    I've come across a bit of a mystery in my Core and I was wondering if anyone might have an idea as to what's going on.

    I have a customer who prepared an exosomal RNA library using a kit from System Biosciences (SBI), specifically their XRNA Exosome RNA-Seq Library Kit. There was much consultation and whatnot but in the end they generated what appeared to be a proper library.

    Upon submitting their library to my Core we did our standard Tapestation check and Kapa qPCR. Tapestation gives a single peak at ~152bp, right where the kit says it should be. qPCR says the library concentration is 2nM, exactly what I asked the submitting lab to normalize to (they quantified with Qubit).

    We load the library on our Miseq for a v3 PE75 run with 20% phiX (an amount requested by the customer). The Miseq errors out on the focus step, no cluster detected. Having a look at the focus images shows they are completely black. In full disclosure this kit uses a special primer for read 2, but that is spiked into the read 2 primer in the Miseq cartridge at the appropriate concentration and shouldn't be a factor in clustering.

    We call up TS and tell them what's going on and we run through all of the system checks we can run without an FS rep coming in. Everything checks out. Must have been a fluke right? Maybe bad reagent kit? Whatever's happening, the customer is in a hurry and needs the data. So....

    We load up the library again, this time bumping phiX up to 30% as suggested by TS. Again, run errors out on focus with no clusters detected. All images black again as well.

    Something's obviously going terribly wrong somewhere. Re-qPCR, re-Tapestation, still looks the same. Qubit quantification agrees with qPCR. As a test of the machine we load up the next person in the queue. It clusters beautifully and runs to completion without issue. Granted this was a PE250 v2 kit so we can't say it's a complete apples-to-apples comparison reagent-wise but I've never seen two kits fail back-to-back.

    In my years of Illumina sequencing I've seen libraries that won't cluster and you only see phiX, but I've never come across a library that will not cluster AND will also keep phiX from clustering. Especially at 20 and 30% loading!

    The customer contacted SBI and they claimed they only repackage reagents and never actually did any of the science to develop the kit. (Uh, what?)

    Thoughts or suggestions would be appreciated.
  • dfhdfh
    Member
    • Jan 2013
    • 46

    #2
    What is the solvent of the customer library?

    Edit: Buffer may be the better word, as the solvent most likely is water
    Last edited by dfhdfh; 05-21-2015, 06:47 AM.

    Comment

    • GW_OK
      Senior Member
      • Sep 2009
      • 411

      #3
      EB buffer. 10mM Tris.

      Comment

      • NextGenSeq
        Senior Member
        • Apr 2009
        • 482

        #4
        I had the same problem with a library made by a commercial company which they gave us to sequence (even though they had their own MiSeq). The only way I was able to get ANY data was to spike-in 50% PhiX. I got a couple of million of usable sequence but the stats were terrible.

        Comment

        • bilyl
          Member
          • Aug 2013
          • 52

          #5
          Originally posted by GW_OK View Post
          I've come across a bit of a mystery in my Core and I was wondering if anyone might have an idea as to what's going on.

          I have a customer who prepared an exosomal RNA library using a kit from System Biosciences (SBI), specifically their XRNA Exosome RNA-Seq Library Kit. There was much consultation and whatnot but in the end they generated what appeared to be a proper library.

          Upon submitting their library to my Core we did our standard Tapestation check and Kapa qPCR. Tapestation gives a single peak at ~152bp, right where the kit says it should be. qPCR says the library concentration is 2nM, exactly what I asked the submitting lab to normalize to (they quantified with Qubit).

          We load the library on our Miseq for a v3 PE75 run with 20% phiX (an amount requested by the customer). The Miseq errors out on the focus step, no cluster detected. Having a look at the focus images shows they are completely black. In full disclosure this kit uses a special primer for read 2, but that is spiked into the read 2 primer in the Miseq cartridge at the appropriate concentration and shouldn't be a factor in clustering.

          We call up TS and tell them what's going on and we run through all of the system checks we can run without an FS rep coming in. Everything checks out. Must have been a fluke right? Maybe bad reagent kit? Whatever's happening, the customer is in a hurry and needs the data. So....

          We load up the library again, this time bumping phiX up to 30% as suggested by TS. Again, run errors out on focus with no clusters detected. All images black again as well.

          Something's obviously going terribly wrong somewhere. Re-qPCR, re-Tapestation, still looks the same. Qubit quantification agrees with qPCR. As a test of the machine we load up the next person in the queue. It clusters beautifully and runs to completion without issue. Granted this was a PE250 v2 kit so we can't say it's a complete apples-to-apples comparison reagent-wise but I've never seen two kits fail back-to-back.

