We have recently been experiencing problems with the amplifcation manifolds from Illumina in their ability to form a tight seal on the cluster station. The result has been decreased reagent flow and very low cluster density when sequencing. Has anyone else been experiencing this issue and has anyone found ways to fix the problem other than contacting Illumina directly? Thanks.
Unconfigured Ad
Collapse
X
-
We experienced the similar problem before.
Try to tighten fluid tubes against the rubber adaptors before using the manifold, it will help. We observed those connection were sometimes not tight enough and the cluster station will pump air occasionally. By tighten the connection manually, we solved the problem.
We have a cBot now, haven't experienced the same problem since then.
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification,...-
Channel: Articles
10-30-2026, 06:46 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Started by SEQadmin2, 11-17-2026, 06:27 AM
|
0 responses
24 views
0 reactions
|
Last Post
by SEQadmin2
11-17-2026, 06:27 AM
|
||
|
Started by SEQadmin2, 11-12-2026, 07:59 AM
|
0 responses
39 views
0 reactions
|
Last Post
by SEQadmin2
11-12-2026, 07:59 AM
|
||
|
Started by SEQadmin2, 11-05-2026, 09:25 AM
|
0 responses
50 views
0 reactions
|
Last Post
by SEQadmin2
11-05-2026, 09:25 AM
|
||
|
Started by SEQadmin2, 11-03-2026, 11:54 AM
|
0 responses
79 views
0 reactions
|
Last Post
by SEQadmin2
11-03-2026, 11:54 AM
|
Comment