Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • Morten
    Junior Member
    • Jun 2010
    • 9

    #1

    Primer contamination in library

    Hi all,

    We have been preparing libraries for Illumina sequencing using the ScriptSeq v2 kit. This gave us a bit of trouble and the resulting library is not really perfect. There is a large peak at 69 bp.
    Please see the bioanalyzer image.

    The peak at 69 bp is about the size of a single primer. Not an adapter-dimer.
    My question is would this be a problem for cluster generation and sequencing?

    This type of contamination would not be amplified, so I guess it would not be a problem, but I might be missing something? Would you feel safe sequencing this library?

    Thank you,
    Morten.
    Attached Files
  • tegudet
    Junior Member
    • Oct 2014
    • 1

    #2
    Hi Morten,
    I think this product could be sequenced and contaminate your reads.
    I suggest you to use BluePippin, where you select the size of the product you want (let say 90bp to 400bp), so that you exclude 69bp primer contamination.

    Good luck,

    tegudet

    Comment

    • Morten
      Junior Member
      • Jun 2010
      • 9

      #3
      Thank you for your reply tegudet.
      However, the yield of this library is not good at all. There is simply not enough material to attempt a size selection.
      As I see it my options are either sequence as is or remake the library.

      Comment

      • microgirl123
        Senior Member
        • Jun 2012
        • 199

        #4
        We sequence libraries with excess adapter on the MiSeq all the time. It's the adapter dimers that cause the problem. Single adapters can bind to the flow cell, but cannot bridge amplify so are not sequenced. Adapter dimers both bind and bridge amplify so they cause significant problems.

        Comment

        • Morten
          Junior Member
          • Jun 2010
          • 9

          #5
          Thanks a lot for the first hand info!
          I will go ahead and have the library sequenced.

          Comment

          • Morgane_AUS
            Member
            • Aug 2014
            • 25

            #6
            Originally posted by Morten View Post
            Thanks a lot for the first hand info!
            I will go ahead and have the library sequenced.
            How did it go ?

            Comment

            • Morgane_AUS
              Member
              • Aug 2014
              • 25

              #7
              Adapter dimer contamination

              Originally posted by microgirl123 View Post
              We sequence libraries with excess adapter on the MiSeq all the time. It's the adapter dimers that cause the problem. Single adapters can bind to the flow cell, but cannot bridge amplify so are not sequenced. Adapter dimers both bind and bridge amplify so they cause significant problems.
              Hi,

              Have you ever attempted to sequence a library with such a high proportion of adapter dimers ? Please see the gel here:


              We planned to use a single end 50 bp read (MiSeq V2 - 15 M reads, 24 samples, genome size 4.41 Mbp). Basically the cell might be flogged with adapters. My reasoning is as follow

              Cartridge: 15 Million reads
              Targeted number of reads: 10 Million reads
              25% is ChiPed DNA: 2.5 Million reads
              24 Samples: 0.1 Million reads/sample
              50 bp per read: 5.2 Million bp /sample ---> 1x genome coverage ->BAD

              Any hint would be greatly appreciated.

              Morgane
              Last edited by Morgane_AUS; 12-02-2014, 09:33 PM.

              Comment

              • cmbetts
                Senior Member
                • Jun 2012
                • 120

                #8
                Are you sure that's actually adapter dimer and not a primer-dimer o the library amplification primers? The full Illumina adapters are about twice the size of your peak (~120bp, assuming these are TruSeq compatible), and primer dimers essentially invisible to the sequencer since they don't have all of the necessary flowcell and priming sequences. I do see a little of the 120bp peak, but I'd expect it to be <5% of reads...

                Comment

                • Morgane_AUS
                  Member
                  • Aug 2014
                  • 25

                  #9
                  Originally posted by cmbetts View Post
                  Are you sure that's actually adapter dimer and not a primer-dimer o the library amplification primers? The full Illumina adapters are about twice the size of your peak (~120bp, assuming these are TruSeq compatible), and primer dimers essentially invisible to the sequencer since they don't have all of the necessary flowcell and priming sequences. I do see a little of the 120bp peak, but I'd expect it to be <5% of reads...
                  Mmm.. maybe you're right.. NEBNext ultra DNA adaptor are about 70 bp long and form a hairpin that would produce a ~35 bp double stranded fragment.

