    Erratic cluster numbers

    Hi,

    Since ~ 2 months we are getting erratic cluster numbers on our GAs (cBot) even if we check the libraries with RT-PCR. Before we were using 7pM (~300.000clusters /tile) and were getting reproducible, nice results. Now a 7pM sample might result in more than 900.000/mm2 clusters. Also single lanes might totally fail without obvious reasons (not even for techsupport).
    I just saw a recent similar thread, but wanted to ask more specific if other/more people are experiencing the same problems.

    Thanks
    Andi
  • GW_OK
    Senior Member
    • Sep 2009
    • 411

    #2
    If you've updated your software that could be the issue. The new software uses a different cluster finding algorithm that's better at identifying clusters than the old one. Our FAS said we ought to re-titrate all our libraries, but since we don't titrate them anyway it's a bit of a moot point. We run ~10-10.5 pM with pretty good results.

    For the lane failures, is it always the same lane?

    Comment

    • NextGenSeq
      Senior Member
      • Apr 2009
      • 482

      #3
      If you spike in PhiX control at 1% or less, it will tell you if its the library or a technical amplification issue.

      Comment

      • Efortier
        Junior Member
        • Nov 2009
        • 8

        #4
        For the lanes that totally failed for no apparent reasons, we had similar problems a few months ago. The problem was the cBot plates, more precisely the reagent HP5 (0.1N NaOH). For some reason during shipping on dry ice, the NaOH pH changed due to the high amount of CO2 produced by the dry ice. This resulted in bad primer hybridisation on the flowcell because the NaOH was not at pH above 12. To work around this problem we just made our own 0.1N NaOH and replaced the HP5 with our NaOH. The new cBot plates should have red foil sealing the HP5 instead of the silicon-type tubes cap. This should fix this issue I think. You can also recover a bad primer hybridisation by redoing the primer hybridisation on the cBot with the appropriate reagents.

        Hope this helps.

        Comment

        • huhrik
          Junior Member
          • Mar 2008
          • 1

          #5
          have you checked raw images for failed lanes?
          how do they looks?

          Comment

          • andibody
            Member
            • Mar 2008
            • 45

            #6
            Indeed, it started happening after we upgraded the software. Could it be that the new image analysis resolves more clusters but at the same time is less unforgivable to high cluster density?
            @NextGenSeq: if the software can't resolve the images it won't also be able to do so for PhiX clusters.
            @huhrik: they are indeed denser than 'good' lanes but far from being 'fully covered'

            Comment

            • simonandrews
              Simon Andrews
              • May 2009
              • 870

              #7
              By any chance are you using libraries which have very biased sequence composition at the start of the first read? This can cause problems with cluster detection which can look like you're getting low cluster numbers.

              Comment

              • andibody
                Member
                • Mar 2008
                • 45

                #8
                no, the density problems were observed even on whole genome samples.

                Comment

                Latest Articles

                Collapse

                • SEQadmin2
                  16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                  by SEQadmin2




                  NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                  10-30-2026, 06:46 AM
                • SEQadmin2
                  Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                  by SEQadmin2



                  Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                  Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                  ...
                  10-14-2026, 09:55 AM

                ad_right_rmr

                Collapse

                News

                Collapse

                Topics Statistics Last Post
                Started by SEQadmin2, 10-28-2026, 09:57 AM
                0 responses
                33 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-28-2026, 09:51 AM
                0 responses
                21 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 11:13 AM
                0 responses
                72 views
                0 reactions
                Last Post SEQadmin2  
                Started by SEQadmin2, 10-22-2026, 10:48 AM
                0 responses
                48 views
                0 reactions
                Last Post SEQadmin2  
                Working...