If you're not quantitating via qPCR, you run the risk of this issue as discussed above:
-
Dear Jessica, thanks for your replay. I am not absolutely sure about the problem is the agilent kit but other libraries prepared with other kits and quantified and denatured in the same way are working properly. We have prepared 2 batches of libraries with 2 different lots and both failed, they were not used before. In fact, we have repeated 4 samples with a new kit sent by techsupport and we have no clusters... The quantification has been done with Qubit and bioanalyzer, in the same way as the 200 libraries that worked fine in previous runs. In fact, we have tested several concentrations and all of them had the same low number of clusters. Perhaps there is somthing in the pcr2 primer sequences, something that makes the DNA more sensible to denaturation... I don't know but today I have launched a new run and I have obtained low clusters again...
Leave a comment:
-
lpalacios-- if you're fairly certain it's the Sureselect kit, I'd consider the number of freeze/thaws the kit components have been through. If you're using a brand new kit, maybe check whether this is a new kit lot number compared to the other ones you've used. I would assume, though, that if the problem was with the kit, you'd see some problems on the bioanalyzer-- reduced yield, etc.
The other question to ask is how are you quantitating your libraries after construction? Are you using a qPCR method?
Leave a comment:
-
Dear all,
I am having problems with cluster density in the last 8 libraries prepared with Agilent Sureselect XT kit. I have run more than 40 libraries prepared with this kits with an average of 800 clusters when I loaded 8pM but in the last runs they have decreased to about 250-300 clusters! I have tried reducing NaOH in the denaturalization, the protocol starting with 2 or 4nM, to load 8, 12.5, 15 or 20pM of library... And nothing worked. I have run other libraries prepared with other kits and they work ok, so it is not an issue of the MiSeq. The final purification is done with agentcourt beads and the bioanalyzer traces are ok, similar to runs that worked fine.
Could anybody help me?
Thanks a lot
Leave a comment:
-
Originally posted by LabRat89 View PostAdditionally, has anyone used kits past expiration date and still have it work? If so, what is the time frame?
Leave a comment:
-
Originally posted by NextGenSeq View PostThe low cluster density is usually caused by your NaOH. Either your NaOH pH is off or you are adding too much Na+ ions. which interfere with library hybridization to the flowcell.
We always check the pH of our NaOH prior to using it by making up a 0.1 M solution of our stock. The pH of 0.1M NaOH should be at least 12.6 or higher.
CO2 from the atmosphere will acidify the NaOH solution over time.
--
Phillip
Leave a comment:
-
The low cluster density is usually caused by your NaOH. Either your NaOH pH is off or you are adding too much Na+ ions. which interfere with library hybridization to the flowcell.
We always check the pH of our NaOH prior to using it by making up a 0.1 M solution of our stock. The pH of 0.1M NaOH should be at least 12.6 or higher.
CO2 from the atmosphere will acidify the NaOH solution over time. Also, make sure you are using the more recent denaturation protocol (attached).
Denature DNA for 4 nM Library
1 Combine the following volumes of sample DNA and freshly diluted 0.2 N NaOH in
a microcentrifuge tube:
• 4 nM sample DNA (5 μl)
• 0.2 N NaOH (5 μl)
2 Discard the remaining dilution of 0.2 N NaOH or set aside to prepare a PhiX control
within the next 12 hours.
3 Vortex briefly to mix the sample solution, and then centrifuge the sample solution to
280 × g for 1 minute.
Denature and Dilute DNA
Preparing Libraries for Sequencing on the MiSeq 9
4 Incubate for 5 minutes at room temperature to denature the DNA into single strands.
5 Add the following volume of pre-chilled HT1 to the tube containing denatured DNA:
• Denatured DNA (10 μl)
• Pre-chilled HT1 (990 μl)
The result is a 20 pM denatured library in 1 mM NaOH.
6 Place the denatured DNA on ice until you are ready to proceed to final dilution.Last edited by NextGenSeq; 08-26-2014, 10:20 AM.
Leave a comment:
-
"Needs Attention" Message on BaseSpace
Thank you nucacidhunter for the attachment.
I encountered another problem that I've never seen before. Under BaseSpace, the run is showing "Needs Attention" even though the MiSeq instrument displays successful completion of the run.
Please see attached image. Have you or anyone else encountered this before?
