    Illumina second read quality drop

    Hello,
    For several months now, we observed on our GAIIx that the intensities for the second read in a paired end run (V4 kits) are often below the intensities of the last cycle of the first read. Before we always obtained a first cycle intensity for read 2 between 100 to 70% of the the first cycle intensity for read 1 (as it is normally expected).
    We have had some extreme runs where the 1st cycle intensity of read 1 is 386 and for the second read it drop at 53 for the phiX lane. The engineer from Illumina checked already several times the peltier and re calibrated its temperature and alignment. Until now the problem isn't solved.
    The quality of the results for read 2 for the phiX lane still fall within the specifications of Illumina on % error Rate but are typically twice as high as the error rate for read 1.
    I checked all the runs we did before this phenomenon started and we always observed almost exactly the same error rate for read 1 and 2.
    As the results are within specifications Illumina pretends their is no problem, but we have the very strong feeling that the machine is running below optimal.
    Does anyone else has observed similar problems and if yes how were they solved?
  • josdegraaf
    Member
    • Mar 2010
    • 33

    #2
    Did you think about the second read primer hyb, maybe something with the PE module?

    Comment

    • ScottC
      Senior Member
      • Jan 2008
      • 244

      #3
      This happened to us once, and our FAS has told us that they are investigating a cause for this type of situation. I'll keep you posted if I hear anything.

      Comment

      • rgrande
        Junior Member
        • Sep 2009
        • 2

        #4
        rgrande

        Did the Illumina’s FAS verified the intensity of laser beams of the device? We had a similar problem with our instrument and it was solved by changing the laser engine because one of the two laser beams intensity was low (which identifies G and T bases).

        Comment

        • rgrande
          Junior Member
          • Sep 2009
          • 2

          #5
          Did the Illumina’s FAS verified the intensity of laser beams of the device? We had a similar problem with our instrument and it was solved by changing the laser engine because one of the two laser beams intensity was low (which identifies G and T bases).

          Comment

          Latest Articles

          Collapse

          • SEQadmin2
            16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
            by SEQadmin2




            NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
            Today, 06:46 AM
          • SEQadmin2
            Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
            by SEQadmin2



            Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

            Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
            ...
            10-14-2026, 09:55 AM

          ad_right_rmr

          Collapse

          News

          Collapse

          Topics Statistics Last Post
          Started by SEQadmin2, 10-28-2026, 09:57 AM
          0 responses
          24 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-28-2026, 09:51 AM
          0 responses
          15 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 11:13 AM
          0 responses
          68 views
          0 reactions
          Last Post SEQadmin2  
          Started by SEQadmin2, 10-22-2026, 10:48 AM
          0 responses
          48 views
          0 reactions
          Last Post SEQadmin2  
          Working...