  • dlbuz
    Member
    • May 2016
    • 10

    #1

    Using TruSeq and Small RNA libraries in same sequencing reaction

    I was wondering if it were possible to combine Small RNA libraries prepared using the Illumina Small RNA kit and mRNA libraries prepared using the TruSeq stranded mRNA kit into the same sequencing reaction?

    I noticed that the PCR primer (RP1) used in the small RNA kit is similar up to 20ish nt from the 5' end of the TruSeq Universal Adapter.

    RP1 (RNA PCR Primer) | After this point it differs from TruSeq Universal Adapter
    5’ AATGATACGGCGACCACCGAGATCTACAC|GTTCAGAGTTCTACAGTCCGA

    RPIX (Small RNA Index Primers)
    5’ CAAGCAGAAGACGGCATACGAGATXXXXXXGTGACTGGAGTTCCTTGGCACCCGAGAATTCCA

    vs.


    TruSeq Universal Adapter
    5’ AATGATACGGCGACCACCGAGATCTACAC|TCTTTCCCTACACGACGCTCTTCCGATCT

    TruSeq Index Adapters
    5’ GATCGGAAGAGCACACGTCTGAACTCCAGTCACXXXXXXATCTCGTATGCCGTCTTCTGCTTG


    Although the Small RNA and TruSeq individual adapters vary widely beginning from the 5' end.

    I'm asking because I called Illumina support and they told me that they don't recommend it but didn't specifically say it would be incompatible. Knowing them they would want me to run 2 sequencing reactions separately.
    Last edited by dlbuz; 07-29-2016, 10:36 AM.
  • cmbetts
    Senior Member
    • Jun 2012
    • 120

    #2
    It's probably fine on a MiSeq or NextSeq, as they automatically use all of the different sequencing primers for the different sample prep methods. I haven't done it with small RNA and TruSeq before, but we regularly mix TruSeq and Nextera libraries (And hybrids that have TruSeq on one end and Nextera on the other) on our MiSeqs/NextSeqs without any apparent issue.
    I think that hardest part would be getting the cluster density right, as the different sized libraries have different clustering/loading characteristics. You also probably be wasting a lot of cycles on the smRNA library because of the read lengths and especially if you were planning on doing paired end.

    Comment

    • dlbuz
      Member
      • May 2016
      • 10

      #3
      I'm using MiSeq instrument and I was also under the impression that the instrument uses all the primers compatible with MiSeq.

      One way I accounted for possible wasting cycles was to dilute the small RNA samples 10 fold as the coverage isn't as necessary as it is with the mRNA samples.

      I was wondering if your reads came up normally with the illumina software platform when you used two different adapters or did you manually fish out the reads from the sequencing instrument?

      Comment

      • cmbetts
        Senior Member
        • Jun 2012
        • 120

        #4
        Originally posted by dlbuz View Post
        I'm using MiSeq instrument and I was also under the impression that the instrument uses all the primers compatible with MiSeq.

        One way I accounted for possible wasting cycles was to dilute the small RNA samples 10 fold as the coverage isn't as necessary as it is with the mRNA samples.

        I was wondering if your reads came up normally with the illumina software platform when you used two different adapters or did you manually fish out the reads from the sequencing instrument?
        The libraries will properly demultiplex as long as the sample sheet has the correct indices for the libraries. The requires manually editing the sample sheet because you can't select indices from multiple kits in Illumina Sample Manager. You would follow a similar procedure as if you've made your own custom indices.

        Comment

        • cmbetts
          Senior Member
          • Jun 2012
          • 120

          #5
          I should also mention that when adding indices manually, there aren't any checks for correct color balancing, so it's up to you to make sure that they're compatible in low-plex situations.

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM
          • SEQadmin2
            New Genomics Technologies Take Aim at Long-Standing Limits
            by SEQadmin2


            Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

            We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
            ...
            09-28-2026, 10:25 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-15-2026, 10:52 AM
          0 responses
          23 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-14-2026, 05:39 AM
          0 responses
          27 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-13-2026, 09:56 AM
          0 responses
          25 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-09-2026, 11:14 AM
          0 responses
          28 views
          0 reactions
          Last Post SEQadmin2  
          Working...