Hello. I am using Nextera XT to do whole genome sequencing of bacteria grown in stress conditions/high concentrations of antibiotics. What I found is, even if I follow the protocol correctly, I end up with a lowi(ish) cluster density (about 140K/mm2) and also The PF and Q30 reads are about 70%. I wonder what i coudl do to improve both cluster density and reads quality. If you have any tips please shout!
Unconfigured Ad
Collapse
X
-
From your post I can't be sure, but it sounds like you're sequencing your Nextera libraries on a NextSeq. If your loading concentration is resulting in 140K/mm2 densities, you can start by targeting a higher density (generally aim for 200K/mm2) with the formula [your starting concentration] x (target density)/(observed density) to get a new concentration to start with. I.E. if I loaded a pool at 1.7pM and observed 150K/mm2, I would multiply 1.7*(200/150) = 2.27pMwould be my new target loading concentration. To play it safe you could target the low end of the recommended density range (180K/mm2) but that is your choice.
As far as read quality, increasing the density without overclustering should help. The other obvious improvement would be spiking in a higher percentage of an ATCG balanced genome library will also help with basecalling. Illumina sells non-indexed PhiX to serve this purpose, but spiking in something easily separated from your bacterial samples might be a good idea.
-
These two issues can have common or separate causes. To diagnose the issue it will be helpful if you can post run summary page (Reads tables) and also %Base from data by cycle plot of analysis page of SAV.Originally posted by thiNGS View PostHello. I am using Nextera XT to do whole genome sequencing of bacteria grown in stress conditions/high concentrations of antibiotics. What I found is, even if I follow the protocol correctly, I end up with a lowi(ish) cluster density (about 140K/mm2) and also The PF and Q30 reads are about 70%. I wonder what i coudl do to improve both cluster density and reads quality. If you have any tips please shout!
Comment
-
What's the %GC of these organisms? you may need to spike in a lot of PhiX(20%) if it's highMicrobial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
Comment
-
Comment
-
We've had poor assembly with nextera genomes above 62%, but decent assembly with the same genomes prepped with truseq.
I've never run high gc on nextseq, just have been told that we need to spike in at least 20% phiX (compared to 5% recommended for high gc on miseq/hiseq)Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
Comment
-
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...-
Channel: Articles
10-30-2026, 06:46 AM -
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, 10-28-2026, 09:57 AM
|
0 responses
41 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:57 AM
|
||
|
Started by SEQadmin2, 10-28-2026, 09:51 AM
|
0 responses
25 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
74 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
48 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Comment