    Running one sample with index primers?

    Hi all,
    I have a question - I am going to prepare library and I want try it with single sample and do a test run on MiSeq. The poblem is that I have just NEBNext Multiplex Oligos which are supposed to be multiplexed. Will the sequencing work if I add PhiX?
  • nucacidhunter
    Jafar Jabbari
    • Jan 2013
    • 1250

    #2
    That will work. You would have option of setting up run with index read or without.

    Comment

    • thermophile
      Senior Member
      • Apr 2015
      • 243

      #3
      Only one index could cause the run to fail during the index sequencing because there isn't any diversity in each cycle. I'd just set up the sample sheet with 0 index
      Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

      Comment

      • minio.cz
        Junior Member
        • Jul 2015
        • 6

        #4
        Thank you both. It should be obvious not to read indexes, but my brain is apparently not working lately

        Comment

        • nucacidhunter
          Jafar Jabbari
          • Jan 2013
          • 1250

          #5
          Originally posted by thermophile View Post
          Only one index could cause the run to fail during the index sequencing because there isn't any diversity in each cycle. I'd just set up the sample sheet with 0 index
          A single index sequencing on HiSeq systems is fine and on average 98% of reads correctly demultiplex. I wonder if you have seen single index sequencing fail on MiSeq or speculating.

          Edit: If PhiX added to run, sequencing with index read reduces follow up work required for removing PhiX reads in some workflows.
          Last edited by nucacidhunter; 03-16-2018, 03:29 PM.

          Comment

          • thermophile
            Senior Member
            • Apr 2015
            • 243

            #6
            Yes I've had runs with 50% phiX fail during indexing on the MiSeq-I'd guess something similar would happen if you only use one index (3/4 colors would be dark and the machine would likely throw the "focus out of range" error). OP said they were going to run this on the miseq
            Microbial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.

            Comment

            • minio.cz
              Junior Member
              • Jul 2015
              • 6

              #7
              Originally posted by nucacidhunter View Post
              A single index sequencing on HiSeq systems is fine and on average 98% of reads correctly demultiplex. I wonder if you have seen single index sequencing fail on MiSeq or speculating.
              I am speculating. I am testing library prep and would like to test library first on MiSeq if it contains what I want. But have only multiplexing oligos.

              Comment

              • GenoMax
                Senior Member
                • Feb 2008
                • 7142

                #8
                If you underload the library deliberately the run should be able to complete even with just one index. I am reasonably sure we have done this in the past but will have to confirm.

                Comment

                • minio.cz
                  Junior Member
                  • Jul 2015
                  • 6

                  #9
                  Ok, final result. Turning off index read really works. Sequenced library without problem. In the end no PhiX has been added.

                  Comment

                  • GenoMax
                    Senior Member
                    • Feb 2008
                    • 7142

                    #10
                    No surprise since index reads are never part of the actual sequence in Illumina technology.

                    Comment

                    Latest Articles

                    Collapse

                    • SEQadmin2
                      Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                      by SEQadmin2



                      Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                      Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                      ...
                      10-14-2026, 09:55 AM
                    • SEQadmin2
                      New Genomics Technologies Take Aim at Long-Standing Limits
                      by SEQadmin2


                      Researchers using sequencing and genomics tools often have to make trade-offs. They can choose between speed or scale, short reads or long-range information, or targeted panels or a view of the whole transcriptome. New technologies that have been released this year are built to address those tough choices.

                      We asked six companies the same four questions to learn about their latest products. The new technologies bring a lot to the table, including rethinking sequencing
                      ...
                      09-28-2026, 10:25 AM

                    ad_right_rmr

                    Collapse

                    News

                    Collapse

                    Topics Statistics Last Post
                    Started by SEQadmin2, 10-15-2026, 10:52 AM
                    0 responses
                    18 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-14-2026, 05:39 AM
                    0 responses
                    26 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-13-2026, 09:56 AM
                    0 responses
                    22 views
                    0 reactions
                    Last Post SEQadmin2  
                    Started by SEQadmin2, 10-09-2026, 11:14 AM
                    0 responses
                    28 views
                    0 reactions
                    Last Post SEQadmin2  
                    Working...