Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • MikeNice
    Junior Member
    • Mar 2010
    • 6

    #1

    Is it possible to re-sequence Index2 on a HiSeq PE dual index run?

    Background:
    Instead of combining 4 50 cycle kits for our HiSeq dual index runs, we typically use standard 200 cycle kits and 1 50 cycle kit to top off up to 4 dual index runs starting on the same day. Further, we split one incorporation mix between FCA and FCB at the start of all our HiSeq runs, and split the second mix at the turn.

    Our problem:
    On this particular dual index run, the reagents were not topped off at the start of the run, thus the incorporation mix ran dry sometime during the dark cycles, causing index2 to fail. The dilemma here is that simply adding incorporation mix and resuming the run at the beginning of index2 will cause the read to be out of phase since an unknown number of bases were added during the dark cycles before the mix ran dry. Since the index2 primer is grafted to the flowcell we cannot denature the strand and start the read over either.

    Any suggestions as to how to save this run? Thanks in advance!
  • kcchan
    Senior Member
    • Jul 2012
    • 186

    #2
    Calling up tech support might be your best bet at this point. I would guess that a strip and rehyb would still work at this point. The only difference here being the fact that you won't be adding a new primer but instead relying on the grafted flow cell primers.

    Comment

    • HeinKey
      Member
      • May 2009
      • 21

      #3
      MikeNice,

      If you flush your system with high salt to remove the air, and then continue, and repeat the second index?
      You first need to tweak the HCS so it will image the dark cycli resulting in 15 cycli imaging instead of 7 dark+8 imaging. How to do this I don't know, this might be impossible...??
      If you just repeat it you will miss the first x-cycli which were done in your first dark cycli attempt.

      You will have to demultiplex your self since you don't know which cycli will be index cycles. After the P5index I expect the turnaround will complete normally so the reverse read can be sequenced normally.

      @kcchan, I am not sure of strip and rehyb are necessary. I think the grafted oligo's are fixed and extended with 1-7 (dark cycle) nucleotides. I expect these extended oligo's to rehyb to the first strands ending up in the same situation as before strip and rehyb.
      Last edited by HeinKey; 05-29-2013, 02:39 AM. Reason: did not read the background, only the question

      Comment

      • MikeNice
        Junior Member
        • Mar 2010
        • 6

        #4
        @HeinKey and @kcchan, thank you both so much for your replies.

        As HeinKey pointed out, I don't think the strip and rehyb work in this scenario because the index2 primer and therefore unknown number of extended bases are grafted to the flowcell (this is unlike all other reads, which are only hydrogen bonded to the grafted strand and can therefore be denatured and washed away).

        I very much like the idea to change the recipe so that the dark cycles are imaged ensuring we capture the entire index2 read. Tech Support is not very forthcoming on editing HiSeq recipes, but if this happens to us again I will try this solution and update this thread with the protocol and outcome.

        As it turns out, in this case we could demultiplex with only index1 and 2 bases from index2.

        Thanks again!

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          Yesterday, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        29 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        18 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        71 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Working...