    Quality drop in the Reverse read of paired end library (illumina Hiseq 1000)

    Hi
    Recently we have prepared 3 paired end libraries (2X100) and sequenced on illumina hiseq 1000. we have checked the quality of the sequenced reads and found that for one library both forward and reverse reads showed good quality fatsQC reports. However for other 2 libraries the fastQC report for forward read (R1) is very good. But the fastQC report for reverse read is looking really bad (especially the per base quality score). I suppose if there is any problem with the library preparation then it should affect both the reads (R1, R2) which is not the case here.Could anyone let me know the reason why this is happening?


    Thanks
    Last edited by anurupa; 10-19-2013, 04:17 AM.
  • kohanlon
    Junior Member
    • Dec 2012
    • 3

    #2
    I saw a similar problem with some of my libraries and after sending data to Illumina, found that my libraries were too concentrated. When the cluster density gets too high, the second read quality can be affected significantly. The instrument has trouble doing the turnaround when there are so many clusters so your Qscores drop off. You should be able to look at your images and see if you have a higher density of clusters at the front of the flow cell or at the back. If the clusters have a higher density at the back of the flow cell, it is an indication of over clustering and could be the cause of your drop in Qscores. I would try using a lower concentration of your library for the input and see if that clears up the problem.

    Comment

    • anurupa
      Member
      • Jan 2012
      • 14

      #3
      we have loaded three libraries in a single lane (multiplexed). As mentioned out of 3 libraries 1 library din't show any problem for forward(R1) and reverse(R2). we have checked the cluster density which is ~600 k/mm2. I think it is optimal value and the problem arises only if it is above 900 k/mm2

      Comment

      Latest Articles

      Collapse

      • SEQadmin2
        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
        by SEQadmin2




        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
        Yesterday, 06:46 AM
      • SEQadmin2
        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
        by SEQadmin2



        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
        ...
        10-14-2026, 09:55 AM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by SEQadmin2, 10-28-2026, 09:57 AM
      0 responses
      29 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-28-2026, 09:51 AM
      0 responses
      20 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 11:13 AM
      0 responses
      71 views
      0 reactions
      Last Post SEQadmin2  
      Started by SEQadmin2, 10-22-2026, 10:48 AM
      0 responses
      48 views
      0 reactions
      Last Post SEQadmin2  
      Working...