  • Lovro
    Member
    • Mar 2015
    • 19

    #1

    myseq error "waiting for FPGA command CAMDBLSNAP"

    Hi,

    I would appreciate any help in regard to the following MySeq error:

    Timeout of 70.000 seconds exceeded while waiting for FPGA command "CAMDBLSNAP 390 390 188 188 34390 150" to complete.

    see image:



    The error emerges any time after the run is started and ruins the run. It first appeared a year ago and now the fifth time since.

    The only advice we are continually getting from Illumina support is that we should reboot our myseq prior to the run (we are doing that since the first time the error occurred) and that the error is "very unlikely to repeat".
    Unfortunately, the error is repeating and keeps ruining runs.

    The MySeq's warranty has expired.

    If anybody had experienced the problem himself or has any information regarding the error (what does the code mean?), we would be very happy to hear from you.

    Thank you,
    Lovro
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    If your MiSeq is not under a maintenance contract then you are living dangerously (or you just don't use it enough to justify the contract).

    This error may be due to some sort of hardware issue that can only be fixed by a visit from an Illumina engineer. You may have to bite the bullet and pay for that visit.

    Comment

    • Jessica_L
      Senior Member
      • Feb 2010
      • 117

      #3
      What Genomax said.

      I'd add that if you don't have your MiSeq under a maintenance agreement, you may want to consider getting one first, as the service visit and repair can easily exceed the cost of a contract. There are usually some hoops to jump through, but that's probably the cheapest way to get your instrument fixed.

      An initial visit with an FSE might be a good idea, at least, and if you have a decent relationship with them, they could give you a ballpark idea of how much repairs might cost and how that would compare to the cost of a service agreement.

      Comment

      • Lovro
        Member
        • Mar 2015
        • 19

        #4
        Hi, thanks for the answers.
        We have considered the latter.
        However, I was more interested in finding out the meaning of the error code.

        We had a hardware problem with a moveable mirror part (bouncing while snapping into position) while the machine was still under warranty. It was replaced by illumina and might still be under warranty.
        "CAMDBLSNAP" sounds a lot like some part of the camera snapping into position, so I want to investigate the issue.

        The error codes descriptions might be impossible to get a hold of though...?

        tnx

        Comment

        • Jessica_L
          Senior Member
          • Feb 2010
          • 117

          #5
          Your best bet on that score is probably your field support from Illumina, for a MiSeq probably your FAS. The error codes aren't closely guarded secrets, but I'm not aware of a collection of codes with descriptions that's accessible in any kind of searchable format.

          Comment

          • misterc
            Member
            • Jan 2016
            • 46

            #6
            CAMDBLSNAP is an FPGA or cabling issue on one of the 2 USB cables to the cameras internally. You can try re-seating the USB cables to the cameras to prevent/reduce the frequency of these errors. The ones you want to re-seat are on the inside of the MiSeq, facing the front of the instrument, up on the top right (behind the monitor), readily accessible when you take the top skin/panel off.

            Comment

            • Lovro
              Member
              • Mar 2015
              • 19

              #7
              Thank you very much. I will mention this to their representative.

              Comment

              Latest Articles

              Collapse

              • SEQadmin2
                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                by SEQadmin2



                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                ...
                10-14-2026, 09:55 AM
              • SEQadmin2
                New Genomics Technologies Take Aim at Long-Standing Limits
                by SEQadmin2


                Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                ...
                09-28-2026, 10:25 AM

              ad_right_rmr

              Collapse

              News

              Collapse

              Topics Statistics Last Post
              Started by SEQadmin2, Yesterday, 10:52 AM
              0 responses
              12 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-14-2026, 05:39 AM
              0 responses
              19 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-13-2026, 09:56 AM
              0 responses
              19 views
              0 reactions
              Last Post SEQadmin2  
              Started by SEQadmin2, 10-09-2026, 11:14 AM
              0 responses
              27 views
              0 reactions
              Last Post SEQadmin2  
              Working...