  • pmiguel
    Senior Member
    • Aug 2008
    • 2328

    #1

    How many HiSeqs still run a control lane?

    A "control lane" is a lane of phiX library -- if I understand correctly it was required back in the "GA" days. Kind of a "tithe" to the gods of sequencing. One lane of eight set aside to honor them. (Joking...)

    I was a little late to the Illumina scene. By the time we started running it was not necessary to run a control lane, although you could. So I think we've run 1 control lane in the last 3 years.

    Anyone still running them?

    --
    Phillip
  • Jessica_L
    Senior Member
    • Feb 2010
    • 117

    #2
    When I was working in a core lab up till about May of 2013, we were running them religiously. I'm not sure if my old lab has changed their workflow in the time I've been gone, even though Illumina has made it pretty clear that you no longer need so much PhiX.

    Short answer: I was running them up till about two years ago.

    Comment

    • kmcarr
      Senior Member
      • May 2008
      • 1181

      #3
      With recent versions of the software it is not possible to designate a control lane.

      From the HiSeq Control Software v2.2.38/Real Time Analysis v1.18.61 Release Notes (July 2014):

      I. HCS v2.2.38
      FEATURES:
      Control lanes are no longer required
      • The control lane option has been removed from all run modes: internal testing by Illumina has shown that a control lane is unnecessary when using RTA 1.18.61, which includes new features to support low diversity sequencing. See the RTA 1.18.61 section (below) for more details. When sequencing challenging samples such as a single amplicon, a 10% PhiX spike-in is recommended in place of a control lane.

      Comment

      • MU Core
        Member
        • Apr 2008
        • 60

        #4
        Our procedure is to now include a 0.1% spike-in of PhiX to each lane for run performance monitoring and troubleshooting.

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by kmcarr View Post
          With recent versions of the software it is not possible to designate a control lane.

          From the HiSeq Control Software v2.2.38/Real Time Analysis v1.18.61 Release Notes (July 2014):
          Good point. I did see that. But it didn't make much of an impression since we never run one anyway...

          --
          Phillip

          Comment

          • ETHANol
            Senior Member
            • Feb 2010
            • 308

            #6
            As mentioned with the new base caller you cannot even specify a control lane if you want to. We do a lot runs that are only whole genome bisulfite sequencing so we use to have to do a control lane, now obviously we don't. We spike in PhiX at 5-10%. The official word from Illumina is you need 10% PhiX, but the unofficial word is that you can go as low as 5% PhiX. We've gone as low as 2.5% and the data still looked good, but we target 5-10%.
            --------------
            Ethan

            Comment

            • pmiguel
              Senior Member
              • Aug 2008
              • 2328

              #7
              Originally posted by ETHANol View Post
              As mentioned with the new base caller you cannot even specify a control lane if you want to. We do a lot runs that are only whole genome bisulfite sequencing so we use to have to do a control lane, now obviously we don't. We spike in PhiX at 5-10%. The official word from Illumina is you need 10% PhiX, but the unofficial word is that you can go as low as 5% PhiX. We've gone as low as 2.5% and the data still looked good, but we target 5-10%.
              Thanks Ethan.
              Hey, I thought you got tired of doing science in Greece and had left?

              --
              Phillip

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-15-2026, 10:52 AM
              0 responses
              26 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-14-2026, 05:39 AM
              0 responses
              34 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-13-2026, 09:56 AM
              0 responses
              31 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-09-2026, 11:14 AM
              0 responses
              28 views
              0 reactions
              Last Post SEQadmin2  
              Working...