    Minimal coverage for accurate variant calling according to Illumina FDA submission?

    Dear all,
    looking up the FDA submission (510K) of the MiSeq Dx platform document, it is stated that the "Recommended minimal coverage per amplicon needed for accurate variant calling (Q(max_gt | poly_site) >=100) is 75x:



    My questions are
    1 - Is there a known rational for this 75x number?
    2 - What does "Q(max_gt | poly_site) >=100" mean?

    Thanks!
  • Brian Bushnell
    Super Moderator
    • Jan 2014
    • 2709

    #2
    If you are not getting accurate calls at 30x coverage over a particular site, 75x won't help you much, because there's some fundamental problem with your methodology. And I think by 75x per amplicon they mean 150x for a human, which seems quite excessive. I imagine that number is designed so that you can screw up a lot of stuff (reduce library complexity by over-amplifying too little input DNA, use exome-capture and an inferior aligner to yield extreme reference-bias, etc) and still have some chance of producing useful data.
    Last edited by Brian Bushnell; 04-30-2015, 08:49 AM.

    Comment

    • alunem
      Junior Member
      • Feb 2011
      • 5

      #3
      Thanks Brian,
      my focus is more on the regulatory point of view though (this is why I was interested by the FDA submission document) so I still need some references/pointers.
      Do you understand this?
      Q(max_gt | poly_site)

      Comment

      • Brian Bushnell
        Super Moderator
        • Jan 2014
        • 2709

        #4
        Sorry, no, I have not the slightest idea what that means.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Originally posted by alunem View Post
          Thanks Brian,
          my focus is more on the regulatory point of view though (this is why I was interested by the FDA submission document) so I still need some references/pointers.
          Do you understand this?
          http://www.illumina.com/documents/pr...provements.pdf

          See page 7 for explanation of Q(max_gt | poly_site) .

          Comment

          • alunem
            Junior Member
            • Feb 2011
            • 5

            #6
            Oh great, thanks.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              Yesterday, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            28 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            18 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            71 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...