  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #46
    Originally posted by TonyBrooks View Post
    Surely anything beginning with 5 G's won't be registered as a cluster. So bubbles wouldn't contain any clusters either - so you don't get any data. I would expect that there is some algorithm there to detect when something is a G and when it's an intermittent bubble. G's would be localised areas with no signal, bubbles would cover much larger areas.
    I would bet no such checking is done and bubble covering a cluster get called as "G's".

    Alternatively, it is technically possible to eliminate or drastically decrease the number of bubbles that traffic through the flowcell. Illumina has never seemed the least bit interested in doing this, that I could tell. But maybe they have sufficient impetus here to deal with this issue.

    --
    Phillip

    Comment

    • ymc
      Senior Member
      • Mar 2010
      • 496

      #47
      Originally posted by AllSeq View Post
      On our blog we've given our take on Illumina's "$1000" genome and why it still isn't available for most people.
      I think it makes sense commercially for them to market it like that in the beginning. Over time, I think lesser players should be able to buy one instead of ten. Application to other species and non-WGS uses should also be available given time.

      Comment

      • ymc
        Senior Member
        • Mar 2010
        • 496

        #48
        I think the three buyers of HiSeq X Ten so far probably has a lot of cancer genomes they want to sequence. Now they can go 300x easily. They may also want to run 1000x or more to study tumor heterogeneity.

        Comment

        • snetmcom
          Senior Member
          • Oct 2008
          • 159

          #49
          There's only so many institutions that can afford/support a Hiseq X. I imagine they were sold so quickly because they were designed specifically for those customers. They probably helped set the spec's of the system.

          Comment

          • ymc
            Senior Member
            • Mar 2010
            • 496

            #50
            What is the reagent cost for NextSeq 500 High output mode for 2x150 and 2x75? Anyone knows? Thanks

            Comment

            • SNPsaurus
              Registered Vendor
              • May 2013
              • 525

              #51
              Here are some prices I've heard:
              FC-404-1002 | NextSeq™ 500 High Output Kit (150 cycles)
              List Price (USD):$2,500.00

              FC-404-1004 | NextSeq™ 500 High Output Kit (300 cycles)
              List Price (USD):$4,000.00

              FC-404-1005 | NextSeq™ 500 High Output Kit (75 cycles)
              List Price (USD):$1,300.00
              Providing nextRAD genotyping and PacBio sequencing services. http://snpsaurus.com

              Comment

              • Genohub
                Registered Vendor
                • Mar 2013
                • 210

                #52
                Really nice piece on the computation costs of the HiSeq X Ten:

                Comment

                • ymc
                  Senior Member
                  • Mar 2010
                  • 496

                  #53
                  Originally posted by Genohub View Post
                  Really nice piece on the computation costs of the HiSeq X Ten:

                  http://glennklockwood.blogspot.co.uk...=1389835567115
                  But BGI's 128 HiSeq 2000s have roughly the same output as 10 HiSeq X.

                  So if they could handle it several years ago, I think the new HiSeq X owners should be ok. It is not like they ordered 100 each this time around.

                  Comment

                  • ymc
                    Senior Member
                    • Mar 2010
                    • 496

                    #54
                    Originally posted by SNPsaurus View Post
                    Here are some prices I've heard:
                    FC-404-1002 | NextSeq™ 500 High Output Kit (150 cycles)
                    List Price (USD):$2,500.00

                    FC-404-1004 | NextSeq™ 500 High Output Kit (300 cycles)
                    List Price (USD):$4,000.00

                    FC-404-1005 | NextSeq™ 500 High Output Kit (75 cycles)
                    List Price (USD):$1,300.00
                    For NextSeq High Output PE150, assuming two years shelf life, there can be 365*2*24/29 = 604 run per machine. The per run equipment cost is $250,000/604= $414. Each run costs $4,000. then for a total of 120Gbp output, the cost is $36.78/Gbp.

                    From a # of read pairs POV, for PE75, 365*2*24/18 = 973 runs per machine. $250,000/973 = $257 equipment cost. $2,500 to run two flow cells. Then for a total of 400M read pairs = $6.89/M read pairs

                    For HiSeq Rapid PE150, $750,000/(2*365*2*24/39) = $834.76 equipment cost. $7,800 to run two flow cells. Then for an output of 180Gbp, the cost is $47.97/Gbp.

                    For HiSeq Rapid PE100, $750,000/2*365*2*24/27) = $577.9 equipment cost. $5,800 to run two flow cells. For an output of 600M read pairs = $10.63/M read pairs.

                    For MiSeq V3 PE300, $100,000/(365*2*24/65) = $371 equipment cost. $1,400 per run. For an output of 15Gbp, the cost is $118/Gbp

                    For MiSeq V3 PE75, $100,000/(365*2*24/24) = $137 equipment cost. $800 per run. For an output of 25M read pairs, $37.48/M read pairs.

                    So the conclusion is NextSeq 500 makes sense as expected....

                    Comment

                    • TonyBrooks
                      Senior Member
                      • Jun 2009
                      • 303

                      #55
                      Originally posted by pmiguel View Post
                      I would bet no such checking is done and bubble covering a cluster get called as "G's".

