    What does it mean by "150/350 PE reads" from MiSeq?

    http://www.ncbi.nlm.nih.gov/sra/SRX573008[accn]

    I downloaded this SRA data. I noticed that the R1 reads are 150bp but the R2 reads are 350bp. How could that be? Aren't they supposed to be the same length???
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Unequal length on MiSeq is permitted.

    Comment

    • ymc
      Senior Member
      • Mar 2010
      • 496

      #3
      Originally posted by nucacidhunter View Post
      Unequal length on MiSeq is permitted.
      So this data was generated by a 2x250 run?

      Why do people want to do unequal length?

      Thanks in advance!

      Comment

      • Brian Bushnell
        Super Moderator
        • Jan 2014
        • 2709

        #4
        All else equal... longer reads are more useful. 350bp reads can span repeats that 250bp reads can't, for example.

        However, all else is NOT equal, and I seriously doubt the quality at the end of those 350bp reads is very good.

        Comment

        • ymc
          Senior Member
          • Mar 2010
          • 496

          #5
          Originally posted by Brian Bushnell View Post
          All else equal... longer reads are more useful. 350bp reads can span repeats that 250bp reads can't, for example.

          However, all else is NOT equal, and I seriously doubt the quality at the end of those 350bp reads is very good.
          Thanks a lot for your reply. Are there any limitations to adjusting the lengths? Can you do 1/499 or even 0/500???

          Comment

          • HESmith
            Senior Member
            • Oct 2009
            • 512

            #6
            Well, you need 4-5 cycles in the first read to call clusters, and at least 25 cycles for base-calling. Plus, as Brian noted, the quality drops precipitously with the longer reads. But I think that the instrument could be programmed for 25/475.

            Comment

            • ymc
              Senior Member
              • Mar 2010
              • 496

              #7
              Originally posted by HESmith View Post
              Well, you need 4-5 cycles in the first read to call clusters, and at least 25 cycles for base-calling. Plus, as Brian noted, the quality drops precipitously with the longer reads. But I think that the instrument could be programmed for 25/475.
              Thanks for your reply. It is good to know that this is possible anyway. 25/575 for the 2x300 kit can come in handy sometimes.

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                Or you could do a 600 cycle SE run.

                Comment

                • ymc
                  Senior Member
                  • Mar 2010
                  • 496

                  #9
                  Originally posted by GenoMax View Post
                  Or you could do a 600 cycle SE run.
                  So you can turn PE into SE in V3 chemistry but not V2?

                  Comment

                  • nucacidhunter
                    Jafar Jabbari
                    • Jan 2013
                    • 1250

                    #10
                    It can be done with v2 too and possibly on HiSeq. In MiSeq after cycle 200 the quality drops and is in lowest value at 300 cycle. Increasing cycle beyond that results in low quality so it may not be useful depending on intended use of data.

                    Comment

                    • matth431
                      Member
                      • Jan 2012
                      • 36

                      #11
                      Originally posted by ymc View Post
                      So this data was generated by a 2x250 run?

                      Why do people want to do unequal length?

                      Thanks in advance!
                      One reason we've done this is where there's a high proportion of short fragments in the library - we've had a user make Nextera libraries from very low amounts of viral DNA which tend to over-tagment, so lots of ~200bp inserts. If the bulk of the reads sequence past the end of the molecules, the machine loses focus and the run will abort prior to the index reads.

                      So we often do 100/400 reads to ensure we get index data for all clusters, then hopefully get as much of a reverse read as possible for the small percentage of larger fragments in the library - if the run aborts on read 2, we can at least recover the data, but if it aborts on read 1, the data is useless as it can't be demultiplexed.

                      Comment

                      Latest Articles

                      Collapse

                      • SEQadmin2
                        16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                        by SEQadmin2




                        NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                        Today, 06:46 AM
                      • SEQadmin2
                        Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                        by SEQadmin2



                        Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                        Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                        ...
                        10-14-2026, 09:55 AM

                      ad_right_rmr

                      Collapse

                      News

                      Collapse

                      Topics Statistics Last Post
                      Started by SEQadmin2, 10-28-2026, 09:57 AM
                      0 responses
                      22 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-28-2026, 09:51 AM
                      0 responses
                      14 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 11:13 AM
                      0 responses
                      67 views
                      0 reactions
                      Last Post SEQadmin2  
                      Started by SEQadmin2, 10-22-2026, 10:48 AM
                      0 responses
                      47 views
                      0 reactions
                      Last Post SEQadmin2  
                      Working...