I have focusing problems with 1.4mm FC. Sometimes a couple of lanes will be totally off focus while the rest of the lanes seems to be within the 15000nm range; Sometimes one end is +10,000nm and the other end gives -10,000nm and this makes it impossible to set a good "Z" value for focusing across the full FC. I think part of the problem is caused by the container FC was stored/shipped because often FC got stuck in it and this may have squeezed and forced the FC to change shape since it's a very thin sandwich. Anybody having similar issues in the field? Any solutions or suggestions? Thanks!
Unconfigured Ad
Collapse
X
-
Not every FC is manufactured perfectly....generally speacking, as long as you have a good focus in the middle of FC, lane 4-1,2 or 5-1,2, you may be fine.Originally posted by mh123 View Post........and this makes it impossible to set a good "Z" value for focusing across the full FC.........
If way off, it seems your case, chances are plate that holds your FC in your GAII may be off-leveled. Illumina engineer has a tool to check and adjust the level. top left, right and bottom left and right should be leveled for even focusing. This off-level may cause increase error rate as it will affect imaging quality. You may want to check with FAS and/or service engineer.Last edited by monad; 11-21-2008, 09:31 AM.
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...-
Channel: Articles
Today, 06:46 AM -
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, 10-28-2026, 09:57 AM
|
0 responses
23 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:57 AM
|
||
|
Started by SEQadmin2, 10-28-2026, 09:51 AM
|
0 responses
15 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
68 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
48 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Comment