    OverClustering - MiSeq v2 reagents - Nextera XT Library Prep

    Hi everyone,

    I am a new user of MiSeq. I am using the Nextera XT Library Prep for sequencing of gDNA along with the MiSeq Reagents v2.

    I am having issues with overclustering, and the Q30 was pretty low (58.3%). However, the Cluster Passing Filter % was still around 79%. The images did not look terribly overclustered tough.

    I have heard that reviewing parameters on Reads 1, 2, 3 and 4 can help understand what happened but I don't know how to interpret those parameters.
    I noticed that for Reads 1 and 4, the Q30 is low : 63%, and 51.6% respectively. However, for Reads 2, and 3, it's pretty good: 90.4%, and 90.8% respectively. What does that mean when the error rates is high for Reads 1 and 4???

    Thanks so much for your help!!!!
  • finswimmer
    Member
    • Oct 2016
    • 60

    #2
    Hello,

    Read 2 und 3 are the index read. As they have just a few cycles the Q30 is alway higher than in Read 1 und 4.

    What was the Cluster density? Have you had better runs before? What kind of gDNA do you try to sequence (human?) ? Maybe you have low diversity in your target regions, which makes it hard for the MiSeq to distiginguish the clusters . Spike in phiX might help.

    fin swimmer

    Comment

    • hopedlamini
      Junior Member
      • Dec 2017
      • 1

      #3
      Hi everyone, I have sequenced bacterial DNA using Illumina MiSeq, and have both forward and reverse reads, after quality trimming, the forward read has high GC content, is there any software of trimming this? Thanks

      Comment

      • clseq2017
        Junior Member
        • Dec 2017
        • 4

        #4
        Originally posted by finswimmer View Post
        Hello,

        Read 2 und 3 are the index read. As they have just a few cycles the Q30 is alway higher than in Read 1 und 4.

        What was the Cluster density? Have you had better runs before? What kind of gDNA do you try to sequence (human?) ? Maybe you have low diversity in your target regions, which makes it hard for the MiSeq to distiginguish the clusters . Spike in phiX might help.

        fin swimmer

        Hello,
        The cluster density was very high 1600 K/mm2 (supposed to be 1000-1200 K/mm2 for V2 reagents).
        Yes, I have had better runs but the cluster density has always been in the upper limit. I am sequencing bacterial gDNA from several species.
        Maybe this is low diversity. How can I double check that in the run parameters?

        Thanks!

        Comment

        • clseq2017
          Junior Member
          • Dec 2017
          • 4

          #5
          Nucleotide Diversity

          Hello,

          You mentioned "low diversity" library but looking at some troubleshooting guidelines it looks like my library is unbalanced (see attached picture).
          What is an unbalanced library? How can I fix it?

          Thanks!
          Attached Files

          Comment

          • GenoMax
            Senior Member
            • Feb 2008
            • 7142

            #6
            You are way over spec for the cluster density. I am surprised you were able to get any data at all. I would not trust the data you have if it is to be used for any type of de novo work. You should think about re-running this at half concentration.

            Comment

            • clseq2017
              Junior Member
              • Dec 2017
              • 4

              #7
              Originally posted by GenoMax View Post
              You are way over spec for the cluster density. I am surprised you were able to get any data at all. I would not trust the data you have if it is to be used for any type of de novo work. You should think about re-running this at half concentration.

              I agree with you! But Illumina looked at the run and they said the parameters are still acceptable. However, I want to repeat and fix this overclustering issue. But my questions are : how do I avoid overclustering???? and how do I lower the loading concentration? since I am using Nextera XT that does the bead-normalization.

              Also, I checked the pooled library and it looks like there is nothing showing up! so I am very confused!!!
              Our bioinformatician said "The 3’-end quality is quite poor for both the R1 and R2 and there are strong 3’-end sequence biases".
              Why is the quality poor on 3'-end? Any explanation?

              Sorry for all those questions, but it's my first MiSEq failed run!!!

              Thanks!

              Comment

              • finswimmer
                Member
                • Oct 2016
                • 60

                #8
                Hello,

                to your question about how to avoid overclustering: What was the concentration of your Library at the end of of the LibPrep? Have you measure it with qubit? What is the average fragment size? And with how many pM have you load the flow cell?

                fin swimmer

                Comment

                • Genetic Librarian
                  Member
                  • May 2017
                  • 31

                  #9
                  Originally posted by clseq2017 View Post
                  and how do I lower the loading concentration?
                  During the denature and dilute process you dilute your library to the final loading concentration with the provided buffer based on your final Qubit quantification of the library pool. So, instead of diluting it to e.g. 10 pM loading concentration, try 6 pM next time.

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    Originally posted by clseq2017 View Post
                    I agree with you! But Illumina looked at the run and they said the parameters are still acceptable.
                    It is interesting if they actually said that. Perhaps they did not want to spook you

                    Our bioinformatician said "The 3’-end quality is quite poor for both the R1 and R2 and there are strong 3’-end sequence biases".
                    Why is the quality poor on 3'-end? Any explanation?
                    That is to be expected. When a run is overclustered the clusters are very close to begin with. Over the run they become fatter and then the software has a tough time telling them apart. You could easily lose a run due to this. In your case though since this is mix of genomic DNA's the run was able to end successfully but the Q-scores suffered when the reliability of basecalls dropped towards the end of the run.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                      by SEQadmin2




                      NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                      10-30-2026, 06:46 AM
                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-28-2026, 09:57 AM
                    0 responses
                    45 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-28-2026, 09:51 AM
                    0 responses
                    27 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 11:13 AM
                    0 responses
                    75 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-22-2026, 10:48 AM
                    0 responses
                    49 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...