  • soleulloa
    Member
    • Aug 2011
    • 34

    #1

    Discordance between Cluster Density and Thumbnail images

    Hi all,

    I ran MiSeq using a 500 cycles kit.
    The value for cluster density was 310K/mm2 but if I'm going to look the thumbnail images, I think I have much more cluster than that.

    Illumina tech looked the results and says "you are overclusterized".
    But for me cluster density is OK when I've reviewed the images.

    I'm attaching here 4 images from cycle 1.

    Could you tell me your opinion?
    Thank you!!
    Soledad
    Attached Files
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    To me reported cluster density is what I would expect from the images. The run has been underclustered for a library that looks high diversity although I am assuming that the base diversity of other cycles are similar to the first cycle.

    Comment

    • Brian Bushnell
      Super Moderator
      • Jan 2014
      • 2709

      #3
      What's the scale on the little yellow boxes?

      Comment

      • misterc
        Member
        • Jan 2016
        • 46

        #4
        That looks about right to me for 300-500K/mm2 raw density (and NOT overclustered). Those intensities look reasonable to me too for cycle 1 given the MiSeq ramps up exposure on the cameras at each cycle to counter the typical decreasing intensity plots on their instruments.

        Comment

        • soleulloa
          Member
          • Aug 2011
          • 34

          #5
          Ok! Thank you.
          I'm using 10 pM (from 4 nM) to get this result.
          What do you think I have to use now? 12 or 15 pM to get a good run?

          Comment

          • microgirl123
            Senior Member
            • Jun 2012
            • 199

            #6
            What's your percentage passing filter?

            Comment

            • soleulloa
              Member
              • Aug 2011
              • 34

              #7
              Hi,

              95.27% PF

              Comment

              • microgirl123
                Senior Member
                • Jun 2012
                • 199

                #8
                If 95% of them are passing filter, then you are definitely not overclustered. If you were overclustered your percentage passing filter would be low.

                Unfortunately, the relationship between loading concentration and clusters is not linear so it's always a guess. I find that getting good cluster density is the Achilles' heel of the Illumina Miseq. My guess would be to go with the 15 pM since your density was so low.

                Comment

                • pmiguel
                  Senior Member
                  • Aug 2008
                  • 2328

                  #9
                  Originally posted by Brian Bushnell View Post
                  What's the scale on the little yellow boxes?
                  I don't know, but when I looked into it in the past, it seemed that the size of a cluster was around 1um in diameter. But that may have been on a HiSeq. I'm not sure whether the scales would be different on a MiSeq.

                  Old-timers like @ECO would be more likely to know. (Alas, I don't think @ callouts actually work on SeqAnswers...)

                  --
                  Phillip

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    Illumina has a note on optimizing cluster density available here.

                    @soledad: Have you checked what the images look like in the middle and at the end of run?
                    Last edited by GenoMax; 02-17-2017, 08:51 AM.

                    Comment

                    • soleulloa
                      Member
                      • Aug 2011
                      • 34

                      #11
                      Someone has quantified libraries before dilution? When you prepare 2 nM dilutions?
                      I've quantified before and after that final dilution now because this is the first run when I have this problem and I'd like to compare these results.
                      Always I was between 1000-1200 K/mm2 using a 300 cycles kit.
                      This time I did the same process... and the only variable is I'm using 500 cycles kit.
                      But the library concentration will be the same.. so I don't understand what happens.

                      Comment

                      • nucacidhunter
                        Jafar Jabbari
                        • Jan 2013
                        • 1250

                        #12
                        It is most likely overestimating library concentration if you have used the same type of library. qPCR is the best method for quantification, other methods that quantify dsDNA are inadequate for libraries that contain non-sequenceable dsDNA fragments.

                        Comment

                        Latest Articles

                        Collapse

                        • SEQadmin2
                          Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                          by SEQadmin2



                          Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                          Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                          ...
                          10-14-2026, 09:55 AM
                        • SEQadmin2
                          New Genomics Technologies Take Aim at Long-Standing Limits
                          by SEQadmin2


                          Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                          We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                          ...
                          09-28-2026, 10:25 AM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by SEQadmin2, 10-15-2026, 10:52 AM
                        0 responses
                        16 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 10-14-2026, 05:39 AM
                        0 responses
                        21 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 10-13-2026, 09:56 AM
                        0 responses
                        22 views
                        0 reactions
                        Last Post SEQadmin2  
                        Started by SEQadmin2, 10-09-2026, 11:14 AM
                        0 responses
                        27 views
                        0 reactions
                        Last Post SEQadmin2  
                        Working...