          In my years of Illumina sequencing I've seen libraries that won't cluster and you only see phiX, but I've never come across a library that will not cluster AND will also keep phiX from clustering. Especially at 20 and 30% loading!

          The customer contacted SBI and they claimed they only repackage reagents and never actually did any of the science to develop the kit. (Uh, what?)

          Thoughts or suggestions would be appreciated.
          Have you thought about doing a bead cleanup of the prepared library before sequencing? Maybe something is carrying over.

          Alternatively, perhaps you can mix in 50/50 the library that just worked, rather than using PhiX.

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            I assume that PhiX and input library is denatured separately and diluted with hyb buffer and then proportionally mixed. In this case it is less likely that any possible library contaminant would inhibit denaturation of PhiX due to dilution of contaminants. Once denatured library is loaded into flow cell it has to hybridize to flow cell oligos for bridge amplification and then hybridised with Read1 primer. At this step two events can result in sequencing failure:

            1- Any oligo with full or partial complementarity to flow cell oligo or its adapter compliment especially if end modified can block or disrupt bridge amplification resulting in clustering failure.

            2- The same goes with oligos with complimentary sequences to Read1 primer that can interrupt extension of primers during sequencing cycles.

            An alternative to oligos is carry over of a blocked nucleotide from library prep step which would have similar effects. A clean up as suggested by bilyl is good starting point for trouble shooting.

            Library prep with this kit follows a less efficient small RNAseq library prep with high number of PCR cycles in comparison with main stream kits. It also has a clean-up step after 3’ linker ligation which reduces the likelihood of 3’ linker carryover. I wonder if there is any trace of oligos in tape profile which can act as blocker during those two events.

            Comment

            • HESmith
              Senior Member
              • Oct 2009
              • 512

              #7
              The other possibility (assuming that you add phiX to the sample before denaturation) is that the sample is highly buffered or acidic, which would neutralize the NaOH used for denaturation. Alternatively, the sample pH could be basic and, after denaturation, exceed the recommended limit and adversely affect hybridization to the flow cell (although that would probably require a fairly high pH to block hybridization completely). In either case, cleanup of the sample would solve the problem.

              Comment

              • GW_OK
                Senior Member
                • Sep 2009
                • 411

                #8
                Thanks guys.

                It's supposedly in EB buffer, so I doubt it's an acidic or high buffer situation. Also, wouldn't stray oligos adversely affect qPCR?

                We do add phiX prior to denaturation. And I will be suggesting a bead cleanup to the customer once they return from vacation. There wasn't much to begin with and even less now after two loads so if anyone is going to be potentially losing the rest it will be them.

                Comment

                • bilyl
                  Member
                  • Aug 2013
                  • 52

                  #9
                  Originally posted by HESmith View Post
                  The other possibility (assuming that you add phiX to the sample before denaturation) is that the sample is highly buffered or acidic, which would neutralize the NaOH used for denaturation. Alternatively, the sample pH could be basic and, after denaturation, exceed the recommended limit and adversely affect hybridization to the flow cell (although that would probably require a fairly high pH to block hybridization completely). In either case, cleanup of the sample would solve the problem.
                  Somewhat relevant: I have seen the occasional improperly made buffers. Things like 10mM Tris Hcl pH 8.5 suddenly became 100mM because someone was tired when dispensing from stock. If that's the case then it would neutralize most of the 0.1N NaOH.

                  Comment

                  • nucacidhunter
                    Jafar Jabbari
                    • Jan 2013
                    • 1250

                    #10
                    Originally posted by HESmith View Post
                    The other possibility (assuming that you add phiX to the sample before denaturation) is that the sample is highly buffered or acidic, which would neutralize the NaOH used for denaturation. Alternatively, the sample pH could be basic and, after denaturation, exceed the recommended limit and adversely affect hybridization to the flow cell (although that would probably require a fairly high pH to block hybridization completely). In either case, cleanup of the sample would solve the problem.
                    If the library and PhiX were pooled prior to denaturation, HESmith’s explanation seems highly possible. An ssDNA assay after denaturation could be used to test this possibility.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                      by SEQadmin2




                      NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                      10-30-2026, 06:46 AM
                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-28-2026, 09:57 AM
                    0 responses
                    45 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-28-2026, 09:51 AM
                    0 responses
                    26 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    75 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    49 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...