No I did not know that those tests even exist...
Unconfigured Ad
Collapse
X
-
Illumina tech support will entertain questions from end-users (so you have done the right thing) but you would need your partner to be involved if they need remote access to the sequencer to check this run/require the partner to run the diagnostic tests.Originally posted by SarahAurora View PostIt's not our MiSeq, we just use it from our cooperation partner. I'll check if they have a maintenance contract with Illumina. Nevertheless, I wrote tech support an email with all details.
Comment
-
Onboard diagnostics are in the "System Check" icon under "Manage Instrument" from the main screen on the Miseq Control Software. The optics checks are fairly fast to run, the fluidics will take some time and babysitting (raising and lowering sippers as prompted).
Has your partner had any successful runs since your failed run?
Edit: Asking me to run these tests is usually what Tech Support has me do first when I call them with an issue. Enough so that I'll sometimes run them prior to calling...
Comment
-
-
If this failure is due to hardware/reagent problem (and the partner has a maintenance contract with Illumina) then Illumina generally will replace the reagents at no cost, so you should be able to re-run your samples for free (or at least no charge for the reagents).
Comment
-
No, we are the only one using the MiSeq. We already used it for 8 runs, but the ninth one was now the fatal one....Originally posted by GW_OK View PostOnboard diagnostics are in the "System Check" icon under "Manage Instrument" from the main screen on the Miseq Control Software. The optics checks are fairly fast to run, the fluidics will take some time and babysitting (raising and lowering sippers as prompted).
Has your partner had any successful runs since your failed run?
Comment
-
That would be great!Originally posted by GenoMax View PostIf this failure is due to hardware/reagent problem (and the partner has a maintenance contract with Illumina) then Illumina generally will replace the reagents at no cost, so you should be able to re-run your samples for free (or at least no charge for the reagents).
Comment
-
I run primarily amplicons too so the metrics that I look for are a bit different than the standards that Illumina uses (our runs don't meet the base diversity that makes some of the standard number make sense).
From the %q30 it looks like something bad happened part way through R1-amplicons aren't really long enough, optics issue, fluidics issue. Whatever it was *seems* to be machine/reagent related because it didn't improve after the turnaround. Are you spiking in primers or do your amplicons use Illumina sequencing primers? If you make an error spiking in the primers, you won't see anything after the turn around.
My list to check (roughly in order)
1. %phiX aligned. is it close to what I thought I was spiking in?
2. %base does it look appropriately horrible (my 16s amplicons should be at 90% for the first ~50 bases, then decline to ~80% for the rest of the read)
3. Cluster density across flowcell, top and bottom
4. thumbnails, does the density seem about right
5. fwmh. The software is not made for 90% single base, focus issues are common, but fwmh shouldn't be too jaggedMicrobial ecologist, running a sequencing core. I have lots of strong opinions on how to survey communities, pretty sure some are even correct.
Comment
Latest Articles
Collapse
-
by SEQadmin2
NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...-
Channel: Articles
Yesterday, 06:46 AM -
-
by SEQadmin2
Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?
Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating...-
Channel: Articles
10-14-2026, 09:55 AM -
ad_right_rmr
Collapse
News
Collapse
| Topics | Statistics | Last Post | ||
|---|---|---|---|---|
|
Genome Sequencing Reveals Large, Repeat-Rich DNA in Trapdoor Spider Named for Stephen Colbert
by SEQadmin2
Started by SEQadmin2, 10-28-2026, 09:57 AM
|
0 responses
29 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:57 AM
|
||
|
Started by SEQadmin2, 10-28-2026, 09:51 AM
|
0 responses
21 views
0 reactions
|
Last Post
by SEQadmin2
10-28-2026, 09:51 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 11:13 AM
|
0 responses
71 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 11:13 AM
|
||
|
Started by SEQadmin2, 10-22-2026, 10:48 AM
|
0 responses
48 views
0 reactions
|
Last Post
by SEQadmin2
10-22-2026, 10:48 AM
|
Comment