    MiSeq Wash question

    We just had our MiSeq installed about a month ago, and haven't yet sequenced anything on it. I've just been performing maintenance washes weekly to keep the fluidics system in good shape.

    I just performed a Standby wash this week since the instrument has been idle for over 30 days. Now that I've done the Standby wash, is it necessary to keep doing Maintenance washes every week given that we haven't run any DNA yet on the machine? Or can we just let it go for now, until we actually start running samples? (at that point I think we'd perform a maintenance wash before running the DNA)
  • bunce
    Member
    • Sep 2012
    • 55

    #2
    Hi Krina,
    Just keep following the wash instructions on the MiSeq - the bottom line is that the more you use the fluidics system the less chance you will have of developing biofilms or blockages. I would continue to do maintenance washes each week.

    On a side note we have been using the Illumina bleach wash protocols after runs and have found this to be easy and is a good safeguard to (at least minimise) carry-over. Cheers, Mike

    Comment

    • Krina796
      Junior Member
      • Jul 2014
      • 7

      #3
      Hi bunce,

      Thank you for your help, I really appreciate it. Did you find the bleach wash protocol on the Illumina website?

      Best,
      Krina

      Comment

      • bunce
        Member
        • Sep 2012
        • 55

        #4
        Hi Krina, the wash protocol is attached together with the xml - there maybe a more up-to-date version of this now? I am not sure how easy this is to find on the illumina site. You just need to make sure the xml finds it into the correct place (v2, v3 etc) - some users were putting it in the wrong location apparently.

        We have replaced the post-run wash with the bleach protocol... then after the bleach was we run a 'pseudo'-maintainance wash (three times with the same wash solution). We have done the latter to make sure the bleach is flushed - it is probably not necessary but we have always done it.
        Attached Files

        Comment

        • Krina796
          Junior Member
          • Jul 2014
          • 7

          #5
          This is very helpful, thank you very much bunce!

          Comment

          • james hadfield
            Moderator
            Cambridge, UK
            Community Forum
            • Feb 2008
            • 224

            #6
            How to get the bleach protocol

            Hi, I just wrote a new post on my blog, http://core-genomics.blogspot.co.uk/...ng-bleach.html which describes the bleach protocol and how to get it.

            Comment

            • GA-J
              Member
              • Jul 2015
              • 28

              #7
              Miseq wash question

              Does anyone know where I may find miseq wash record from miseq computer? Any folder or file? Thanks in advance.

              Another one: if I have next run maybe in 2 or 3 weeks, I don't want to do stand-by wash, what will you suggest? Thanks again.

              Comment

              • Jessica_L
                Senior Member
                • Feb 2010
                • 117

                #8
                I'm not sure about a log for the wash dates and times on the MiSeq.

                If you plan to not run anything for 2 weeks, just do a post-run wash each week and you should be fine. I've heard lately that Illumina does not recommend doing the standby wash and then leaving the instrument idle for multiple weeks-- apparently they prefer if you just do a post-run wash each week with monthly maintenance washes. As others have said already, the more you use the fluidics system, the less likely the instrument is to develop blockages or films in the lines.

                Comment

                • GA-J
                  Member
                  • Jul 2015
                  • 28

                  #9
                  Thank you, Jessica L, sounds very clear and simple to follow.

                  Comment

                  • mlo
                    Junior Member
                    • Feb 2016
                    • 1

                    #10
                    Does anyone know where I may find WashStatus.xml files on the MiSeq?

                    Comment

                    • misterc
                      Member
                      • Jan 2016
                      • 46

                      #11
                      I think it's root level on the C:\ or D:\ drive.

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM
                      • SEQadmin2
                        New Genomics Technologies Take Aim at Long-Standing Limits
                        by SEQadmin2


                        Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                        We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                        ...
                        09-28-2026, 10:25 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, Yesterday, 10:52 AM
                      0 responses
                      8 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-14-2026, 05:39 AM
                      0 responses
                      19 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-13-2026, 09:56 AM
                      0 responses
                      15 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-09-2026, 11:14 AM
                      0 responses
                      27 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...