    Water blanks

    Some colleagues of mine are running an Illumina TSCA panel on FFPE samples using the MiSeq, they have just starting using a water blank and putting it through the whole process including sequencing on the MiSeq. As a diagnostic lab using a water blank is common practise for every other technique we use.

    They were surprised at how many reads were assigned to the blank. I was just wondering if anyone else had tried running this sort of blank on a MiSeq or any other NGS machine for that matter. Or if anyone has any thoughts on using water blanks on NGS?

    Thanks in advance!!
  • dpryan
    Devon Ryan
    • Jul 2011
    • 3478

    #2
    What are they hoping to achieve with a water blank? The amount of noise in such an experiment would likely be over-inflated.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      Is the water blank part of the pool that is being run? If so you would be interested in looking at JackieBadger's post (#3) in this thread: http://seqanswers.com/forums/showthread.php?t=38697

      Comment

      • AStretton
        Member
        • Jul 2012
        • 25

        #4
        Hi dpryan - with the TSCA you run a gel after the PCR step, some of the samples on the gel, look as if there is no band but they are still run on the miseq and they still some times get good results from these and these results are then reported back. Sometimes they get quite low coverage less than 150x but are reporting if a mutation is found (we are looking at somatic mutations in FFPE DNA). So when they ran a water blank which is also showing as negative on a gel, to check that the mutations they are reporting are genuine and not an artefact of the assay and this water blank then had quite good coverage up in 100x coverage they are now worried that they are reporting false positives. We have a few things we are going to look at to try to address this but I was just wondering what other people thought about running water blanks.

        Hi GenoMax - Yes it is part of the pooled samples, thanks I'll have a read of that paper
        Last edited by AStretton; 01-14-2014, 05:48 AM. Reason: Missed a bit

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          There are approved assays for CF diagnostics for MiSeqDX so you should ask your local FAS about what kind of controls that assay is using.

          This kit for MiSeq is referring to a master negative control (which may be the water blank you are referring to).

          Comment

          • AStretton
            Member
            • Jul 2012
            • 25

            #6
            The TSCA kit doesn't come with a negative control, so we run a negative of our own, running through the entire prep but we haven't actually put any sample in. I will have a chat with our FAS about what the negative control is in that kit though, thanks for the suggestion.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              Today, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            23 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            15 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            68 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...