                      Alternatively, it is technically possible to eliminate or drastically decrease the number of bubbles that traffic through the flowcell. Illumina has never seemed the least bit interested in doing this, that I could tell. But maybe they have sufficient impetus here to deal with this issue.

                      --
                      Phillip
                      We''ll have to wait and see. I have yet to see a bubble on the MiSeq, (although we don't look for one, we have yet to see any spike in the per base N plot of fastqc that we see with HiSeq BMS). So they may have learnt a couple of tricks (like embedding the gaskets on the flow cell).

                      Comment

                      • pmiguel
                        Senior Member
                        • Aug 2008
                        • 2328

                        #56
                        Originally posted by TonyBrooks View Post
                        We''ll have to wait and see. I have yet to see a bubble on the MiSeq, (although we don't look for one, we have yet to see any spike in the per base N plot of fastqc that we see with HiSeq BMS). So they may have learnt a couple of tricks (like embedding the gaskets on the flow cell).
                        Alright, different instrument. Probably won't have the issues associated with the HiSeq (bubbles.)

                        Let's go back to cluster recognition. Looks like 5 initial G's and the cluster will not be seen at all. But, as you mention, that would be less than 0.1% of the amplicons. So the bias would be unlikely to cause issues.

                        How about 4G's? Do you really think the software would be confident enough of a cluster which only was visible in a single cycle out of the first 5?

                        How about 3G's?

                        This whole inferring signal from a lack of signal makes me twitchy...

                        --
                        Phillip

                        Comment

                        • TonyBrooks
                          Senior Member
                          • Jun 2009
                          • 303

                          #57
                          Originally posted by pmiguel View Post
                          Alright, different instrument. Probably won't have the issues associated with the HiSeq (bubbles.)

                          Let's go back to cluster recognition. Looks like 5 initial G's and the cluster will not be seen at all. But, as you mention, that would be less than 0.1% of the amplicons. So the bias would be unlikely to cause issues.

                          How about 4G's? Do you really think the software would be confident enough of a cluster which only was visible in a single cycle out of the first 5?

                          How about 3G's?

                          This whole inferring signal from a lack of signal makes me twitchy...

                          --
                          Phillip
                          I guess that could be why cluster density is so low.

                          Comment

                          • GW_OK
                            Senior Member
                            • Sep 2009
                            • 411

                            #58
                            So I've been watching Jay Flatley's taped talk to investors from yesterday and I've taken some screenshots and notes for your enjoyment:


                            Better chemistry including new cluster method, better scan mix, speeding up the paired end turnaround chemistries, new custom camera that can do *bi-directional* scanning (allows for 6x speedup), higher powered lasers, new computer system (bye-bye Dell tower) and re-written software


                            Remember when Ion Torrent said they were going to use chip foundries to change the sequencing world? Wafer-scale flowcell generation for patterned FCs, 12 FC/wafer.


                            400nm feature size. Top and bottom of FC.


                            Example patterned FC picture.


                            700nm center-to-center distance for 400nm-sized features. Plenty of headspace to pack more and more in.


                            Seeding the current FCs is a problem b/c DNA binds randomly and you can get polyclonal clusters (Poisson distribution). Current size of clusters and camera power can discern polyclonal clusters fairly well, but with 400nm features there's no way to tell them apart.


                            New cluster chemistry has simultaneous seeding and amplification. Instantaneous amplification of the template as soon as it sits on a cluster site prevents two templates from seeding the same feature via competition. 80% monoclonal features vs 40% from regular FCs.


                            "Typical" data qualities from a HiseqX, ~120Gbp/lane. 1.8Tbp/run is conservative. 36x human genome/lane.

                            System is *licensed only for human WGS*, so that's how they keep you from running other stuff.

                            Comment

                            • GW_OK
                              Senior Member
                              • Sep 2009
                              • 411

                              #59
                              More from the same presentation, NextSeq500:



                              Larger than normal flowcell. Like, the size of your hand.


                              2-dye chemistry.


                              LED light sources and cellphone cameras. Low cost, light weight, no alignment needed on install.


                              Fluidics module. "Valve and stagger" technology allows for chemistry and imaging at the same time.


                              2 different flowcells.


                              Price per Gbp

                              Comment

                              • BBoy
                                Member
                                • Oct 2010
                                • 50

                                #60
                                Illumina controls the clustering adapters and sequencing primers. It should be trivial to ensure no Gs at the end of the cluster.

                                Comment

                                Latest Articles

                                Collapse

                                • SEQadmin2
                                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                  by SEQadmin2



                                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                  ...
                                  10-14-2026, 09:55 AM
                                • SEQadmin2
                                  New Genomics Technologies Take Aim at Long-Standing Limits
                                  by SEQadmin2


                                  Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                  We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                  ...
                                  09-28-2026, 10:25 AM

                                ad_right_rmr

                                Collapse

                                News

                                Collapse

                                Topics Statistics Last Post
                                Started by SEQadmin2, 10-15-2026, 10:52 AM
                                0 responses
                                21 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-14-2026, 05:39 AM
                                0 responses
                                26 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-13-2026, 09:56 AM
                                0 responses
                                22 views
                                0 reactions
                                Last Post SEQadmin2  
                                Started by SEQadmin2, 10-09-2026, 11:14 AM
                                0 responses
                                28 views
                                0 reactions
                                Last Post SEQadmin2  
                                Working...