                  The primers though are 60 bp and could be responsible of the 120 bp peak.

                  Now what I'm wondering is whether the adaptors can make concatemers ?

                  Are you saying that you would sequence a library looking like this ?

                  Thanks !
                  Last edited by Morgane_AUS; 12-02-2014, 05:58 PM.

                  Comment

                  • cmbetts
                    Senior Member
                    • Jun 2012
                    • 120

                    #10
                    I guess it depends on how precious your sample is. That's a slightly higher MW peak than I'm used to seeing for library amplification primers (but I've also nevr used this particular kit), and while I wouldn't expect them to sequence, if they are true primer-dimers and not excess primers, they might still have some of the necessary sequences to interfere with proper sequencing of the library. I'd also be concerned about that super-wide library size distribution as well. It would be difficult to know just how much to load on the sequencer since I'd expect that a sizable amount of it is too large to successfully cluster. I'd still be tempted to try a two-sided Ampure size selection and elute in a small volume to concentrate.

                    Comment

                    • Morgane_AUS
                      Member
                      • Aug 2014
                      • 25

                      #11
                      Originally posted by cmbetts View Post
                      I guess it depends on how precious your sample is. That's a slightly higher MW peak than I'm used to seeing for library amplification primers (but I've also nevr used this particular kit), and while I wouldn't expect them to sequence, if they are true primer-dimers and not excess primers, they might still have some of the necessary sequences to interfere with proper sequencing of the library. I'd also be concerned about that super-wide library size distribution as well. It would be difficult to know just how much to load on the sequencer since I'd expect that a sizable amount of it is too large to successfully cluster. I'd still be tempted to try a two-sided Ampure size selection and elute in a small volume to concentrate.

                      I didn't do the two sided size selection because the protocol didn't recommend to do it when starting the lib prep with less than 100 ng (and I started with 0.25 ng !). I'm worried I will loose the DNA if I do it now. These samples are quite precious (results of working the last 5 week-ends - it is a side project) and I won't be able to repeat the whole experiment due to other time constraints.

                      I was thinking about doing one bead purification step to remove the 120 bp fragment. Do you think large fragments can affect the whole sequencing run?

                      Thanks for you advice !

                      Comment

                      • nucacidhunter
                        Jafar Jabbari
                        • Jan 2013
                        • 1250

                        #12
                        Originally posted by Morgane_AUS View Post
                        Hi,

                        Have you ever attempted to sequence a library with such a high proportion of adapter dimers ? Please see the gel there: https://bioinfo4biologists.wordpress...dapter-dimers/
                        Could you attach your figure. I cannot open this link.

                        Comment

                        • Morgane_AUS
                          Member
                          • Aug 2014
                          • 25

                          #13
                          Originally posted by nucacidhunter View Post
                          Could you attach your figure. I cannot open this link.
                          sorry I don't know what's wrong with the link.

                          Here it is. Normal means that ChIPed DNA was used directly as input of the lib prep. WGA means ChIPed DNA was amplified using a whole genome amplification kit before lib prep. I started the lib prep with 0.25 ng of ChIPed DNA and 2.5 ng of amplified ChIPed DNA.
                          Attached Files

                          Comment

                          • Morgane_AUS
                            Member
                            • Aug 2014
                            • 25

                            #14
                            updated link

                            Comment

                            • nucacidhunter
                              Jafar Jabbari
                              • Jan 2013
                              • 1250

                              #15
                              It is difficult to figure out the nature of band in 100-200 bp mostly because of overloaded ladder. A TapeStation or Bioanalyser would have been better for diagnosis. If you had a no-template library prep that also would have helped with working out the possible source of that band. In any case, you can clean it with 0.9x bead twice. Larger fragments won’t an issue as cluster generation favours smaller fragments.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              43 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              26 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              74 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...