Thank you again!Attached Files
Leave a comment:
-
Cluster densities given for Illumina systems are for sequencing libraries prepared with Illumina kits and does not apply to third party kits or methods. This attachment is a useful source for troubleshooting cluster densities issues:Attached Files
Leave a comment:
-
Originally posted by nucacidhunter View PostIllumina's stand on max of 1mM NaOH is pretty firm, so it must make a difference. It is easy to achieve this and follow Illumina's instruction. If library concentration is low it can be concentrated. You can also increase loading to 15-16 pM (maybe in small steps). Just have to make sure that low cluster number is not due to over-clustering where RTA struggles to recognise very closely located clusters and do not use them for template generation.
Additionally, has anyone used kits past expiration date and still have it work? If so, what is the time frame?
Thanks again!!
Leave a comment:
-
Originally posted by nucacidhunter View PostIllumina's stand on max of 1mM NaOH is pretty firm, so it must make a difference. It is easy to achieve this and follow Illumina's instruction. If library concentration is low it can be concentrated. You can also increase loading to 15-16 pM (maybe in small steps). Just have to make sure that low cluster number is not due to over-clustering where RTA struggles to recognise very closely located clusters and do not use them for template generation.
Leave a comment:
-
Illumina's stand on max of 1mM NaOH is pretty firm, so it must make a difference. It is easy to achieve this and follow Illumina's instruction. If library concentration is low it can be concentrated. You can also increase loading to 15-16 pM (maybe in small steps). Just have to make sure that low cluster number is not due to over-clustering where RTA struggles to recognise very closely located clusters and do not use them for template generation.
Leave a comment:
-
Hi everyone,
So I understand that the final concentration of NaOH should be a max at 1mM. However, we've been loading it at 1.2mM final concentration with 12pM sample input. Our cluster densities have been roughly 500k/mm2 for the past 3 runs. Would you say the 0.2mM difference is negligible?
Thanks in advance!
Leave a comment:
-
The solution to the problem comes definitely by controlling the NaOH concentration to a max of 1mM. I tried loading the libraries at 10pM concentration with 1mM NaOH final concentration and that did the trick.
Leave a comment:
-
Hi Memento,
I agree with your observations and have observed a similar trend.
I thus stick with the Qubit measurement, which I believe uses a Picogreen or similar kind of a dye, and is quite accurate.
How would one get to check if the valve is a problem? The machine is relatively new, and not too many runs have been conducted (probably 10 -11 runs).
Would I need to do a PhiX flow cell? Does Illumina help with providing a test kit?
ThanksLast edited by abyss; 07-17-2013, 01:50 PM.
Leave a comment:
Latest Articles
Collapse
-
by seqadmin
The field of immunogenetics explores how genetic variations influence immune responses and susceptibility to disease. In a recent SEQanswers webinar, Oscar Rodriguez, Ph.D., Postdoctoral Researcher at the University of Louisville, and Ruben Martínez Barricarte, Ph.D., Assistant Professor of Medicine at Vanderbilt University, shared recent advancements in immunogenetics. This article discusses their research on genetic variation in antibody loci, antibody production processes,...-
Channel: Articles
11-06-2024, 07:24 PM -
-
by seqadmin
Next-generation sequencing (NGS) and quantitative polymerase chain reaction (qPCR) are essential techniques for investigating the genome, transcriptome, and epigenome. In many cases, choosing the appropriate technique is straightforward, but in others, it can be more challenging to determine the most effective option. A simple distinction is that smaller, more focused projects are typically better suited for qPCR, while larger, more complex datasets benefit from NGS. However,...-
Channel: Articles
10-18-2024, 07:11 AM -
ad_right_rmr
Collapse
News
Collapse
Topics | Statistics | Last Post | ||
---|---|---|---|---|
Started by seqadmin, Today, 11:09 AM
|
0 responses
22 views
0 likes
|
Last Post
by seqadmin
Today, 11:09 AM
|
||
Started by seqadmin, Today, 06:13 AM
|
0 responses
20 views
0 likes
|
Last Post
by seqadmin
Today, 06:13 AM
|
||
Started by seqadmin, 11-01-2024, 06:09 AM
|
0 responses
30 views
0 likes
|
Last Post
by seqadmin
11-01-2024, 06:09 AM
|
||
New Model Aims to Explain Polygenic Diseases by Connecting Genomic Mutations and Regulatory Networks
by seqadmin
Started by seqadmin, 10-30-2024, 05:31 AM
|
0 responses
21 views
0 likes
|
Last Post
by seqadmin
10-30-2024, 05:31 AM
|
Leave a comment: