Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • d00b
    Member
    • Jan 2009
    • 17

    #1

    "Out of focus" on Miseq cluster

    Hi Miseq users,

    We have low Q score issue with Miseq V3 600 cycles kit.
    Cluster density was ~1000K, PhiX was ~5% aligned and PF was higher than 87.8%.

    Based on illumina spec, we supposed to have over 70% Q30 score. However, we have very low Q score (67% of Q30). We found this lower Q score was caused by cluster focus issue (we haven't got any error message during the run). We found tile image that contains out-of-focus clusters as following pictures.






    Is there anyone has experience on this?

    by the way, we have checked phiX run after on-site service. No problem with PhiX test run.
    Last edited by d00b; 03-03-2016, 10:48 AM.
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    What version of MCS are you running? If you are running v.2.6.x, the tech support will likely ask you to downgrade to MCS v.2.5.x. Have you contacted them?

    Comment

    • d00b
      Member
      • Jan 2009
      • 17

      #3
      Originally posted by GenoMax View Post
      What version of MCS are you running? If you are running v.2.6.x, the tech support will likely ask you to downgrade to MCS v.2.5.x. Have you contacted them?
      We have MCS 2.5.05.
      Yes, we have contacted them several times and we are still waiting for their review on this run. Do you have same issue with MCS V.2.6.x? Would you share your experience?

      Thanks

      Comment

      • GenoMax
        Senior Member
        • Feb 2008
        • 7142

        #4
        We were running v.2.6.x for a while on several machines and finally hit a "lost focus" error on one of the machines mid-run. At that point advice from tech support was to downgrade to v.2.5. After that downgrade things appear to be working ok (within limits, there is no solution for the Q-score drop for 600 cycle V3 kits).

        Comment

        • Jessica_L
          Senior Member
          • Feb 2010
          • 117

          #5
          I know several other users have reported issues with the v3 600 cycle kits, though I don't remember the specifics. I think at least some may have been related to the known MCS v.2.6 issue, but you might try searching around here for more information. I seem to remember several instances of people seeing low Q-scores on 600 cycle v3 kits, unrelated to the MCS v2.6 issue, as GenoMax has said.

          How are your 600 cycle runs normally? Do the images look blurry on the left side even early on or does it drift more out of focus as the run progresses? From an SAV/basespace Charts screen, what does the FWHM plot look like?

          Comment

          • d00b
            Member
            • Jan 2009
            • 17

            #6
            V3 600 cycles kit's Q score issue has been well noticed.
            However, our low Q score issue was also related with focusing problem.
            Our normal Q score with V3 600 is over 70% as illumina spec.
            Definitely, it is lower than V2 500 cycle kit. FWHM plot is terrible after 250cycles in read1 and 180cycles in read 2.

            Comment

            • GenoMax
              Senior Member
              • Feb 2008
              • 7142

              #7
              @d00b: The focusing issue sounds specific for your MiSeq. Something an engineer visit should fix.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                by SEQadmin2




                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                10-30-2026, 06:46 AM
              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, 10-28-2026, 09:57 AM
              0 responses
              45 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-28-2026, 09:51 AM
              0 responses
              26 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 11:13 AM
              0 responses
              74 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-22-2026, 10:48 AM
              0 responses
              48 views
              0 reactions
              Last Post SEQadmin2  
              Working...