    Index read Q30 problems on NextSeq

    Hello, I'm running a singal read dual indexed library on NextSeq, with a custom primer for index 2. While my general run metrics looked fine (clustering and pass filter rates ok) and my Read 1 Q30 was great (91%), both of my index reads crashed out with Q30 values in the 60s. Looking at the flowcell %>Q30 on SAV, once we hit the index reads most of the tiles start coming up grey. Obviously, we didn't get many reads assigned to samples, but I've never seen this before. Any ideas on why this happened, and how to remedy in the future?
    Thanks!
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    Four possible reasons:

    1- Chemistry or mechanical fault, more likely causing failure in stripping read1
    2- Sequence of index starting with GG
    3- Faults in adapter or primer design or synthesis where index primer does not bind properly, especially non-complementary 3’ base
    4- Interaction between Illumina and custom primers

    Comment

    • Vesperholly2
      Junior Member
      • Nov 2014
      • 4

      #3
      Hmmm, I wonder about faults in synthesis or chem/mech glitch. The adaptor/primer setup has always worked, so I'm not so curious about that, and we have many different indices in here that all tanked in unison. I wonder about how to best check that beyond simply running it again? If there was a glitch would it show up in the run logs, I wonder, or do those simply collect pass/fail data? No one likes wasting those kits....
      Thanks so much for your help!

      Comment

      • nucacidhunter
        Jafar Jabbari
        • Jan 2013
        • 1250

        #4
        Calling Illumina tech support can help at least to rule out some possibilities.

        Comment

        • GenoMax
          Senior Member
          • Feb 2008
          • 7142

          #5
          Get illumina tech support to take a look at this run. They can look at the instrument remotely (if your's is accessible, they will ask for certain log files) and determine if this problem is reagent/hardware related. In that case you will get a free replacement kit or other additional information to address the problem.

          Comment

          • HESmith
            Senior Member
            • Oct 2009
            • 512

            #6
            Another possibility is over-clustering. We've seen the same phenomenon (good reads, failed indices) at high cluster densities.

            All of nucacidhunter's explanations would yield index images with no clusters. Check the thumbnails to see if that's the case or not.

            Comment

            Latest Articles

            Collapse

            • SEQadmin2
              16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
              by SEQadmin2




              NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
              Yesterday, 06:46 AM
            • SEQadmin2
              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
              by SEQadmin2



              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
              ...
              10-14-2026, 09:55 AM

            ad_right_rmr

            Collapse

            News

            Collapse

            Topics Statistics Last Post
            Started by SEQadmin2, 10-28-2026, 09:57 AM
            0 responses
            29 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-28-2026, 09:51 AM
            0 responses
            18 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 11:13 AM
            0 responses
            71 views
            0 reactions
            Last Post SEQadmin2  
            Started by SEQadmin2, 10-22-2026, 10:48 AM
            0 responses
            48 views
            0 reactions
            Last Post SEQadmin2  
            Working...