.

                        Cheers
                        h

                        Comment

                        • thomasblomquist
                          Member
                          • Jul 2012
                          • 68

                          #42
                          This message string has been very helpful for our specific purposes. So, I'm thinking that if we have three batches of barcoding reagents that we rotate through that should be sufficient amount of inter-run washes to eliminate the tubing contamination of barcoded product from a previous runs that used the same barcode. I.e. every third to fourth run may have the same barcode.

                          Thanks again. Very helpful.

                          -Tom

                          Comment

                          • nhunkapiller
                            Junior Member
                            • Dec 2012
                            • 6

                            #43
                            Is there a consensus as to whether this issues affects the HiSeq 2500 in Rapid Mode?

                            Comment

                            • james hadfield
                              Moderator
                              Cambridge, UK
                              Community Forum
                              • Feb 2008
                              • 224

                              #44
                              The answer is in Kameran's post, although why Illumina only put this on iCom and not out there publicly is a little galling.

                              Originally posted by kameran View Post
                              Hi everyone, this is Kameran and I’m part of the Illumina Technical Support team. Illumina has posted a bulletin, “Best Practices for High Sensitivity Applications: Minimizing Sample Carryover”, which can be found by following this link:
                              https://icom.illumina.com/MyIllumina...applications-m


                              In summary, run to run sample carryover is more likely to have an effect on very low detection threshold applications. Our internal testing found that, when the instrument is properly washed and maintained, sample carryover typically remained below 0.1%, representing 1 read in 1000. Carryover may arise from a number of steps in the sequencing workflow, including incomplete removal of template DNA before the next run and various steps of library preparation. Proper maintenance procedures with water and Tween are recommended to minimize carryover. We currently do not advise the use of other wash additives, such as bleach, Triton or other decontamination solutions.

                              Of course, we’re always available by phone or email ([email protected]) to discuss any additional questions or concerns.
                              They say "To avoid direct introduction of samples on the HiSeq 1500/2500, the TruSeq® Rapid Duo cBot Sample Loading Kit (CT-402-4001) can be used to load samples onto the flow cell. The Rapid Duo kit contains disposable parts, thereby limiting the opportunity for carryover during clustering."

                              I think it is clear there is a fail-safe work-around for HiSeq 2500. Now we just need something a little better for MiSeq and we're all happy once more!

                              Comment

                              • id0
                                Senior Member
                                • Sep 2012
                                • 130

                                #45
                                I am just trying to understand this issue more. If the contamination is less than 1%, it seems largely irrelevant. Based on my experience, trying to call variants of that frequency will yield a huge amount of false positives. Isn't the sequencing error rate a much bigger problem that makes this a moot point?

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                  by SEQadmin2




                                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                  Today, 06:46 AM
                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-28-2026, 09:57 AM
                                0 responses
                                22 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-28-2026, 09:51 AM
                                0 responses
                                15 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 11:13 AM
                                0 responses
                                68 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-22-2026, 10:48 AM
                                0 responses
                                48 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...