    Library prep with <100ng DNA?

    What is the lowest amount of DNA I can get away with using for a library prep for a 36bp single read run on the IIX?
  • NextGenSeq
    Senior Member
    • Apr 2009
    • 482

    #2
    Nextera's protocol only uses 50ng. However, the library insert size is not very tight with it. You could probably use almost any protocol if you increase the amplification cycle number. This would also increase artifacts tho.

    Comment

    • sstarkenburg
      Junior Member
      • Aug 2010
      • 5

      #3
      Thanks for the tip. Have you had any success/experience using Illumina's library prep kit at 100ng? If so, how many amplification cycles would you recommend?

      Comment

      • henry.wood
        Member
        • Apr 2010
        • 63

        #4
        I've fairly routinely used around 50ng. I've occasionally dipped down lower. According to my picogreen machine, my lowest is 5ng. We tweaked the method a bit, doing 18 cycles of PCR before the gel cut rather than 12 cycles afterwards. It doesn't always work, and sometimes we get a bit of PCR duplicatation, but 50ng is reasonably well behaved.

        Comment

        • Efortier
          Junior Member
          • Nov 2009
          • 8

          #5
          We routinely use 10 ng and 12 cycles without any problems. Just be careful to reduce the amount of adapters and PCR primers to take into consideration the lower amount of initial DNA.

          Comment

          • suludana
            Member
            • May 2008
            • 61

            #6
            Are you talking about the preparation of Genomic DNA?

            Comment

            • henry.wood
              Member
              • Apr 2010
              • 63

              #7
              I'm talking about genomic DNA, mostly for copy number purposes. I've put all the method details in a recent paper http://nar.oxfordjournals.org/cgi/co...&pmid=20525786

              Comment

              • sstarkenburg
                Junior Member
                • Aug 2010
                • 5

                #8
                Many thanks to everyone for the input! Happy sequencing....

                Comment

                • sstarkenburg
                  Junior Member
                  • Aug 2010
                  • 5

                  #9
                  Originally posted by Efortier View Post
                  We routinely use 10 ng and 12 cycles without any problems. Just be careful to reduce the amount of adapters and PCR primers to take into consideration the lower amount of initial DNA.
                  How are your purifying the prep after ligating the adapters?

                  Comment

                  • mimi_lupton
                    Member
                    • Oct 2008
                    • 19

                    #10
                    Originally posted by henry.wood View Post
                    I'm talking about genomic DNA, mostly for copy number purposes. I've put all the method details in a recent paper http://nar.oxfordjournals.org/cgi/co...&pmid=20525786
                    I am doing some sequencing to look at copy number variants at low depth. Can I ask why in your paper you use 200bp fragment size, did you use paired end reads? Would it have been possible to have a larger fragment size to allow you to call larger inversions for example?

                    Thanks

                    Comment

                    • henry.wood
                      Member
                      • Apr 2010
                      • 63

                      #11
                      Originally posted by mimi_lupton View Post
                      I am doing some sequencing to look at copy number variants at low depth. Can I ask why in your paper you use 200bp fragment size, did you use paired end reads? Would it have been possible to have a larger fragment size to allow you to call larger inversions for example?

                      Thanks
                      Hello,
                      The 200bp fragment was partly due to the samples. DNA from FFPE tissues is often already fairly fragmented, so getting 500bp fragments might have been tricky. It's also partly due to convenience, that's what was already happening in the lab, so we didn't have to play around with the covaris too much.
                      We only used single end reads, again partly due to the fragmentation and partly because the sequencer we use mostly does single end runs. Purely from a copy number point of view, as long as you have enough sequence to trust the alignment for each read, then that's fine. To spot inversions etc using paired end reads we would have needed higher coverage. At the depth we used, we're only getting a read every 3kb or so. That's fine for copy number, and we keep the libraries, so we can always go back and resequence a sample if we want greater resolution.
                      Hope that makes sense.

                      Comment

                      • maggie
                        Junior Member
                        • Dec 2009
                        • 5

                        #12
                        I understand using less adaptor for lower input amounts during Illumina library construction and then altering the number of PCR cycles. Why/how should the PCR primer amounts be adjusted (how much should be used for 50ng input, or 200ng input, etc)?

                        Comment

                        • zwirko
                          Junior Member
                          • Feb 2011
                          • 3

                          #13
                          I'm trying to automate the Nextera prep, and I'm wondering how to go about the tagmentation cleanup. Yields following Ampure (SPRI) have been substantially lower than with MinElute columns, but columns can't easily be scaled up and the elution volumes on Qiagen plates are way too large not to mention highly variable. Ideally, I'd like to eliminate the tagmentation cleanup all together, but I'm not sure how since simply skipping it doesn't seem to work. Ideas?

                          Comment

                          • HoneyBee
                            Junior Member
                            • Jun 2012
                            • 4

                            #14
                            Originally posted by henry.wood View Post
                            Hello,
                            The 200bp fragment was partly due to the samples. DNA from FFPE tissues is often already fairly fragmented, so getting 500bp fragments might have been tricky. It's also partly due to convenience, that's what was already happening in the lab, so we didn't have to play around with the covaris too much.
                            We only used single end reads, again partly due to the fragmentation and partly because the sequencer we use mostly does single end runs. Purely from a copy number point of view, as long as you have enough sequence to trust the alignment for each read, then that's fine. To spot inversions etc using paired end reads we would have needed higher coverage. At the depth we used, we're only getting a read every 3kb or so. That's fine for copy number, and we keep the libraries, so we can always go back and resequence a sample if we want greater resolution.
                            Hope that makes sense.
                            Have you tried exome sequencing on FFPE samples? We are thinking of using TruSeq Exome kit but don't know if it works with low input + FFPE samples..

                            Comment

                            • henry.wood
                              Member
                              • Apr 2010
                              • 63

                              #15
                              We are currently awaiting the results of FFPE exomes. We used the illumina nextera exome kits. They worked with <100ng DNA from FFPE samples. The libraries looked fine, but we won't know how the data turns out for a week or so.
                              I know people have used sureselect kits with FFPE in the past, but I think they need higher amounts of template. I don't know about truseq.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM
                              • SEQadmin2
                                New Genomics Technologies Take Aim at Long-Standing Limits
                                by SEQadmin2


                                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                                ...
                                09-28-2026, 10:25 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, Yesterday, 10:52 AM
                              0 responses
                              11 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-14-2026, 05:39 AM
                              0 responses
                              19 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-13-2026, 09:56 AM
                              0 responses
                              18 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-09-2026, 11:14 AM
                              0 responses
                              27 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...