    Hi All,

    Anyone is facing instability of rRNA removal using ribozero kit of Illumina? Thanks.
  • dpryan
    Devon Ryan
    • Jul 2011
    • 3478

    #2
    If by "instability" you mean that the you often get a relatively large percentage of rRNA sequenced after using the kit, then yes, we see that all the time. We've tried a bunch of different kits and the results are hit or miss. I think we only get good results if the kits are used more or less immediately after RNA extraction.

    Comment

    • pmiguel
      Senior Member
      • Aug 2008
      • 2328

      #3
      It usually works okay for us. The biggest consistent failure mode is that if you don't use a sample with enough rRNA, the rRNA capture probe oligos can bleed over into your library. You can tell if they are the source of your "rRNA" contamination because they will map to the wrong strand of your rRNA reference.

      So we insist people bring us at least 1ug of total RNA when using this kit.

      --
      Phillip

      Comment

      • dpryan
        Devon Ryan
        • Jul 2011
        • 3478

        #4
        @pmiguel: I hadn't thought about checking the strand, that's a good idea! What sort of left-over rRNA levels do you tend to get? Ours range from low (<0.5% rRNA) to absurdly high (up to 20% or so).

        Comment

        • pmiguel
          Senior Member
          • Aug 2008
          • 2328

          #5
          Originally posted by dpryan View Post
          @pmiguel: I hadn't thought about checking the strand, that's a good idea! What sort of left-over rRNA levels do you tend to get? Ours range from low (<0.5% rRNA) to absurdly high (up to 20% or so).
          As long as they are below 30% or so I call it a win. (Some plant tissues produce total RNA that seems to be >95%rRNA). But I think we probably are getting similar results to you most of the time.

          The samples that showed the wrong strand hits produced >90% rRNA reads.

          Oh, the other "trick" -- only buy the Ribozero "Gold" library kit. This is for "human, mouse, rat" but we've used it for other animals, including moths. If you get the non-Gold kit, there are no probes to the mitochondrial rRNA. Fine for some tissue--but disastrous for others. (Eg, muscle tissue.)

          Fortunately the plant Ribozero library construction kit includes probes to nuclear, mitochondrial and chloroplast rRNA.

          --
          Phillip

          Comment

          • dpryan
            Devon Ryan
            • Jul 2011
            • 3478

            #6
            Our sequencing folks were getting worried whenever the numbers were greater than the 2% advertised levels. Glad it's not just us (not that I'm surprised, the folks in our sequencing facility are really good).

            Comment

            • maggielaw
              Junior Member
              • Aug 2010
              • 3

              #7
              Originally posted by pmiguel View Post
              It usually works okay for us. The biggest consistent failure mode is that if you don't use a sample with enough rRNA, the rRNA capture probe oligos can bleed over into your library. You can tell if they are the source of your "rRNA" contamination because they will map to the wrong strand of your rRNA reference.

              So we insist people bring us at least 1ug of total RNA when using this kit.

              --
              Phillip
              Hi Philip,

              Which kits are you referring to? TruSeq Stranded Total RNA with Ribo-Zero Human/Mouse/Rat (Cat# RS-122-2201)? TruSeq Stranded Total RNA with Ribo-Zero Globin(RS-122-2501)? or others?

              Comment

              • pmiguel
                Senior Member
                • Aug 2008
                • 2328

                #8
                Originally posted by maggielaw View Post
                Hi Philip,

                Which kits are you referring to? TruSeq Stranded Total RNA with Ribo-Zero Human/Mouse/Rat (Cat# RS-122-2201)? TruSeq Stranded Total RNA with Ribo-Zero Globin(RS-122-2501)? or others?
                I guess this would be an issue for all these kits. But the one we use most frequently is:

                TruSeq Stranded Total RNA with Ribo-Zero Gold, Set A Indexes (Cat# RS-122-2301)

                and/or

                TruSeq Stranded Total RNA with Ribo-Zero Gold, Set B Indexes (Cat# RS-122-2302)

                --
                Phillip

                Comment

                • subxero
                  Member
                  • Jul 2013
                  • 16

                  #9
                  Originally posted by pmiguel View Post
                  It usually works okay for us. The biggest consistent failure mode is that if you don't use a sample with enough rRNA, the rRNA capture probe oligos can bleed over into your library. You can tell if they are the source of your "rRNA" contamination because they will map to the wrong strand of your rRNA reference.

