    Shelf life

    Hi All,

    What is the oldest kit that anyone has used? Do you see a drop-off in output quality and/or quantity with kits that have been in storage for a while? We're contemplating a large order simply because it's significantly cheaper that way, but I'm a little worried about how well they keep (so far, we've only ever used kits that are used within a week or so of their arrival).

    Cheers,

    Scott.
  • swbarnes2
    Senior Member
    • May 2008
    • 910

    #2
    Older reagents

    We had a badish run, and the first thing Illumina told us was that our reagents we older than the shelf life. Not by a lot either (and no one had told us that the shelf-life was so short!), less than 6 months.

    So if your reagents are old, and you have a problem, Illumina will tell you that you have old reagents, and to just run again with fresh ones. Not a big help.

    Because of that, we haven't dared to try any runs with old reagents.
    Last edited by swbarnes2; 08-27-2008, 03:37 PM. Reason: added extra info

    Comment

    • ScottC
      Senior Member
      • Jan 2008
      • 244

      #3
      Hi,

      Thanks for the info. How old were the reagents?

      Comment

      • SCurB
        Junior Member
        • Oct 2008
        • 2

        #4
        I have used kits at least as old as the shelf life (3 months) and possibly as old as 4 months and had them work fine.

        However, I probably wouldn't want to push your luck given the price of these kits.

        Comment

        • andibody
          Member
          • Mar 2008
          • 45

          #5
          I suppose we are talking about the Sequencing kits.
          What about the others, sample prep and cluster gen kit? They should in principle have a much longer shelf life.
          Mostly Enzymes and buffers at -20C...

          Comment

          • kwebb
            Member
            • Jul 2008
            • 21

            #6
            Does anyone have any recent experience with older kits/reagents? What is the oldest kit that has been used successfully?

            Thanks!

            Comment

            • kwebb
              Member
              • Jul 2008
              • 21

              #7
              To elaborate further - we have cluster generation kits, flowcells and sequencing kits (v2, 36 cycle, single end) all purchased in Aug 2008. Can anyone comment on whether or not they think we should even bother trying to sequence with these kits?
              Does anyone know the shelf life of the cluster generation kits and/or flowcells?

              Comment

              • james hadfield
                Moderator
                Cambridge, UK
                Community Forum
                • Feb 2008
                • 224

                #8
                We have ued PE kits that were many months out of dat, I would need to check in the lab but at least August of 2008. Results were absoutley fine, PhiX QC was good.

                Comment

                • picabo
                  Junior Member
                  • Jun 2009
                  • 5

                  #9
                  I believe the cluster kits have a shelf life of 6 months and the sequencing kits have a shelf life of 3 months.

                  We used an old kit and the run failed.

                  The cluster kit has formamide and NaOH and both of those go bad over time. If you use fresh supplies of those reagents, you might be able to get a successful run.

                  When we place orders, we place a large order and have them send the kits in multiple shipments. That way we get the discount, but don't have to worry about the kits going bad.

                  Comment

                  • csquared
                    Member
                    • May 2008
                    • 67

                    #10
                    Cluster kits have a very long shelf life. Just used a couple kits that were between 8 and 12 months old with no trouble. Only issue is that the small volume of 2M NaOH suplied in the kit tends to evaporate slightly in the older kits. Since the vol of 2M NaOH that is supplied in the PE kits is much more than needed, we just borrow some from that kit. As long as the flow cell tubes are stored upright, we haven't seen any issues with age. We have not been as confident in the seq kits and so far have used all in the 3 month shelf life.

                    The instrument is pretty forgiving for the most part. Including the time from primer hyb to seq that Illumina says needs to be 4 hours or less. We've let PE runs sit on the cluster station overnight and started seq in the morning with no issues. If first base looks poor, redo the primer hyb and all is good.
                    HudsonAlpha Institute for Biotechnology
                    http://www.hudsonalpha.org/gsl

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                      by SEQadmin2




                      NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                      Yesterday, 06:46 AM
                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-28-2026, 09:57 AM
                    0 responses
                    25 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-28-2026, 09:51 AM
                    0 responses
                    15 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    71 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    48 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...