  • GenoMax
    replied
    Originally posted by WhiteSeal View Post
    @GenoMax: I have not had the opportunity to look into the data myself, but customer is complaining that Qscores at final 50 cycles is well below expected.
    Should one overly worry about Q-scores, specially for a re-sequencing run (not sure what kind of a run this is)? If a reference genome was available (and you are not trying to call SNP's in that region) the sequence may be still usable (irrespective of the Q-scores). If this was a de novo run then the worry would be justifiable.

    Leave a comment:


  • NextGenSeq
    replied
    Originally posted by WhiteSeal View Post
    @GenoMax: samples were checked on BioAnalyzer to check the length of the fragments (all >700bp) so this should not happen. Also, we have seen this with almost all our 600PE v3 runs on the MiSeq.

    @NextGenSeq: MiSeq run 600PEv3 with 566k clusters gave - according to illumina output - ~14M reads, but in read two at cycle 550 already 5M reads have a Qscore <Q30 (not good), and at cycle 600 12M reads have a Qscore <Q30 (really not good). And then this was a 'good' run.
    This is really low. Did you spike in PhiX? If so what are those reads like?

    If they are good it's your library.

    Leave a comment:


  • WhiteSeal
    replied
    @nucacidhunter: indexing was fine. 76 samples, barcode variation for each base was evenly distributed. So no issue there.

    @GenoMax: I have not had the opportunity to look into the data myself, but customer is complaining that Qscores at final 50 cycles is well below expected.

    @avo: I was wondering.. you say you have seen worse and you just trim the data, but with so many reads not making it till the end then what is the use of running a 600PE kit. I (and customer also) expect at least 70% of all data to have a >Q30. In this case 70% is around cycle 550...

    As for faulty reagents. This was actually a replacement kit from Illumina for one of their faulty ones. Also, I was proactive and complained and replacement has been received to run the pool again.
    Last edited by WhiteSeal; 09-26-2014, 06:48 AM.

    Leave a comment:


  • avo
    replied
    I'm a bit confused by the %Q30 graph shown here. I have seen 600V3 runs on our MiSeq that looked much worse. Yet we generally cluster well above 566k, so that might be a factor. Nevertheless I would be perfectly satisfied to see a Q30 graph like this. Even the example runs I looked at on Illuminas BaseSpace for the 600V3 kit are worse (SmallGenomesBcereusRhodo).

    Because all our runs met the Illumina specs of a Total % ≥ Q30 of 70%, i never thought about complaining.
    We observe this for a variety of different samples using TruSeq LT as well as TruSeq PCR-free libraries. Based on the %Base plots and the output of the adapter trimming there shouldn't be a significant problem with the fragment size (>800 bp) either.

    Despite that we lose bases from the ends of both reads during the quality trimming it doesn't have a major impact on our analysis.

    Does anyone experience similar issues?

    Leave a comment:


  • GenoMax
    replied
    @WhiteSeal: Has this data been analyzed and you are sure the sequence (irrespective of Q-scores) looks ok (no adapters etc).

    Points mentioned by nucacidhunder are very relevant. If the nucleotide diversity drops late into the read then that would also lead to Q-score drops. Since you have been in touch with Illumina tech support if you used reagents that were recalled they should alert you. Wouldn't hurt to proactively check on that possibility too.

    Leave a comment:


  • nucacidhunter
    replied
    The explanation so far leaves two possibilities open. One is library diversity (particularly in the positions with low Q scores) and the other is faulty reagents. Illumine has had some recalls recently. I wonder what was %PF for the run.

    Leave a comment:


  • WhiteSeal
    replied
    @GenoMax: samples were checked on BioAnalyzer to check the length of the fragments (all >700bp) so this should not happen. Also, we have seen this with almost all our 600PE v3 runs on the MiSeq.

    @NextGenSeq: MiSeq run 600PEv3 with 566k clusters gave - according to illumina output - ~14M reads, but in read two at cycle 550 already 5M reads have a Qscore <Q30 (not good), and at cycle 600 12M reads have a Qscore <Q30 (really not good). And then this was a 'good' run.

    Leave a comment:


  • NextGenSeq
    replied
    IMHO the v4 chemistry isn't worth the extra money.

    How many reads did you get?

    Leave a comment:


  • GenoMax
    replied
    Drop-off in quality accompanies presence of short inserts (when one would start reading to the adapter on other end). Have you checked/eliminated that possibility for this data set. It seems to happen more often that one would expect.

    Leave a comment:


  • WhiteSeal
    replied
    We are seeing a lot of issues with these 600 runs. There is a significant decrease in Qscores around 200 cycles (first read) and then around cycle 450 of the second read (a picture of %>Q30 is attached). We've had a lot of data that just isn't good enough for data analysis.. and I mean a lot of data. Certainly it is not what the kit should return as a result. I mean it is a 2x300 kit. Not a 2x maybe 300 if you are lucky. We've contacted illumina about this and have received compensation for 3 kits already (hoping for more) but has anyone else tried this.

    Leave a comment:


  • DJ Flowcell
    replied
    Yeah, we are experiencing the same thing. The chemistry has been improved though, although the cycle time is back up to 5min (up from 4.3min for the v2 kits). By the end of the first 300bp read, for %perfect reads, the 0 error metric is down to about 45%

    Leave a comment:


  • nickloman
    replied
    Yes, we are too.

    Leave a comment:


  • Luvs_2_sequence
    started a topic MiSeq error rate for 600 cycle PE

    MiSeq error rate for 600 cycle PE

    Was a bit alarmed at the run for 600 cycle chemistry we did the other day. Error profile shows that along the read, the error rate increases dramatically. Even only 200 cycles into the first read, the error rate has already surpassed 1%
    The error rate for read 2 is similar.

    Is this expected? We were a bit shocked to see this high of any error rate and was wondering if others were seeing that doing 600 cycle runs.

    Let me know

Latest Articles

Collapse

  • SEQadmin2
    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
    by SEQadmin2




    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
    Yesterday, 06:46 AM
  • SEQadmin2
    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
    by SEQadmin2



    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
    ...
    10-14-2026, 09:55 AM

ad_right_rmr

Collapse

News

Collapse

Topics Statistics Last Post
Started by SEQadmin2, 10-28-2026, 09:57 AM
0 responses
29 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-28-2026, 09:51 AM
0 responses
18 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 11:13 AM
0 responses
71 views
0 reactions
Last Post SEQadmin2  
Started by SEQadmin2, 10-22-2026, 10:48 AM
0 responses
48 views
0 reactions
Last Post SEQadmin2  
Working...