    Too many "Undetermined Read" on HiSeq

    HiSeq 2500 Rapid run mode, PE 100bp, on-board cluster generation, dual index 8+8, DNA libraries prepared by Agilent QXT kit.

    We had several runs with high percentage undetermined read. Normally it is around 2~3% but in those runs they were 20~30%. The % Perfect Index Reads dropped from ~97% to 60~70%, and the % One Mismatch Reads (Index) increased from 2-3% to 30~40%. When we use the same sequencer to repeat the same sample, the repeat run data was good as normal.

    We carefully checked the index sequence of those "undetermined read", and found the index 1 sequences are consistent with some samples in the pool, but index 2 sequences have some mismatched bases.

    We don't know if the flowcell has defects or the PE cluster kit has problem. Dose anyone have the similar experience? Thanks a lot!
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    What was the cluster concentration? This sounds like the classic case of a flowcell that is overloaded (but not so much as to lose entire index 2 reads). Were you able to recover data allowing mismatches in the tag read?

    Comment

    • AndyCai
      Junior Member
      • Sep 2014
      • 2

      #3
      Originally posted by GenoMax View Post
      What was the cluster concentration? This sounds like the classic case of a flowcell that is overloaded (but not so much as to lose entire index 2 reads). Were you able to recover data allowing mismatches in the tag read?
      Actually the bad runs had low cluster density:
      Lane Tiles Density (K/mm2)
      1 64 949 +/- 61
      2 64 797 +/- 38
      1 64 1008 +/- 45
      2 64 1040 +/- 64
      1 64 905 +/- 63
      2 64 849 +/- 41
      1 64 864 +/- 34
      2 64 732 +/- 41
      1 64 972 +/- 49
      2 64 919 +/- 50

      The normal runs (good runs) are like this:
      1 64 1127 +/- 96
      2 64 1129 +/- 96
      1 64 1208 +/- 37
      2 64 1161 +/- 29
      1 64 1254 +/- 92
      2 64 1254 +/- 91
      1 64 1218 +/- 88
      2 64 1237 +/- 89
      1 64 1156 +/- 79
      2 64 1202 +/- 69

      The index read2 %Q30 decreased sharply to 20~60%, but read1, index read1, read2 looked fine (>85%), although slightly lower than normal (>90%).

      The cluster PF of bad runs are 84~87%, normal runs are 88~93%.

      We didn't try allowing mismatch index. We have to use perfect match reads only.

      Thanks!

      Comment

      • jdk787
        josh kinman
        • Apr 2014
        • 72

        #4
        Have you contacted Illumina about this?
        Like GenoMax stated, low quality index reads are normally a result of overclustering, so I would be interested in hearing what Illumina says about what you're seeing.
        Josh Kinman

        Comment

        Latest Articles

        Collapse

        • SEQadmin2
          16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
          by SEQadmin2




          NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
          Yesterday, 06:46 AM
        • SEQadmin2
          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
          by SEQadmin2



          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
          ...
          10-14-2026, 09:55 AM

        ad_right_rmr

        Collapse

        News

        Collapse

        Topics Statistics Last Post
        Started by SEQadmin2, 10-28-2026, 09:57 AM
        0 responses
        29 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-28-2026, 09:51 AM
        0 responses
        21 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 11:13 AM
        0 responses
        71 views
        0 reactions
        Last Post SEQadmin2  
        Started by SEQadmin2, 10-22-2026, 10:48 AM
        0 responses
        48 views
        0 reactions
        Last Post SEQadmin2  
        Working...