    MiSeq v3 600

    Hi everyone,

    we have troubles with our MiSeq Runs. At about cycle 80 we have a massive decrease in Q30scores.In general the quality of the runs is not good. Q30 scores of 50 or lower. In addition, during my last two runs- the intensity does not increase . I am very new to this technique and have therefore attached some Plots from basespace.
    Maybe someone had the same problem? Or has an idea about what is going wrong- what are the possible causes? I went through the troubleshooting guides, but could find yet any answer.

    thanks :-)
    Attached Files
  • GenoMax
    Senior Member
    • Feb 2008
    • 7142

    #2
    What kind of libraries are these? Are you spiking in phiX and if so at what %? Do you expect to have low nucleotide diversity after cycle 80 in sequencing? This run is also borderline overclustered and that could be one the reasons for this.

    Comment

    • HESmith
      Senior Member
      • Oct 2009
      • 512

      #3
      The run is significantly over-clustered, as indicated by the low PF (only 68%; should be mid-80s to mid-90s) in combination with the high error rate (4.4%; should be <1%). Quantify your libraries by qPCR and load at the recommended concentration (actually, a bit lower for PE-300bp runs).

      Comment

      • thermophile
        Senior Member
        • Apr 2015
        • 243

        #4
        Are you sure your template is more than 80bp?
        Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

        Comment

        • MU Core
          Member
          • Apr 2008
          • 60

          #5
          Is there a possibility of adapter dimers? May need to perform an additional library clean-up.

          Comment

          • HESmith
            Senior Member
            • Oct 2009
            • 512

            #6
            Short templates/adapter dimers produce a drop-off in signal intensity, which @ATGCT doesn't observe (first thumbnail). And adapter dimers produce a diagnostic spiky signal, also not observed.

            Comment

            • nucacidhunter
              Jafar Jabbari
              • Jan 2013
              • 1250

              #7
              I wonder if you can post %Base from Data by cycle pan.

              Comment

              • ATGCT
                Junior Member
                • Oct 2017
                • 4

                #8
                Hi, thanks for your answers !!!
                We are doing Rad-seq libraries. Yes, I am quite sure, that my templates are bigger than 80bp- but we are checking this now using the bioanalyzer kit.
                We already contacted Illumina, and they said, that we should increase PhiX concentration and lower our library concetration (from 8pM)- I think because of low diversity. I attached now % Base- at first it looks like problem of low intensities- then I dont know what happens
                Attached Files

                Comment

                • ATGCT
                  Junior Member
                  • Oct 2017
                  • 4

                  #9
                  Ah yes @ GenoMax - we spike with 5% PhiX. I thin we should increase to 10%

                  Comment

                  • nucacidhunter
                    Jafar Jabbari
                    • Jan 2013
                    • 1250

                    #10
                    Two issues:
                    1- low sequence diversity at the start which has resulted in low PF%. This should not happen if following standard RAD-seq protocol

                    2- short inserts and adapter-dimers that start reading into adapter in cycles 80-120

                    Both can be fixed by more careful library prep.

                    Comment

                    • GenoMax
                      Senior Member
                      • Feb 2008
                      • 7142

                      #11
                      @ATGCT: You should take this data and use BBMap to estimate the insert size. You can do it in different ways. See this link for examples. You likely have very short inserts as @nucacidhunter has said. This will just confirm that observation.

                      Comment

                      • ATGCT
                        Junior Member
                        • Oct 2017
                        • 4

                        #12
                        Hi, thanks :-) I will have a look at it. But the insert size should be around 500 bp - I cut them out of a Agarose gel - so I have seen the size- unless something happens after gel extraction. We will have our Bioanalyzer results soon. Then I can also have a look at that. But I will try also BBMap.
                        Thanks for your suggestions :-) !!!

                        Comment

                        • GenoMax
                          Senior Member
                          • Feb 2008
                          • 7142

                          #13
                          Don't be surprised. Keep in mind that small inserts preferentially cluster so if they are in there they are attaching to flowcell first.

                          Comment

                          • thermophile
                            Senior Member
                            • Apr 2015
                            • 243

                            #14
                            your %base also suggest short fragments and reading into the flowcell.
                            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                            Comment

                            Latest Articles

                            Collapse

                            • SEQadmin2
                              Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                              by SEQadmin2



                              Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                              Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                              ...
                              10-14-2026, 09:55 AM
                            • SEQadmin2
                              New Genomics Technologies Take Aim at Long-Standing Limits
                              by SEQadmin2


                              Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                              We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                              ...
                              09-28-2026, 10:25 AM

                            ad_right_rmr

                            Collapse

                            News

                            Collapse

                            Topics Statistics Last Post
                            Started by SEQadmin2, Yesterday, 10:52 AM
                            0 responses
                            9 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-14-2026, 05:39 AM
                            0 responses
                            19 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-13-2026, 09:56 AM
                            0 responses
                            16 views
                            0 reactions
                            Last Post SEQadmin2  
                            Started by SEQadmin2, 10-09-2026, 11:14 AM
                            0 responses
                            27 views
                            0 reactions
                            Last Post SEQadmin2  
                            Working...