    using NEB kit to make libraries for Illumina MiSeq

    Hello,

    I used NEB multiplex kit to make libraries for Illumina MiSeq. Trying to figure out the amount of sample to load in the flow cell to get a decent cluster density. Last round I loaded 12pM of total sample and it over clustered with my coverage very low. My colleague uses 7pM of the same kind of sample on a HiSeq, how does that compare with MiSeq? should Miseq sample amount be lesser or higher?

    Thanks for your suggestions.
  • Isequencestuff
    Member
    • Nov 2012
    • 21

    #2
    what MiSeq kit (flowcell) are you using? I assume you're using the V2 kit since you're aiming to load 12 pM of total sample. If your quantitation method is accurate, aiming to load 12 pM should give you a perfect cluster density. We usually get around 850-900. Our quantification process: qPCR post library construction, then based on those quantities dilute all library samples to 2 nM and pool. Then run a last qPCR on that one pooled library sample to make sure it's close to 2 nM (this saves us from wasting a flow cell). From there we dilute to 12 pM final (including about 1-5% phiX), then load and go.

    Comment

    • MWN
      Junior Member
      • Aug 2011
      • 8

      #3
      For V2, I load 11 pM based on Qubit. For V3, 14 pM.

      Comment

      • kcchan
        Senior Member
        • Jul 2012
        • 186

        #4
        Comparing loading concentrations isn't very beneficial. Long inserts also don't behave in the same way as short inserts, so a 12pM TruSeq DNA library may be fine whereas a 12pM miRNA library will overcluster. Also, each instrument has its own different sweet spot. The same sample at 12pM may be perfect on one MiSeq but overcluster on a different MiSeq. It's better being more conservative (4-6pM) and working your way up over the course of a few runs.

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          10-30-2026, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        45 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        26 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        75 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        49 views
        0 reactions
        Last Post SEQadmin2  
        Working...