Unconfigured Ad

Collapse
X
 
  • Time
  • Show
Clear All
new posts
  • JakobHedegaard
    Member
    • Mar 2008
    • 62

    #1

    NovaSeq index issue (vs MiSeq)

    Hi,
    We have conducted a NovaSeq S1 run of different library types and noticed a lower %perfect reads from a single sample as compared to a MiSeq QC run of the same libraries.
    The libraries were a mixture of 21x QIAseq FX, 9x KAPA RNA Hyper libraries (both with 8 bp dual indexes) and 36x KAPA Hyper DNA libraries prepared with NextFlex48 single indexes from BiooScientific (cat#514104).
    We notice that a single KAPA Hyper DNA library with index sequence GTAGAGAT+TCTTTCCC (Barcode Adapter 17) were found to have 76 % perfect index match on the NovaSeq while it had 97.8 % perfect index match on the MiSeq. The insert sequences from the NovaSeq were furthermore much more noisy from this specific library.
    All other KAPA Hyper DNA libraries performed fine on both MiSeq and NovaSeq.
    The NextFlex48 indexes are single 6 bp index but can be read as 8 bp dual indexes resulting in an additional and universal terminal AT and an universal TCTTTCCC index 2 from all KAPA Hyper DNA libraries (see the attached pdf).

    Wonder why a single library can behave so different on a MiSeq and NovaSeq- while similar libraries essentially only differing in the index sequence behave identical.
    I have asked the same question to Illumina Tech Support and was asked to pass the question on to BiooScientific. I have done so, but wonder if any of you experienced NovaSeq users have encountered similar issues with specific indexes.
    Best, /Jakob
    Attached Files
    Last edited by JakobHedegaard; 01-15-2019, 05:00 AM. Reason: Updating the header
  • JakobHedegaard
    Member
    • Mar 2008
    • 62

    #2
    Come on - I have heard several rumors about indexes behaving non-optimal on the NovaSeq. Please share your experiences :-)

    Comment

    • GenoMax
      Senior Member
      • Feb 2008
      • 7142

      #3
      MiSeq is a sequencing champ as far as illumina libraries are concerned. We look at results from MiSeq nano runs only qualitatively (rather than quantitatively) before the pools are transferred for final NovaSeq runs. You always need to keep in mind G = no signal limitation of 2-color chemistry as well.

      If you are consistently observing that index combination performing badly then BioO would need to know that.

      Comment

      • pmiguel
        Senior Member
        • Aug 2008
        • 2328

        #4
        You are running non-UDI indexes on the NovaSeq? Are you not concerned about index swapping?

        We demultiplex allowing 0 mismatches for all Illumina runs these days. So I haven't taken note of non-concordant perfect index matches between the MiSeq and NovaSeq.

        However, this is an issue mentioned in the Illumina White Paper on using an iSeq to titrate reactions for a NovaSeq.
        --
        Phillip

        Comment

        • luc
          Senior Member
          • Dec 2010
          • 469

          #5
          I would suggest to check the index reads for this particular library to see which base is affected.

          Comment

          • JakobHedegaard
            Member
            • Mar 2008
            • 62

            #6
            I have been told that two TruSeq UD indexes have shown lower than expected perfect index reads: UDI0025 (ACTAAGAT, AACCGCGG) and UDI0055 (AGGCAGAG, AGAATGCC).
            Anyone else seen this?
            Last edited by JakobHedegaard; 02-12-2019, 11:52 AM.

            Comment

            • JakobHedegaard
              Member
              • Mar 2008
              • 62

              #7
              And in a recent S2 (2x150) NovaSeq run of eight Nextera DNA Flex libraries prepared from human gDNA using indexes from the CD Indexes (96 Indexes, 96 Samples, cat# 20018708) we observed a single library with only 24-27% perfect reads. The index was from well F04, H701 (TAAGGCGA) + H522 (TTATGCGA) and the top non-perfect index was TAAGGCGA+TTTTGCGA (so a T in pos 3 in index2 where it should have been a A).
              The other indexes in the run were find with a high fraction of perfect indexes.
              Other observations of indexes with high fraction of non-perfect indexes?
              We allow a single mismatch so we do get the data we need. We are though a bit worried that the quality may be affected.
              And it would be great with an explanation of this index issue on the NovaSeq!

              Comment

              • JakobHedegaard
                Member
                • Mar 2008
                • 62

                #8
                In our hands, we do not find any significant differences in the quality between insert data with perfect index reads and insert data with 1-mismatch index reads.
                So despite a much larger fraction of 1-mismatch index reads the data is fine. Nice...

                But it would be great with an explanation of how/why this index issue on the NovaSeq actually happen.
                In our data, the issue is more frequent in index2 than in index1.
                Can it have something to do with a more "loose" binding to the surface oligo when priming for reading index2? Or a sequence-specific higher (and systematic) error rate in index2 reading?
                Any ideas?

                And a list of indexes where one should expect a higher fraction of non-perfect indexes would be nice. Should we start one?

                Comment

                • ralyma
                  Junior Member
                  • Aug 2018
                  • 1

                  #9
                  Dear Jakob,

                  We also have observed problems with H522=TTATGCGA. In demultiplexed data (default settings accepting 1-mismatch) 97% of observed adapters are TTTTGCGA.

                  To my knowledge TTTTGCGA is not an Illumina adaptor, so this makes it less likely that we have mixed H522 with another adaptor. Then it is more likely that there has been has been an issue in the production of H522 so that sequence TTTTGCGA has been produced instead of TTATGCGA.

                  Comment

                  Latest Articles

                  Collapse

                  • SEQadmin2
                    16S/ITS Profiling vs. Shotgun Metagenomics: What Changes in Your Workflow?
                    by SEQadmin2




                    NGS has revolutionized clinical microbiology by offering new ways to identify and surveil pathogens without prior knowledge of the target. One popular approach is 16S rRNA and ITS amplicon sequencing, which targets specific microbial markers. But studies are increasingly incorporating shotgun metagenomics, a technique that sequences all DNA from a sample without targeting specific microbial markers. This approach allows more detailed taxonomic classification, detection...
                    10-30-2026, 06:46 AM
                  • SEQadmin2
                    Multiomic Liquid Biopsy: What DNA/RNA Panels Reveal That DNA Alone Can’t
                    by SEQadmin2



                    Tumors contain complex mixtures of cell populations that evolve in response to treatment and host-specific factors. Tissue biopsies provide only static snapshots of a tumor’s state. But what if we could follow that evolution over time using a minimally invasive technique?

                    Liquid biopsies can detect tumor-derived material circulating in the bloodstream, providing a way to study treatment response, detect emerging resistance, and discover new biomarkers. Circulating
                    ...
                    10-14-2026, 09:55 AM

                  ad_right_rmr

                  Collapse

                  News

                  Collapse

                  Topics Statistics Last Post
                  Started by SEQadmin2, 10-28-2026, 09:57 AM
                  0 responses
                  44 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-28-2026, 09:51 AM
                  0 responses
                  26 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 11:13 AM
                  0 responses
                  74 views
                  0 reactions
                  Last Post SEQadmin2  
                  Started by SEQadmin2, 10-22-2026, 10:48 AM
                  0 responses
                  48 views
                  0 reactions
                  Last Post SEQadmin2  
                  Working...