  • Customprimers-SampleSheet error during secondary analysis

    Hi everyone,

    Yesterday i did a 16s MiSeq run using custom primers, when i tried to demultiplex using samplesheet generated through IEM, i found error during secondary analysis to generate Fastq files. I replaced the illumina index sequences with custom index sequences during samplesheet generation. But still I could not overcome the issue. It would be great help, if any one can help to sort this issue. Many thanks in advance.

  • #2
    What was the error?

    --
    Phillip

    Comment


    • #3
      Dear Phillip, Sorry I was not clear in my message. The error was during the demultiplex execution according to the analysis error log. When MSR executes the secondary analysis, the error message was "aborting analysis. Please check error log and sample sheet issues". For your information please find the below link for sample sheet. Thank you.
      https://www.dropbox.com/s/6z28l7bx1n...Sheet.csv?dl=0

      Comment


      • #4
        Nothing leaps out at me as being wrong with that sample sheet. You will probably need to take this up with Illumina.

        --
        Phillip

        Comment


        • #5
          Well, one thing. You mentioned you changed the sample sheet after the run. What sample sheet was used by the MiSeq during the run?

          The length of the i7 index read is determined by the length of the sequences in the sample sheet. So if you did not have 12 base indexes in the run sample sheet, then that index read would be whatever length the indexes you did use were. And, if you didn't provide any indexes, then no index read would have been done and it would likely be impossible to demultiplex you run.

          --
          Phillip

          Comment


          • #6
            Following up @Phillip's suggestion, post this part from your ReadInfo.xml file for this run. It should look something like this
            Code:
            <Read Number="1" NumCycles="151" IsIndexedRead="N"/>
            <Read Number="2" NumCycles="8" IsIndexedRead="Y"/>
            <Read Number="3" NumCycles="152" IsIndexedRead="N"/>

            Comment


            • #7
              Look for WorkflowLog.txt (I demultiplex in basespace where it's in fastqc/files), scroll to the bottom and it will tell you exactly what has caused the sample sheet failure.
              Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

              Comment


              • #8
                I'm experiencing a similar problem. Was this resolved?

                I ran a MiSeq run with 16s custom libraries/primers and got the same error "abort analysis. Please check error log and sample sheet." No fastq files were produced for my run, but the overall quality of the run was good. I also noticed the base call intensity for the i7 read was pretty low. Could this be a software issue or maybe a primer issue? PhiX was also spiked in at 15%.
                Last edited by athuyvo; 03-13-2018, 04:08 PM. Reason: added more details

                Comment


                • #9
                  Here's how my header looks for custom amplicon sequencing.

                  Code:
                  Workflow	GenerateFASTQ
                  Application	FASTQ Only
                  Assay	Nextera
                  Description	
                  Chemistry	Amplicon
                  Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                  Comment


                  • #10
                    Thanks @thermophile. I have same exact format for my sample sheet. I was able to requeue the analysis by essentially ignoring the i7 reads and demultiplex using only i5 barcodes. We think the problem was suboptimal custom i7 primers so the MiSeq couldn't demultiplex using i7 reads.

                    Comment


                    • #11
                      you reverse complimented your i7 barcodes right?
                      Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

                      Comment


                      • #12
                        i7 barcodes were reverse complimented. We were able to salvage our data by mining the adapter reads and found the i7 barcodes.

                        Comment

                        Latest Articles

                        Collapse

                        • seqadmin
                          Best Practices for Single-Cell Sequencing Analysis
                          by seqadmin



                          While isolating and preparing single cells for sequencing was historically the bottleneck, recent technological advancements have shifted the challenge to data analysis. This highlights the rapidly evolving nature of single-cell sequencing. The inherent complexity of single-cell analysis has intensified with the surge in data volume and the incorporation of diverse and more complex datasets. This article explores the challenges in analysis, examines common pitfalls, offers...
                          Today, 07:15 AM
                        • seqadmin
                          Latest Developments in Precision Medicine
                          by seqadmin



                          Technological advances have led to drastic improvements in the field of precision medicine, enabling more personalized approaches to treatment. This article explores four leading groups that are overcoming many of the challenges of genomic profiling and precision medicine through their innovative platforms and technologies.

                          Somatic Genomics
                          “We have such a tremendous amount of genetic diversity that exists within each of us, and not just between us as individuals,”...
                          05-24-2024, 01:16 PM

                        ad_right_rmr

                        Collapse

                        News

                        Collapse

                        Topics Statistics Last Post
                        Started by seqadmin, Today, 08:18 AM
                        0 responses
                        8 views
                        0 likes
                        Last Post seqadmin  
                        Started by seqadmin, Today, 08:04 AM
                        0 responses
                        10 views
                        0 likes
                        Last Post seqadmin  
                        Started by seqadmin, 06-03-2024, 06:55 AM
                        0 responses
                        13 views
                        0 likes
                        Last Post seqadmin  
                        Started by seqadmin, 05-30-2024, 03:16 PM
                        0 responses
                        27 views
                        0 likes
                        Last Post seqadmin  
                        Working...
                        X