Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • MisUser
    Member
    • Jan 2017
    • 13

    #1

    Low diversity on Nextseq 500

    Hi all,

    Does anyone have already sequenced low diversity libraries on the Nextseq and how did it perform? What were the conditions (PhiX % and library quantities)?
    How many reads did you get?
    Thank you for your input!
    Last edited by MisUser; 03-29-2017, 07:04 AM.
  • jteeee2
    Member
    • Oct 2014
    • 42

    #2
    What kind of libraries are they? Are you expecting diversity to be low throughout the entire read length or just in specific parts?

    We have had success with certain types of low diversity libraries on the NextSeq. Specific performance will depend on the exact library type, but a good place to start would be a PhiX spike-in of 10-20%.

    -JT

    Comment

    • MisUser
      Member
      • Jan 2017
      • 13

      #3
      Thanks for your quick reply.
      The sequence is identical for the first 25 nts and the diversity is good for the next 20nts.

      Comment

      • jteeee2
        Member
        • Oct 2014
        • 42

        #4
        We haven't dealt with this specific situation before, but with that length of identical sequence I would probably aim a little higher with the PhiX for the first attempt (20-30%). Would be interested to hear if other folks have sequenced this kind of library on the NextSeq. Once you convince yourself you can get good data from the high PhiX spike-in, you can titrate it down on future runs so you don't waste so many reads.

        Comment

        • cmbetts
          Senior Member
          • Jun 2012
          • 120

          #5
          I think that 25% is a pretty good starting number. I used to regularly sequence libraries that looked like (8bp diversity + 15bp linker) x 3 on a NextSeq, and things tended to go badly when we dropped the PhiX below 20%

          Comment

          • nucacidhunter
            Jafar Jabbari
            • Jan 2013
            • 1250

            #6
            NextSeq seems to perform good with low diversity and unbalanced libraries such as RRBS as shown in the following:

            Comment

            • dlin82
              Junior Member
              • May 2017
              • 2

              #7
              Hi - sorry to revive an old thread, we are gearing up our own very first RRBS run on the NextSeq500 so wanted to check in if 30% is actually a good amount of PhiX to spike in. Are the quality of clustering consistent?

              Consulting with Illumina and other sources they suggest a starting amount of 40-50% (more like 50%) PhiX, and you really do lose quite a lot of reads with that amount so I would like to avoid that at all if possible. Thanks so much everyone!

              Comment

              • nucacidhunter
                Jafar Jabbari
                • Jan 2013
                • 1250

                #8
                Two issues with RRBS sequencing:
                1- unbalanced base composition: NextSeq performs fine
                2- low diversity at the start of reads: depends on method and if you are using NuGEN kit 1% PhiX would fine but if you are doing end repair, A tailing and ligation based method then 20% PhiX spike in is good starting point

                Comment

                • dlin82
                  Junior Member
                  • May 2017
                  • 2

                  #9
                  Thanks for the quick reply nucacidhunter!

                  Yes it's homebrew so I guess I'll start with 20-30% PhiX spike-in. I am reading the Meissner mRRBS paper more in detail and noticed they do dark cycles. Does it help in this case (or can the Nextseq even do this)?

                  Thanks so much again, I am basically thrown into this NGS business with zero previous training and am still trying to figure things out!

                  Comment

                  • nucacidhunter
                    Jafar Jabbari
                    • Jan 2013
                    • 1250

                    #10
                    For dark cycles you would need Illumina's assistance in your own risk if they agree and generally is worthwhile if you are doing multiple runs. I have not heard of dark cycles on NextSeq.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                      by SEQadmin2




                      NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                      Yesterday, 06:46 AM
                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-28-2026, 09:57 AM
                    0 responses
                    29 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-28-2026, 09:51 AM
                    0 responses
                    21 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    71 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    48 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...