Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • frossit
    Member
    • Jun 2011
    • 13

    #1

    Focus error in Miseq

    We started a MiSeq running and after the first cycle there was an error: "Best focus is too near edge of range: AutoFocus would have moved z to 0, outside soft limits os -0,295001144426642... -0,00499885557335775: ThroughFocusSacnCommand".
    Is it a camera problem?
    May I start the running againg?

    Thanks in advance for your attention!
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    Contact Illumina tech support and let them take a look at your sequencer.

    Comment

    • thermophile
      Senior Member
      • Apr 2015
      • 243

      #3
      There's no restarting a run with focus problems. It's unlikely to be a camera problem, more likely cluster failure, lens or z stage issue. Absolutely call tech support
      Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

      Comment

      • Samm
        Junior Member
        • Mar 2018
        • 1

        #4
        We have had this error several times as well. The run just stops without images to show whether there is under- or over-clustering. I am very doubtful if this is just a clustering issue. I think it is a hardware problem.

        Comment

        • chuanwu
          Junior Member
          • May 2018
          • 8

          #5
          I had the same issue twice in a row at Miseq platform. How did you guys resolve the problem. I called Illumina and they just questioned a lot about my library.

          Thanks

          Comment

          • ikripp
            Member
            • Jan 2018
            • 19

            #6
            I had the same issue - "Best focus too near edge of range" and after contacting tech support and being directed to run these tests:

            Y Stage test
            M3 mirror test
            Z Stage test
            Compensator test
            Conduct volume test
            Thermal Ramp test
            Prime Reagent Lines
            Optics test

            I was told the run was significantly overclustered and to drop my loading concentration and reload.

            Comment

            • chuanwu
              Junior Member
              • May 2018
              • 8

              #7
              All the system checks passed.
              The only issue for me why the quant of library measured by KAPA qPCR, Qubit, BioAnalyzer and LabChip GX are not consistent with each other. Which platform is reliable?
              Thanks

              Comment

              • gabrieltw
                Member
                • Mar 2013
                • 24

                #8
                KAPA is the most accurate, provided you calculate it correctly. We usually run librarys on BioA or AA to get the correct size distribution and use it to do the calculation with KAPA results.

                Comment

                • chuanwu
                  Junior Member
                  • May 2018
                  • 8

                  #9
                  ikripp, have you had a successful run recently? What is your loading concentration with how much % of PhiX? Thanks

                  Comment

                  • ikripp
                    Member
                    • Jan 2018
                    • 19

                    #10
                    Originally posted by chuanwu View Post
                    ikripp, have you had a successful run recently? What is your loading concentration with how much % of PhiX? Thanks
                    We do approximately a run each week and vary rarely do we have what is considered an unsuccessful run. I base the loading concentration off of the pool concentrations as determined by a HSD1000 tape run on a tapestation 4200, again previous run pools. So its hard to say what loading concentration I have gone with but our standard PhiX is 10% which changes when we have a unique pool. I work with 16S amplicon pools for reference.

                    Comment

                    • chuanwu
                      Junior Member
                      • May 2018
                      • 8

                      #11
                      What is the loading concentration ( ng/ul) you got based on tape station?

                      Comment

                      • ikripp
                        Member
                        • Jan 2018
                        • 19

                        #12
                        Originally posted by chuanwu View Post
                        What is the loading concentration ( ng/ul) you got based on tape station?
                        5pM made from the 20pM library as per Illumina protocol.

                        Comment

                        • ArciMol
                          Member
                          • Apr 2014
                          • 10

                          #13
                          Hi everyone!
                          Here I was searching about failed runs on MiSeq, and found this thread
                          In my case, we do runs on MiSeq each month, and recently we've had problems with 2 runs (MiSeq v3, 600c). In the first failed run, it was a problem of focus ("Best focus is too near edge of range(...)"); on the 2nd failed run, it has no clustering ("No usable signal found in the images"). I'll post lot number, in case it helps someone else:

                          MiSeq Reagent v3 600c box 1 of 2: LOT 20183241 and LOT 20270202
                          MiSeq Reagent v3 box 2 of 2: LOT 20178686 and LOT 20262956

                          We've also had the same answer from our local Illumina manager ... something like "it's your libraries fault" or "did you even load library on cartridge?". As we're no newbies, these answers doesn't help
                          Has someone else had failed runs with MiSeq, or reagents v3?
                          Science is ok, but I'm hungry.

                          Comment

                          • ikripp
                            Member
                            • Jan 2018
                            • 19

                            #14
                            Originally posted by ArciMol View Post
                            Hi everyone!
                            Here I was searching about failed runs on MiSeq, and found this thread
                            In my case, we do runs on MiSeq each month, and recently we've had problems with 2 runs (MiSeq v3, 600c). In the first failed run, it was a problem of focus ("Best focus is too near edge of range(...)"); on the 2nd failed run, it has no clustering ("No usable signal found in the images"). I'll post lot number, in case it helps someone else:

                            MiSeq Reagent v3 600c box 1 of 2: LOT 20183241 and LOT 20270202
                            MiSeq Reagent v3 box 2 of 2: LOT 20178686 and LOT 20262956

                            We've also had the same answer from our local Illumina manager ... something like "it's your libraries fault" or "did you even load library on cartridge?". As we're no newbies, these answers doesn't help
                            Has someone else had failed runs with MiSeq, or reagents v3?
                            My experience with "best focus" issues is that it does come down to an over-clustered run, although we have a very over-clustered run on the machine at the moment and it hasn't failed - it just has poor QC metrics. I have had runs with lower clustering (still over-clustered) that have failed yet this one has persisted.

                            I've never seen an error of no usable signal. What do the thumbnail images look like?

                            Comment

                            • ArciMol
                              Member
                              • Apr 2014
                              • 10

                              #15
                              Originally posted by ikripp View Post
                              My experience with "best focus" issues is that it does come down to an over-clustered run, although we have a very over-clustered run on the machine at the moment and it hasn't failed - it just has poor QC metrics. I have had runs with lower clustering (still over-clustered) that have failed yet this one has persisted.

                              I've never seen an error of no usable signal. What do the thumbnail images look like?
                              Hi ikripp,
                              in the case of "no usable signal", we didn't even have thumbnail images, but we do have folder "Images", and there's a "Focus" -> "C1" folder, with .tif images file.
                              Here an example of name: FocusImage_0_exp_200_Focus T_IM_zpos_-0.15_method_CoarseFocus.tif

                              It looks like a huge black square. Literally no signal
                              Science is ok, but I'm hungry.

                              Comment

                              Latest Articles

                              Collapse

                              • SEQadmin2
                                16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                                by SEQadmin2




                                NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                                10-30-2026, 06:46 AM
                              • SEQadmin2
                                Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                                by SEQadmin2



                                Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                                Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                                ...
                                10-14-2026, 09:55 AM

                              ad_right_rmr

                              Collapse

                              News

                              Collapse

                              Topics Statistics Last Post
                              Started by SEQadmin2, 10-28-2026, 09:57 AM
                              0 responses
                              40 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-28-2026, 09:51 AM
                              0 responses
                              23 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 11:13 AM
                              0 responses
                              73 views
                              0 reactions
                              Last Post SEQadmin2  
                              Started by SEQadmin2, 10-22-2026, 10:48 AM
                              0 responses
                              48 views
                              0 reactions
                              Last Post SEQadmin2  
                              Working...