    384 multiplexed 16S libraries - low quality

    Hello Everyone,

    I have done two 2*300bp runs on the Miseq with 384 multiplexed 16S libraries (derived from mice feces). For each run the quality of the reads has been really poor. I am even using 10% PhiX double what illumina recommends. For the first run I had a cluster density of 1109k/mm2 with 89% cluster passing filter and Q30 77% and this last run 1118k/mm2 passing filter of 87.7% and QC30 67%.

    Have attached the results of the last run….Any idea what I can do have better runs? Or why the quality is so bad.

    (I think it has to do with the mice that are producing my sample material live in a control and they should all have the same intestinal flora. )
    Attached Files
  • microgirl123
    Senior Member
    • Jun 2012
    • 199

    #2
    This looks like pretty good data for a low diversity run, which this is even though you have a 10% phiX spike. Your cluster density is is a bit high for a low diversity run though - I'd try to drop it down to ~800-900K.

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      Is this a V3 or V2 run?

      Like microgirl123 said this is a classic signature of low nucleotide diversity samples.

      In spite of the poor quality values you may find the sequence adequate (if you are aligning to a reference). If you are looking to do de novo assembly then you would want to re-run at lower concentration.

      Comment

      • kmcarr
        Senior Member
        • May 2008
        • 1181

        #4
        You data look very typical for V3 PE300 runs of 16S amplicons, which is exactly why we advise researcher against using MiSeq V3 PE300 runs for amplicon sequencing. The V2 PE250 chemistry works very well for 16S sequencing but all the V3 PE300 runs we have attempted look very much like yours. We get just as much quality data from a V2 PE250 as we do from a V3 PE300 at lower cost.

        Comment

        • JBKri
          Member
          • Jan 2014
          • 80

          #5
          Originally posted by kmcarr View Post
          You data look very typical for V3 PE300 runs of 16S amplicons, which is exactly why we advise researcher against using MiSeq V3 PE300 runs for amplicon sequencing. The V2 PE250 chemistry works very well for 16S sequencing but all the V3 PE300 runs we have attempted look very much like yours. We get just as much quality data from a V2 PE250 as we do from a V3 PE300 at lower cost.
          Do you know why the v3PE300 runs have problems with amplicon sequencing? In theory the quality should be better...
          Fadrosh et al (http://www.microbiomejournal.com/content/2/1/6) has figures showing better quality from V3 PE300 cycles than V2 PE250 cycles (in supplementary figures).
          Jon

          Comment

          • kmcarr
            Senior Member
            • May 2008
            • 1181

            #6
            Originally posted by JBKri View Post
            Do you know why the v3PE300 runs have problems with amplicon sequencing? In theory the quality should be better...
            Fadrosh et al (http://www.microbiomejournal.com/content/2/1/6) has figures showing better quality from V3 PE300 cycles than V2 PE250 cycles (in supplementary figures).
            Jon
            Jon,

            That paper uses in-line barcodes (i.e. that are the first part of each read instead of contained in independent index reads) plus 0-7bp "heterogeneity" spacers. In other words they designed their library construction protocol to create as much base diversity as possible, especially at the beginning of the reads. This is different from the case I was describing in which all clusters have fragments starting at exactly the same spot of the 16S and thus have virtually no base diversity.

            Comment

            • JBKri
              Member
              • Jan 2014
              • 80

              #7
              Originally posted by kmcarr View Post
              Jon,

              That paper uses in-line barcodes (i.e. that are the first part of each read instead of contained in independent index reads) plus 0-7bp "heterogeneity" spacers. In other words they designed their library construction protocol to create as much base diversity as possible, especially at the beginning of the reads. This is different from the case I was describing in which all clusters have fragments starting at exactly the same spot of the 16S and thus have virtually no base diversity.
              So, are you saying the V3 reagents seem more sensitive to low diversity than the V2 reagents?

              Comment

              • microgirl123
                Senior Member
                • Jun 2012
                • 199

                #8
                My experience is that it has nothing to do with the reagents. It's just that the extra 50 bp per read in a 600-cycle V3 kit are junk with low diversity samples (they're not great with a balanced genome either). You can get more clusters with a V3 kit because it reads more of the flow cell, which is the reason some of our customers like to use it.

                So... we perform 600-cycle V3 low-diversity runs all the time, but warn our customers to have plenty of overlap so they don't need data from the last 50-bp of each read.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  Beyond CRISPR/Cas9: Understand, Choose, and Use the Right Genome Editing Tool
                  by SEQadmin2



                  CRISPR/Cas9 sparked the gene editing revolution for both research and therapeutics.1 But this system still showed severe issues that limited its applications. The most prominent were the heavy reliance on PAM sequences, delivery limitations, double-stranded breaks that prompt unintended edits and cell death, and editing inefficiency (both in targeting and in knock-in reliability).

                  Despite this, “CRISPR helped turn genome editing from a specialized technique into
                  ...
                  Yesterday, 11:01 AM
                • SEQadmin2
                  Proteomic Platforms: How to Choose the Right Analytical Strategy to Improve Detection and Clinical Applications
                  by SEQadmin2


                  Proteomics platforms are evolving rapidly, with advances in mass spectrometry and affinity-based approaches expanding what researchers can detect and at what scale. As the field moves toward deeper proteome coverage and clinical applications, scientists face an increasingly complex landscape of tools. This article will explore how researchers are navigating these choices to find the right platform for their work.

                  The systematic characterization of the human proteome has
                  ...
                  07-20-2026, 11:48 AM
                • SEQadmin2
                  Advanced Sequencing Platforms Tackle Neuroscience’s Toughest Genomics Problems
                  by SEQadmin2



                  Genomics studies in neuroscience face a special challenge due to the brain’s complexity and scarcity of samples. Mapping changes in cell type and state using conventional next-generation sequencing methods remains challenging. Advances in technologies like single-cell sequencing, spatial transcriptomics, and long-read sequencing have opened the door to deeper studies of the brain and diseases like Alzheimer’s, amyotrophic lateral sclerosis (ALS), and schizophrenia.
                  ...
                  07-09-2026, 11:10 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, Yesterday, 02:55 AM
                0 responses
                8 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-24-2026, 12:17 PM
                0 responses
                12 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-23-2026, 11:41 AM
                0 responses
                12 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 07-20-2026, 11:10 AM
                0 responses
                24 views
                0 reactions
                Last Post SEQadmin2  
                Working...