Has anybody tried generating clusters from libraries that are already below 10nM? How do you denature the sample and maintain the number of clusters you are aiming for? If you add NaOH it dilutes the sample down even further. Has anyone tried using a lower concentration of NaOH or a different method for denaturation (eg heat denaturing)?
Unconfigured Ad
Collapse
X
-
We have loaded as low as 1nM libraries. We just scale the reaction. Take a 5nM reaction that we want to load at 10pM:
Load 6 ul of DNA (to denature at 1.5nM), 1ul of NaOH and 15ul EB buffer. Then dilute 6.7ul of the denatured library.
For 1nM it is slightly more complicated - 19ul DNA (denatured at 0.95nM), 1ul NaOH. The dilute 7.37ul for a 7pM reaction
-
Thanks hiosinjl. I just realised my typo mistake. Thats also what we do for libraries down to 1nM.
What I meant to ask is - how to dilute libraries that are below 1nM? If we just scale things at this point, the NaOH becomes too concentrated in the final dilution for clustering to be effective.
Comment
-
Neutralize the NaOH with HCl
We do this all the time, down as low as 10's of pM for some libraries. After the 5 min. incubation in 0.1N NaOH, we place the tube on ice and add an equal volume of HCl at the same concentration as the NaOH we added. Top up with the appropriate amount of cold hyb solution and you're good to go. pH is now neutral, just a little extra NaCl in there.
If you look at the Sanger method papers, they have a more complicated version for loading low concentration libraries if you look in the supplemental methods.
Comment
-
Hi,
I'm just going through these calculations to make sure I'm doing this right...
Do you mean 6uL of DNA, 1uL NaOH and 13uL EB... or have I misunderstood?Originally posted by hoisinjl View PostLoad 6 ul of DNA (to denature at 1.5nM), 1ul of NaOH and 15ul EB buffer. Then dilute 6.7ul of the denatured library.
Cheers,
Comment
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...-
Channel: Articles
10-30-2026, 06:46 AM -
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, 10-28-2026, 09:57 AM
|
0 responses
45 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:57 AM
|
||
|
Started by SEQadmin2, 10-28-2026, 09:51 AM
|
0 responses
28 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
76 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
49 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Comment