                  So we insist people bring us at least 1ug of total RNA when using this kit.

                  --
                  Phillip
                  Is this the only solution you tried, increasing the input?

                  We are having this problem right now but we have very limited amounts of RNA so are hands are a bit tied as far as the amount of input RNA we can use.

                  Is it possible to decrease the amount of probes used or something along those lines?

                  Comment

                  • pmiguel
                    Senior Member
                    • Aug 2008
                    • 2328

                    #10
                    Actually, I re-thought this recently. The ribo-zero probes (biotinylated oligonucleotides) would not be removed more effectively if they are double stranded -- the streptavidin beads are what removes them--duplex or not.

                    Also, the cDNA produced by 1st strand synthesis, would be the same strand as the ribozero probes. So telling the two apart would require the ability to distinguish human from mouse and rat (or whatever species your RNA doesn't come from) to be sure it is ribozero probes that are contaminating your sample.

                    I was thinking this morning that what we needed was a few sets of rRNA primers so one could assay RNAseq libraries for rRNA levels directly with qPCR prior to sequencing.

                    --
                    Phillip

                    Comment

                    • chenyudr
                      Junior Member
                      • Feb 2016
                      • 9

                      #11
                      Originally posted by pmiguel View Post
                      Actually, I re-thought this recently. The ribo-zero probes (biotinylated oligonucleotides) would not be removed more effectively if they are double stranded -- the streptavidin beads are what removes them--duplex or not.

                      Also, the cDNA produced by 1st strand synthesis, would be the same strand as the ribozero probes. So telling the two apart would require the ability to distinguish human from mouse and rat (or whatever species your RNA doesn't come from) to be sure it is ribozero probes that are contaminating your sample.

                      I was thinking this morning that what we needed was a few sets of rRNA primers so one could assay RNAseq libraries for rRNA levels directly with qPCR prior to sequencing.

                      --
                      Phillip
                      I think the mentioned "a few sets of rRNA primers" are necessary tools for checking rRNA removal efficiency.

                      Comment

                      • dwong
                        Junior Member
                        • Jun 2016
                        • 5

                        #12
                        Have you guys ever considered using NuGEN's InDA-C customized depletion? They also have standard depletion kits for human, Drosophila and mouse. The depletion is only performed after the library is actually made. Just a thought.

                        Comment

                        • pmiguel
                          Senior Member
                          • Aug 2008
                          • 2328

                          #13
                          Originally posted by dwong View Post
                          Have you guys ever considered using NuGEN's InDA-C customized depletion? They also have standard depletion kits for human, Drosophila and mouse. The depletion is only performed after the library is actually made. Just a thought.
                          Yes, I've thought about it. We've actually had a customer that used it to deplete a complex environmental (insect gut flora) RNA sample set of prokaryotic and eukaryotic rRNA.

                          --
                          Phillip

                          Comment

                          • dwong
                            Junior Member
                            • Jun 2016
                            • 5

                            #14
                            I used their Drosophila kit and it knocked down the rRNA quite a bit. Its great that its specific to the organism of choice, but that's just my personal opinion

                            Darren

                            Comment

                            • kmcarr
                              Senior Member
                              • May 2008
                              • 1181

                              #15
                              Originally posted by dwong View Post
                              Have you guys ever considered using NuGEN's InDA-C customized depletion? They also have standard depletion kits for human, Drosophila and mouse. The depletion is only performed after the library is actually made. Just a thought.
                              NuGEN will also work with you to create a custom set of oligos to use with this kit. These can be either for rRNA from a non-model organism or any other RNA you would like to deplete from your library (e.g. globin transcripts in blood). Of course you need to know the sequence of the transcripts you wish to design oligos for.

                              A client of ours had NuGEN add some custom oligos for some specific targets they wanted depleted in a host/pathogen mix which reportedly worked well.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              45 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              26 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              75 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              49 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...