Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Number6
    Member
    • Feb 2009
    • 19

    #1

    PCR-free clustering density issues?

    Here are some observations for clustering density with PCR-free DNA preps, and I'm wondering if anyone else has observed the same? On several PCR-free preps, we run the samples at our usual concentration, and get ~60-70% of the expected yield. This probably isn't unexpected, but when we increase the concentration on subsequent runs, the cluster density hits a ceiling and then starts decreasing (sometimes dramatically). Here are some numbers from a HiSeq2000 high output flowcell run with three different concentrations of the same library:

    Library at 8 pM: 177M total reads/164M PF reads
    Same library at 10 pM: 171M total reads/160M PF reads
    Same library at 12 pM: 141M total reads/134M PF reads

    The cluster plots for the higher concentrations don't look like an overclustered lane (with a lower PF rate). Images look fine....just lower density.

    I assume this is because of the library molecules in the mix with only one adaptor ligated on the ends, not both, interfering with cluster formation.

    Any ideas/comments? Is this too much DNA going into the ligation step? Any other solutions?
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    This probably isn't unexpected, but when we increase the concentration on subsequent runs, the cluster density hits a ceiling and then starts decreasing (sometimes dramatically).
    This seems like typical over clustering and your proceeding data also indicates it. But, then you are saying that this is not the case based on other evidence.

    I assume this is because of the library molecules in the mix with only one adaptor ligated on the ends, not both, interfering with cluster formation.
    That logically is not possible. Fragments with adapter in one end can hybridise to flow cell but they will not be amplified to form clusters, so their interference with clusters would be minimal.

    Any ideas/comments? Is this too much DNA going into the ligation step? Any other solutions?)
    I wonder what DNA you are referring to? Providing some thumbnail images from those runs of the same library with various input and actual cluster density reported in SAV might give some further ideas to explore. Also was the library for different runs was denatured in one batch or were they denatured separately?

    Comment

    • Number6
      Member
      • Feb 2009
      • 19

      #3
      Thanks, nucacidhunter. This is the only time we've set up a titration to look at these, but we've noticed this occasionally with (our) PCR-free preps. Adding more DNA to the flowcell doesn't increase the cluster density and the PF rate doesn't drop like a typical over-clustered lane (like lane 5 in the ppt I've uploaded).

      I agree that partially-ligated DNA fragments wouldn't be expected to interfere; I only suggest this because this is the only explanation that makes any sense to us....seems to me there's *something* in this library that's interfering with cluster generation.
      Attached Files

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        If your cluster number in PhiX control lane and libraries in other three lanes were as you expect, issues with clustering and SBS reagents and the machines would be ruled out. If occasional variation also is only seen for PCR free libraries in the same flow cell that rules out the occasional sequencer issues as well. I have not noticed any difference between these and Nano or other TruSeq libraries clustering. Maybe logging an Illumina technical support case would be good to see what is their explanation.

        Comment

        • NextGenSeq
          Senior Member
          • Apr 2009
          • 482

          #5
          PCR free libraries have higher GC content. You need to denature your libraries using fresh NaOH.

          Illumina tech support suggested that we use the 1ml aliquots of 1N NaOH from Teknova and use a fresh tube every week. NaOH solutions absorb CO2 and change pH over time.

          Explore high-quality sodium hydroxide (NaOH) solutions for pH adjustment, reagent preparation, and laboratory workflows.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
            by SEQadmin2




            NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
            Today, 06:46 AM
          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-28-2026, 09:57 AM
          0 responses
          23 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-28-2026, 09:51 AM
          0 responses
          15 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          68 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          48 views
          0 reactions
          Last Post SEQadmin2  
          Working...