  • Sequencing beyond the end of the fragment

    Hi everyone

    We've had a problem with a sequencing run and I would like to ask for a re-run with justification.

    We submitted 4 libraries to size-select via pippin prep, pool and run SE 150 on a MiSeq. The pippin prep was to select between 200-600bp fragments, and apparently worked well (image of one of the libraries attached, the others looked the same).

    The run came back and it's pretty obvious that something went wrong. between 70-80% of the reads are primer dimers, sized 135bp. I'm attaching the FastQC per-base quality. The first 6bp were in-line barcode. After that there is a big drop in quality at the 77bp mark, which is the end of the primer dimer fragment. However, the fragments keep giving base calls, albeit at very poor quality. Is this what the software does when the reads go beyond the end of the fragment?

    The discrepancy between the reads and the bioanalyzer profile post-pippinprep makes me think that they messed up and loaded the pre-pippinprep sample into the flow cell. But I can also imagine that the bioanalyzer could give a wrong size profile in some conditions, or that there could be some primer dimer contamination that would not be visible on bioanalyzer.

    Does anyone have experience in this kind of problem? Should the read lengths from the MiSeq run be concordant with the bioanalyzer profile, or is that too much to ask?

    Thank you in advance!
    Attached Files
    Last edited by zaratieg; 02-28-2019, 07:01 AM.

  • #2
    The trace that you are presenting here, shows fragments that go beyond 600 bp that are even overlapping with the upper marker. Is this picture from the library before size-selection? That's what it looks like.

    Comment


    • #3
      Thanks for your reply. No, this is supposed to be post size selection. I didn't pay attention to that part, I thought that was their size marker. I saw the well defined bump between 200-600 and assumed that that's what was selected. Does that mean that the pippin prep purification failed?

      Comment

      Latest Articles

      Collapse

      • seqadmin
        Best Practices for Single-Cell Sequencing Analysis
        by seqadmin



        While isolating and preparing single cells for sequencing was historically the bottleneck, recent technological advancements have shifted the challenge to data analysis. This highlights the rapidly evolving nature of single-cell sequencing. The inherent complexity of single-cell analysis has intensified with the surge in data volume and the incorporation of diverse and more complex datasets. This article explores the challenges in analysis, examines common pitfalls, offers...
        Today, 07:15 AM
      • seqadmin
        Latest Developments in Precision Medicine
        by seqadmin



        Technological advances have led to drastic improvements in the field of precision medicine, enabling more personalized approaches to treatment. This article explores four leading groups that are overcoming many of the challenges of genomic profiling and precision medicine through their innovative platforms and technologies.

        Somatic Genomics
        “We have such a tremendous amount of genetic diversity that exists within each of us, and not just between us as individuals,”...
        05-24-2024, 01:16 PM

      ad_right_rmr

      Collapse

      News

      Collapse

      Topics Statistics Last Post
      Started by seqadmin, Today, 08:18 AM
      0 responses
      10 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, Today, 08:04 AM
      0 responses
      12 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 06-03-2024, 06:55 AM
      0 responses
      13 views
      0 likes
      Last Post seqadmin  
      Started by seqadmin, 05-30-2024, 03:16 PM
      0 responses
      27 views
      0 likes
      Last Post seqadmin  
      Working...
      X