  • docbio
    Member
    • Jul 2012
    • 26

    Half reactions for NexteraXT

    Howdy all,

    Wondering if anyone has had any success running "half reactions" of NexteraXT libraries or otherwise stretching those reagents? At the end of the library cleanup, if memory serves you only carry a little more than half of the reaction into bead based balancing. The bulk of our sequencing cost is in the actual library preparation steps, so cutting that in half is a tempting experiment but it would involve pipetting some very small volumes (0.5 ul). Does anyone have any experience with trying this?

    Best,
    docbio
  • dfhdfh
    Member
    • Jan 2013
    • 46

    #2
    I've also thought about this. However, I think it's fine as is, since, as you already mentioned, you would need to pipet very small volumes. Additionally, I like to use the left-over of the library clean-up to do some QC (although you need only a few µls for that).

    Comment

    • kwaraska
      Senior Member
      • Nov 2008
      • 131

      #3
      The Fluidigm C1 which is single cell cDNA uses 1/4 reactions so I believe it is tied to DNA/cDNA quantity. The only change in the 1/4 reaction is the first PCR step is 10 min instead of 5 min. I assume that is to give the tagmentation better working conditions since the volumes are so small.

      Then we just use Tape Station to normalize the samples.

      Comment

      • docbio
        Member
        • Jul 2012
        • 26

        #4
        Originally posted by kwaraska View Post
        The Fluidigm C1 which is single cell cDNA uses 1/4 reactions so I believe it is tied to DNA/cDNA quantity. The only change in the 1/4 reaction is the first PCR step is 10 min instead of 5 min. I assume that is to give the tagmentation better working conditions since the volumes are so small.

        Then we just use Tape Station to normalize the samples.
        Interesting. This is useful info, I'm assuming the Fluidigm C1 protocol is online? I'm sure the concentration of DNA is critical to the tagmentation reaction. If we could move to half reactions it would save a lot of money. The only trick would be consistency. When we're normalizing for HiSeq (not going through bead based) the concentration of the finished libraries can vary somewhat.

        Tape Station looks interesting. We're using Qubit or Quant-IT at the moment.

        Best,
        docbio

        Comment

        • docbio
          Member
          • Jul 2012
          • 26

          #5
          FYI, just as a follow up, we found that half reactions and even one quarter reactions work very well with NexteraXT in our hands, although the one quarter reaction size requires very precise pipetting that is only really amenable to individual pipetting instead of using a multichannel. We've stuck with half reactions and have great success in cutting costs while churning out good sequence data.

          docbio

          Comment

          • Simone78
            Senior Member
            • Oct 2010
            • 208

            #6
            if you are interested in making your own Tn5, you could take a look at our most recent paper: http://www.ncbi.nlm.nih.gov/pubmed/25079858
            The plasmid is available from Addgene (ID 60240). Making the other reagents is really a piece of cake. The only thing we still buy from Illumina are the index oligos...and we dilute them 1:5! For inputs <1 ng DNA that amount is more than enough!

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
              by SEQadmin2



              Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
              ...
              07-09-2026, 11:10 AM
            • SEQadmin2
              Cancer Drug Resistance: The Lingering Barrier to Rising Survival
              by SEQadmin2



              Cancer survival rates have significantly increased in the last few decades in the United States, reaching a combined 70% 5-year survival rate by 2021. Behind this number, there are years of research to find new therapies, drug targets, and early detection methods. But there is one core challenge that keeps slowing down these advances, and it’s about drug resistance.

              There is no single reason why many patients don’t respond to treatment as expected. Cancer is...
              07-08-2026, 05:17 AM
            • GATTACAT
              Reply to Nine Things a Sample Prep Scientist Thinks About Before Sequencing
              by GATTACAT
              Love this - good data definitely starts from good input, and poor input can only give relatively poor data. I particularly like the mention of Nanodrop/absorbance based methods for quantification. It's such a toss up if you'll get an accurate reading or what amounts to a randomly generated number, and a lot of library/sequencing related issues can be traced back to poor quant.
              07-01-2026, 11:43 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, Today, 10:26 AM
            0 responses
            7 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-09-2026, 10:04 AM
            0 responses
            24 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-08-2026, 10:08 AM
            0 responses
            15 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 07-07-2026, 11:05 AM
            0 responses
            33 views
            0 reactions
            Last Post SEQadmin2  
            Working...